Defining Expected Oxygen Leakage for Safety Testing
bassa
I am struggling with the test procedures to determine safety in oxygen rich environments.
Our product has socket outlets, and gas terminals (including oxygen)
Both are placed in separate compartments.
clause 11.2.2.1.b.2 states the following:
- [FONT=&quot]The oxygen concentration is measured for such a period that the highest
[/FONT]
[FONT=&quot]concentration of oxygen occurs. The least favourable control settings are selected.[/FONT][FONT=&quot]The leaking conditions of oxygen are selected such that they provide the minimum[/FONT][FONT=&quot] leak that could be detected by the [/FONT][FONT=&quot]OPERATOR [/FONT][FONT=&quot](e.g. because of a failure of the function[/FONT][FONT=&quot] of the device). If the concentration of oxygen exceeds 25 % in the presence of parts or[/FONT] [FONT=&quot]components that could be a source of ignition including at the moment energy is applied, it constitutes a failure.[/FONT]
The testing laboratory that is testing our product for 60601 3rd edition asks us what the expected oxygen leak rate could be. :confused:
When a leak is only to be detected only by a loss of function, the leak rate could be enormous; it's a supply hose at 8 bar pressure! I can imagine the hose system has a leak that is wildly roaring; while the connected device (needing only 200miliBar) is still working.
Anyone has experience on how to approach this?
Peter Selvey
Because air is already 21% O2, it does not take much of a leak to get up to 25%. The case (b) in the standard would really only work if the source of O2 is very small and the compartment is well ventilated.
Typically in normal condition, high pressure O2 will be confined to the pneumatic system (tubes, solenoid valves etc), so in normal condition OK. To handle the fault condition (e.g. hose disconnects), a typical approach would be to then enclose the pneumatic system inside a separate sealed box away from the main electrical parts. There is some electrical parts in the box (e.g. O2 controller, valves), but they are usually low power and not a source of ignition. Using a metal box would also limit the risk of fire in the worst case.
That approach follows (c) in the standard.
For the final risk management, you could consider (1) undetected failure of the pneumatic system (2) undetected failure of the pneumatic system enclosure, and (3) fault leading to an ignition source in the electronics. It's a fairly remote sequence of events. Also to note that exceeding 25% does not immediately mean "danger danger", it's just that we don't know about the effectiveness of flame retardants at increased O2 levels. Clearly at 90% O2, most don't work, but it could be that at 30 or 40% O2 the flame retardants are still reasonably effective.
cadevon
We are having the same trouble identifying reasonable test conditions for our device (a neonatal CPAP).
We have venting slots in the upper chamber of our device which houses the electronics (approach B from standard), and the pneumatic components are below a lid with an o-ring seal (approach C from standard). There is grating on the bottom of the sealed portion to vent the gas out of the box.
Even with all precautions in place, I believe it's unreasonable to expect to pass if 100 PSI of oxygen is allowed to flow directly into the bottom (sealed) portion. This could occur in SFC where a filter was serviced improperly and the connection subsequently fails.
Is there a precedent for determining the actual pressure to test at? Other steps we could take in order to ensure safety even in this extreme situation (improve seal, fans to vent oxygen out, use of an orifice to limit flow)?
A Defining Expected Service Life in Risk Management File Reliability Analysis - Predictions, Testing and Standards 5
K Defining risk control measures IEC 62304 - Medical Device Software Life Cycle Processes 13
J Defining staff competence - Small mechanical workshop Occupational Health & Safety Management Standards 20
D Question regarding ECO process, specifically for Life Science products and defining form fit and function ISO 13485:2016 - Medical Device Quality Management Systems 1
M Defining frequency of measurement tools callibration Calibration and Metrology Software and Hardware 3
M Defining and Documenting Record Retention CE Marking (Conformité Européene) / CB Scheme 5
G Defining performance metrics for DFMA implementation Design and Development of Products and Processes 2
S Defining a Quality System from scratch - Preferred system and documentation names Document Control Systems, Procedures, Forms and Templates 4
C Defining Approvals Required for Design Control Documents ISO 13485:2016 - Medical Device Quality Management Systems 6
K Defining Acceptance Quality Level, I need clarity on AQL 1.5, 2.5, 4.0 AQL - Acceptable Quality Level 5
M Defining the lifetime of orthopedic implants joints Other Medical Device and Orthopedic Related Topics 2
C AS9100 rev D 8.5.1 c 2 - Defining the Machine in-process frequency per ANSI/ASQ Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
V Defining Safety Precautions for Category 4,5 Molecules Occupational Health & Safety Management Standards 2
E European Regulations defining the terms Repair and Refurbish EU Medical Device Regulations 5
T Defining Major vs. Minor Changes to Procedures ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
GStough Adequately Defining Which Suppliers to Audit and Frequency Supplier Quality Assurance and other Supplier Issues 8
E Quality Techucuan (Technician) in Electronics - Defining Postion Requirements Career and Occupation Discussions 4
moritz Defining a good Scope for Critical SOPs ISO 13485:2016 - Medical Device Quality Management Systems 7
T Standards for defining audible alarms/warnings for OR instruments IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Defining Critical Vs. Non-Critical Suppliers/Service Providers (API Q1, 9th. Ed.) Oil and Gas Industry Standards and Regulations 2
B IEC 60601-2-24 - Defining Storage Volume IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Defining Sub-Disciplines for Chemical Testing Laboratory Employee Proficiency Testing General Measurement Device and Calibration Topics 1
T Defining Nonconformances in a Service Organization ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
M Defining Reliability and Confidence Levels Reliability Analysis - Predictions, Testing and Standards 8
J Defining Martial Arts and Gymnastics Statistical Techniques Statistical Analysis Tools, Techniques and SPC 4
V Defining the criteria for equipment to be qualified or requalified Qualification and Validation (including 21 CFR Part 11) 2
R Need help on defining scope for Design Verification File for Class III IVD 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
J Defining CCP (Critical Control Points) in a Rice Mill Plant Food Safety - ISO 22000, HACCP (21 CFR 120) 9
S Process Map and defining KPIs Misc. Quality Assurance and Business Systems Related Topics 5
5 Major Nonconformance for not "clearly" defining the "device lifetime" ISO 13485:2016 - Medical Device Quality Management Systems 2
E Defining the lifetime of an Implantable Medical Device Other Medical Device and Orthopedic Related Topics 5
G Defining Post Mold Cure Ramp-Down Temperature Manufacturing and Related Processes 2
K Audit Nonconformity on Defining 'Outsourced' Infrastructure Maintenance Quality Manager and Management Related Issues 21
G Points to consider while defining the Quality Policy AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
G Defining Quality Objectives for Product Realization and Design and Development AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
S Developing Documentation and Defining Processes as Subcontractor IATF 16949 - Automotive Quality Systems Standard 6
C Defining ISO 9001:2008 Scope for a Sterilization Company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Defining Skilled vs. Semi-Skilled vs. Unskilled Labor Manufacturing and Related Processes 1
I Defining the scope for ISO 9001 Registration - Software, Hardware and Customer Care ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M Product Specification vs. Information Defining Product - The differences? 7.3.3.1 IATF 16949 - Automotive Quality Systems Standard 6
R Defining Interaction of Processes in a Software Company Software Quality Assurance 3
N Where to begin defining and monitoring Quality Metrics in a Machine Shop Manufacturing and Related Processes 9
A Testing Process Audit - Defining a Process Compliance Mechanism Software Quality Assurance 2
R Defining the type of Applied Part - Metal Probe (Applied Part) employs Water Cooling IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Defining Security Interfaces for Scope for ISMS - Need help IEC 27001 - Information Security Management Systems (ISMS) 10
A Defining the differences between Prototype vs. Production Guidelines? Contract Review Process 5
R Help defining Eyewear Customer Complaint Categories Customer Complaints 5
G Quality Objectives - Where to start defining Quality Objectives? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11