Definition of Manufacturer, Sponsor, Distributor in Australia Therapeutic Goods Act
W
wangyang
Dear all:
I have noticed in the Therapeutic Goods Act , the definition of the "manufacturer" is as follows:
(1) Manufacturers of medical devices
1. The manufacturer of a medical device is the person who is responsible for the design, production, packaging and labelling of the device before it is supplied under the person’s name, whether or not it is the person, or another person acting on the person’s behalf, who carries out those operations.
2. If subsection (1) does not apply to a medical device, the manufacturer of the device is the person who, with a view to supplying the device under the person’s name, does one or more of the following using ready made products:
a. assembles the device;
b. packages the device;
c. processes the device;
d. fully refurbishes the device;
e. labels the device;
f. assigns to the device its purpose by means of information supplied, by the person, on or in any one or more of the following:
i. the labelling on the device;
ii. the instructions for using the device;
iii. any advertising material relating to the device;
iv. technical documentation describing the mechanism of the device.
3. However, a person is not the manufacturer of a medical device if:
a. the person assembles or adapts the device for an individual patient; and
b. the device has already been supplied by another person; and
c. the assembly or adaptation does not change the purpose intended for the device by means of information supplied by that other person, on or in any one or more of the following:
iii. any advertising material relating to the device.
iv. technical documentation describing the mechanism of action of the device
A person is not the manufacturer of a medical device if the person is included in a class of persons prescribed by the regulations for the purposes of this subsection.
From the above definition, i am not sure in the OEM mode, does the contracted manufactuer will be considered to be the manufactuer, or the OEM will be the manufactuer?
In another mode, if the manufactuer sell the medical device to an U.S distributor, then sold to the Australia market, what kind of responsibility should the U.S distributor serve as?
(2): Difference between the sponsor and the distributor.
In TGA 's regulation, i have found the definition and the responsibility of the " sponsor", however, i am not sure if there is any difference between the sponsor and the distributor. If there is more than one distributor, it seems that they all have to include the medical device into ATRG. The evidance can be found from ARGMD:
"If someone intends to supply a device that is identical to a device that is already in the ARTG, even if both devices are made by the same manufacturer, an application to include the device in the ARTG must still be made to the TGA. This is because the ARTG is not only a record of the devices that can be supplied in Australia; it is also a record of all the sponsors who are legally responsible for the medical devices on the market. "
Does it mean that every distributor should include the same medical device into the ATRG and serve as a SPONSOR in fact?
Any response will be appreciated.
MIREGMGR
Re: Question about Manufacturer, sponsor and distributor in Australia
In the general case, and using EU definitions of terms, the OEM is the regulatorily responsible Manufacturer, even if another company contract-makes the device.
None, as long as the manufacturer is the regulatorily responsible Manufacturer and (assuming that the Manufacturer is located outside Australia) there is an Australian Sponsor.
You may have one distributor serve as Sponsor. This means that they're identified on all product labels, including those going to other distributors.
Or, you may have multiple distributors, all serving as Sponsors for the product versions they sell. This however multiplies your regulatory costs and complications.
Or, you may have an office or own operation in Australia, in which case the rules are different and you can represent your own products as an Australian company.
Or, you may hire a third party as your Sponsor, in order to keep all of your distributors on an even footing. Various companies offer this as a regulatory service.
W
wangyang
Thank you for your quick response, MIREGMGR.
Does it mean that only the regulatorily responsible Manufacturer need to apply for TGA certification, and the contracted manufactuer doesn't have to.
Yes i know, thank you.
Thank you.If i have many distributors and i choose to hire the third party as my sponsor, does all of the distributors, especially the domestic distributors, have to include my medical device into the ARTG?
W
wangyang
I have another question:
Many manufactuers may hold CE mark for a single medical device, why it is that? I mean, After my company apply for CE mark, my Client also apply for CE mark. I am totally confused.
pkost
I'm not entirely sure I understand your question here wangyang, however I suspect your client also CE marks the device so that they can put their name on it and not yours!
MIREGMGR
Note that if your customer is CE Marking a device that you make, and they do not have a relationship with you that gives them (1) control of the unchanged characteristics and performance of that device, and (2) access to, or the ability to regulatorily reference, your tech file on that device, their CE Marking basis may be questionable.
In general, that's not your problem, except that it's never good to have customers with questionable regulatory stances.
MIREGMGR
Yes.
No. Only your Sponsor lists the products.
W
wangyang
Yes.
I agree with you, however, when i saw the above sentences, i am not sure if " someone" mean the "sponsor" or the "distributor"?
W
wangyang
Is such way acceptable according to EU directive/regulation?
MIREGMGR
Only Sponsors list products.
If you have one distributor as your Sponsor, or you contract with a third party to be your Sponsor, or you open an office in Australia, then there would be only one listing.
If you followed the approach of having each distributor be a Sponsor, then they all would list.
If you contract with a third party in Australia to be your Sponsor, and separately sell product to someone else outside of Australia and unbeknownst to you they then sell that product to an independent distributor within Australia, that independent distributor would have to Sponsor the product sold to it, even though identical product was already on the Australian market.
J Definition of repackager v. manufacturer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
M Definition of Medical Device Manufacturer under CMDR Canada Medical Device Regulations 4
C Definition Contract Manufacturer - Definition of Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
K Legal Manufacturer - MDD Definition of "Legal Manufacturer" EU Medical Device Regulations 12
bio_subbu GHTF definition of Manufacturer, Authorized Representative, Distributor and Importer Other Medical Device and Orthopedic Related Topics 2
T Definition Servicing Activities - Definition - Medical device accessory manufacturer Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 1
S Definition OEM (Original Equipment Manufacturer) - Exact definition of OEM Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 9
V Definition OEM - Definition of OEM - Original Equipment Manufacturer Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 8
M Definition Aviation - Definition of Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
J Where is the definition of a critical supplier? ISO 13485:2016 - Medical Device Quality Management Systems 5
A DMDIV-Variants : definition and technical documentation CE Marking (Conformité Européene) / CB Scheme 4
O Clarifying FDA definition of "finished device" and “capable of functioning” 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
L 'Distributor' definition & MDR/MSR relationship EU Medical Device Regulations 3
D Definition of equipment for ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 1
M ISO 13485:2016 Complaint Definition Clarity Customer Complaints 2
DuncanGibbons Technical Data Package vs Digital Product Definition APQP and PPAP 0
S Definition of "worldwide" in view of IATF 16949 and Product conformity IATF 16949 - Automotive Quality Systems Standard 3
S Importer definition EU Medical Device Regulations 0
A Software as Medical Device (SaMD) definition and its applicability Other Medical Device and Orthopedic Related Topics 4
M Definition of Done - New Product Phase to the sustaining phase Manufacturing and Related Processes 3
M Help me improve the definition for internal PPM Manufacturing and Related Processes 3
D Tolerance definition based on expected Cp/cpk Reliability Analysis - Predictions, Testing and Standards 14
D BS EN 62304 - Medical-Relevant Data C.5 - Definition of IEC 62304 - Medical Device Software Life Cycle Processes 5
P Definition of production tooling in IATF16949 clause 8.5.1.6 IATF 16949 - Automotive Quality Systems Standard 4
J Sample size definition in an Automotive SMT pilot lot run Misc. Quality Assurance and Business Systems Related Topics 1
J Definition Defect Mapping - Here is my definition Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 9
D Definition of "Critical Laboratory Equipment"? General Measurement Device and Calibration Topics 1
Z Definitive definition of design? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
R Definition Key Supplier - Definition of Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 7
A Definition of customer in ITP (Inspection & Test Plan) Contract Review Process 3
F MEDDEV 2.7/1 rev 4 and MDR - Definition of Description of Principles of operation EU Medical Device Regulations 4
S Definition of "servicing activities" - Spinal Implants ISO 13485:2016 - Medical Device Quality Management Systems 4
T Custom made medical device MDD vs MDR definition CE Marking (Conformité Européene) / CB Scheme 12
C Function definition for DFMEA FMEA and Control Plans 1
W Key functional characteristic definition - Tolerances and Deviations Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 0
Marc Definition Waiver - Definition Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 4
M Informational MHRA guidance – Assistive technology: definition and safe use Medical Device and FDA Regulations and Standards News 0
R Critical suppliers (Definition of) and MDSAP (Medical Device Single Audit Program) ISO 13485:2016 - Medical Device Quality Management Systems 15
Pancho Definition Quality Assurance Program - Definition Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 9
M Inputs on definition of very similar processes for multi site audit sample - IAF MD1 2018 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
U Hand-Held dosing device has no PATIENT - Interpretation of the PATIENT definition IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
L Medical Device Accessory - Health Canada guidance / definition on what an accessory is Canada Medical Device Regulations 3
C MDR Importer/Distributor Definition Questions EU Medical Device Regulations 14
G Medical Device "Immediate Container" Interpretation of Definition US Food and Drug Administration (FDA) 2
S Definition of Convenience Kit for a Combination Product US Food and Drug Administration (FDA) 0
M Are Risks and Opportunities Required as Part of the Process Definition ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
A Definition of "Sensitive Products" Clause 8.5.4 (c) in AS9100 Rev. D AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
Marc Definition Jidoka - Definition Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 2
Marc Definition DFMEA (Design FMEA) - Definition Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 0
S MDR Annex II Design & Manufacturing info - 'Adjuvants' definition EU Medical Device Regulations 3