Dental Treatment Unit (aka Dentist's Chair) requirements
N
nieuwlaat
Since this is my first post here I will introduce myself briefly.
I am working for a company that has been marketing dental equipment in the Netherlands for almost 20 years. I have been working for this company for almost a year now. I am a lawyer by training, but I have no specific knowledge of application of standards, methods etc. relating to medical devices.
At the moment we are working on the development of a new product line. At first we will be exclusively marketing the new product to the European market. (mainly BeNeLux, Nordic countries and Germany).
The product is a Dental Treatment Unit, sometimes referred to a dentist's chair (although the actual chair in which the patient sits is not included in our product!)
My first priority is to work out what are the technical and procedural requirements that apply to our product category, to which we have to comply to be able to put the product on the market in the first place. Also, a reasonable level of protection from claims is desirable.
The main problem that I have is that it is not clear to me where I can find an overview of the requirements we have to fulfil. The information I have found so far has been detailed enough, but I have no idea what is actually important to me and what isn't.
I have worked out that the product can probably be classified as a Medical Device, Class I according to the European Medical Devices Directive (93/42/EEG:2007)
It seems that the procedure to be able to put the CE mark on it is pretty straightforward and does not require auditing, etc. But perhaps there is more to it.
What steps should I take next? Am I missing important stuff that might interfere with market introduction?
If you need more info, examples or anything, please let me know.
Thanks In Advance! :agree1:
Wes Bucey
Re: Dentist's Chair - help needed to get oversight on requirements
Is your company registered to ISO 13485? If so, the registrar who has issued your certificate is one resource you can mine. At the very least, if it doesn't supply the oversight itself, it can steer you to a competent, authoritative source.
N
nieuwlaat
Thanks for the quick response!
As far as I know the company is not registered to ISO 13485, likely never has been. Is this a must-have (as in: legal obligation) or simply a more or less convenient way to tackle the issue?
Best regards
Wes Bucey
Best regards
The first question becomes, "Is this truly a medical device according to the legal definition of the governmental jurisdiction where it will be sold?" If yes, many European countries require registration and thus oversight of the product. Many of those same jurisdictions subscribe to a "harmonization" pact whereby registration to ISO 13485 by an approved registrar can save the trouble and expense of individual registration in each country. It's a legal question, not really a quality one.
Ronen E
Problem Solver
Moderator
The Medical Devices Directive is your best friend. Basically, everything you need to know is there. When in doubt, consult the rich guidance available on line; best start with the MEDDEV official guidance (then you could move on - if need be - to NB-MED, GHTF etc.).
If indeed the device at question is a class I device (with no measuring function), then the process is pretty much straightforward as you've depicted and very little guidance would be required. Read the entire MDD text, then focus on annex I and annex VII.
All the best,
Ronen.
pkost
You have said that you intend to market in the EU, and are therefore required to follow the Medical device directive (93/42/eec). This document tells you all you need to know and what you need to do to market in the EU.
A quick run through for you:
Under this directive a medical device is defined as
‘medical device’ means any instrument, apparatus, appliance, software, material or other article, whether used alone or in combination, including the software intended by its manufacturer to be used specifically for diagnostic and/or therapeutic purposes and necessary for its proper application, intended by the manufacturer to be used for human beings for the purpose of:
— diagnosis, prevention, monitoring, treatment or alleviation of
disease,
— diagnosis, monitoring, treatment, alleviation of or compensation
for an injury or handicap,
— investigation, replacement or modification of the anatomy or of
a physiological process,
— control of conception,
and which does not achieve its principal intended action in or on the human body by pharmacological, immunological or metabolic means, but which may be assisted in its function by such means; (b) ‘accessory’ means an article which whilst not being a device is intended specifically by its manufacturer to be used together with a device to enable it to be used in accordance with the use of the device intended by the manufacturer of the device;
Your dentist chair falls under this definition (an accessory is to be treated as a device).
The next step is to identify the classification, you can do this in Annex IX of the directive. I would check for yourself, but I would imagine a dentist chair falls under rule 1 (non invasive) or 12 (active device) which makes it a class 1 device.
Assuming that you agree it is class 1, Article 11 tells you how you should go about conforming with the requirements.... specifically:
article 11 para 5 said:
In the case of devices falling within Class I, other than devices which are custom-made or intended for clinical investigations, the manufacturer shall, in order to affix the CE marking, follow the procedure referred to in Annex VII and draw up the EC declaration of conformity required before placing the device on the market.
Article VII is probably too long to copy out here, but to summarise you have to
1.put together a technical file (detailed in para 3); this includes demonstrating that you comply with all the essential requirements in annex I.
2. Put in place procedures to handle vigilance, including the requirements of Annex X - clinical evaluation and complaint reporting
3. Write a declaration of conformity.
There is no specific requirement to have a quality management system in place, although if you do, ISO 13485 is the way to go. I would certainly refer to this document when it comes to writing your vigilance procedures and for document retention.
I would also recommend that you look through ISO 14971 which relates to risk management for medical devices; again there is no requirement for this.
When demonstrating compliance with the essential requirements, it is best to use harmonised standards (you can find a list here.. http://ec.europa.eu/enterprise/poli...-legislation/list-references/medical-devices/) as these lead to assumed compliance (both 13485 and 14971 are harmonised), failing that, national standards or guidance is good. Failing that clinical evaluation (annex X) and risk assessment.
Another small point which may be missed...
Under article 14, when you have all of this in place, but before you market, it is necessary to register with your competent authority to inform them that you are placing a class I medical device on the market - you only need to do this for your CA, not all the countries you are selling into.
Final note - these regulations apply to EU countries only, other regulations apply territories outside the EU
[edit] realised dentist chairs tend to be powered to provide light and or move patients up/down/left/right/tilt etc. therefore are active devices and not covered under rule 1 - rule 12 applies in this case.
N
nieuwlaat
Hi there!
just wanted to thank all contributors to this thread, your comments have been helpful. :tg:
best regards,
w. nieuwlaat
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
Q Old products new class - Dental Devices - Choosing tests EU Medical Device Regulations 2
P Manufacturing equipment (e.g.: Dental Welder): medical device or not? US Food and Drug Administration (FDA) 5
A Help establishing regulatory requirements - Plastic dental cartridge syringe US Food and Drug Administration (FDA) 0
P Change on Medical Device Classifications - Dental Crowns Other Medical Device Regulations World-Wide 2
M Informational US FDA paper – Epidemiological Evidence on the Adverse Health Effects Reports in Relation to Mercury from Dental Amalgam: Systematic Literature Review Medical Device and FDA Regulations and Standards News 0
M Informational MHRA Guidance – Dental radiographic X-ray imaging: dose to patients Medical Device and FDA Regulations and Standards News 0
K Is Compliance Testing of IEC 60601-2-63 mandatory for Europe - Dental X-Ray machine CE Marking (Conformité Européene) / CB Scheme 3
B Can Class IIa Dental Medical Devices be sold into the EU without CE EU Medical Device Regulations 2
F Bioburden Testing requirements for Non-Sterile Dental Products EU Medical Device Regulations 6
P Single Use Disposable Dental Syringe - CE Marking Advice EU Medical Device Regulations 4
R Air-powered dental handpiece with LED IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Is IEC 80601-2-60 necessary for Dental Laser? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Standards for Emergency Button for a Dental Laser IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P DMR (Device Master Record) for Class I and Class II Dental Product 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
I Is ARIMA (Time Series Analysis) Model appropriate for this Dental Research? Using Minitab Software 4
A Dental Laser Certification IEC 62304 for a Firmware IEC 62304 - Medical Device Software Life Cycle Processes 1
A Type of Materials to comply with IEC 60601 (Dental Laser Case) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Unique Device Identifier for Small Dental Screws ISO 13485:2016 - Medical Device Quality Management Systems 3
2 Is it common to employ part time dental assistants ? Coffee Break and Water Cooler Discussions 3
G Determination of dental plaque disclosing products Other ISO and International Standards and European Regulations 1
J Product Liability for Dental Gel - Class I Medical Device EU Medical Device Regulations 1
J Question about CE mark for Dental Gel which is probably Medical Device Class 2a EU Medical Device Regulations 2
R Does dental unit need to meet the requirements of clause 14 (PEMS)? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
N Dental Implants and their Accessories - CE Mark Conformity Assesment EU Medical Device Regulations 3
W Is a Dental Laboratory Device a Medical Device? EU Medical Device Regulations 9
E AQL Level Change for FDA Validated Unclassified Dental Product Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
R ASME of Sterilizers from China for the US Dental Industry China Medical Device Regulations 3
J Nonconforming Implantable Dental Devices ISO 13485:2016 - Medical Device Quality Management Systems 4
E FDA Quality Systems and DAMAS (Dental Appliance Manufacturers Audit Scheme) US Food and Drug Administration (FDA) 3
E Registration & Listing of a Dental Device - Class Enforcement Descretion 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
W What is the applicable standard for Chair Mounted Dental Unit for CE IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L Dental Handpieces - Equipment Classification and CE Marking EU Medical Device Regulations 24
P Does the the EU ban BPA and/or Bis-GMA in Medical (Dental) Devices EU Medical Device Regulations 2
M Is Dental Impression Paste (Zinc Oxide and Eugenol) a Drug 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
M Does a dental handpiece need FDA permission? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
R Finding an ADA (American Dental Association) Standard Testing Laboratory Various Other Specifications, Standards, and related Requirements 4
W Classification of Dental Implants - MDD 2.4. Rule 8 EU Medical Device Regulations 8
M CE Mark for Dental Imaging Software EU Medical Device Regulations 5
Ajit Basrur FDA Issues Final Regulation on Dental Amalgam 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
L Dental handpiece with product code EFB for 510k submission Other US Medical Device Regulations 5
M Topical Anesthetic for Dental use - Where to start to get it registered? Other US Medical Device Regulations 10
A Research in Mexico, anyone? US class I (non exempt) 510(K) Dental ISO 13485:2016 - Medical Device Quality Management Systems 1
K Need DETAILED Procedure Template for Dental Insurance Co. Document Control Systems, Procedures, Forms and Templates 21
B FDA vs. Materials requiring Refrigeration - Dental adhesives 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 18
R MSA studies for Heat Treatment Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
K Please help identify appropriate statistical treatment Statistical Analysis Tools, Techniques and SPC 13
1 ISO Registrar with waste water treatment experience Registrars and Notified Bodies 1
J GLP Animal Study - Change number of treatment sites Other US Medical Device Regulations 2
Q Risk Analysis - Same Risk Treatment for Context and Interested Parties ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7