Does the term calibration frequency mean the period of time something is in calibration or is it the ongoing "required" periods by which you must calibrate? If our equipment has a one year calibration frequency and it is stored away and during the storage period its calibration expires. At the moment it expires (still in storage) is this an audit finding because we have failed to follow our own determined calibration frequency?
Additional information; There is no record showing we stored it away. It was not found in use or on a workbench by the auditor. We had removed it from storage and sent it in for calibration so we could use it again. A gap was created between the old calibration period and the new one. The auditor looks at the two calibration certificates and notes the gap between the two and assumes we must have been using the equipment out of calibration but there is no evidence of usage. Since there is no evidence of usage he then determines that we still failed to calibrate at our on determined annual interval and quotes ISO 7.6 as evidence to that fact.
In ISO 7.6 Control of monitoring and measuring devices section a) states; be calibrated or verified at specified intervals, or prior to use, against measurement standards traceable to international or national measurement standards; where no such standards exist, the basis used for calibration or verification shall be recorded;
My contention is that the calibration interval is the period of time "we" determine that the equipment calibration is likely to remain stable. (one year in our case) If the calibration period ends and the equipment becomes out of cal, but in storage we would simply be required to calibrate it again prior to use. The auditors contention was that when it expired in storage, regardless of usage we were out of compliance since we could not show evidence that it had been moved to storage and we had failed to follow our own determined interval of calibration of one year.
Additional information; There is no record showing we stored it away. It was not found in use or on a workbench by the auditor. We had removed it from storage and sent it in for calibration so we could use it again. A gap was created between the old calibration period and the new one. The auditor looks at the two calibration certificates and notes the gap between the two and assumes we must have been using the equipment out of calibration but there is no evidence of usage. Since there is no evidence of usage he then determines that we still failed to calibrate at our on determined annual interval and quotes ISO 7.6 as evidence to that fact.
In ISO 7.6 Control of monitoring and measuring devices section a) states; be calibrated or verified at specified intervals, or prior to use, against measurement standards traceable to international or national measurement standards; where no such standards exist, the basis used for calibration or verification shall be recorded;
My contention is that the calibration interval is the period of time "we" determine that the equipment calibration is likely to remain stable. (one year in our case) If the calibration period ends and the equipment becomes out of cal, but in storage we would simply be required to calibrate it again prior to use. The auditors contention was that when it expired in storage, regardless of usage we were out of compliance since we could not show evidence that it had been moved to storage and we had failed to follow our own determined interval of calibration of one year.