Depth Mic Calibration

Jrupnow

Registered
How are you calibrating depth mics or any equipment that has interchangeable rods? For the last 12 years our have been calibrated and stickered as "calibrate at use" because of the rods being interchanged. We are getting a minor non-conformance from a customer because we do not have records of "calibrate at use". Is everyone else just calibrating the tools and assuming when the rods are interchanged they are correct? What are other people doing for this situation?
 

Ron Rompen

Trusted Information Resource
calibrate the micrometer head to validate the wear (or lack of wear) on the screw thread. Measure/calibrate each rod with a height gauge (start by measuring the initial value when received, and compare to that0.
 

UncleFester

Involved In Discussions
It looks like you're verifying rather than calibrating, as you're not measuring squareness of the datum face to the axis of the rod, not flatness of the datum face as would be expected on a certificate of calibration. What tolerances are you using per rod length at verification? As a simple guide, you could have a series of tolerances per rod length and check this when changing each rod. You could get the table from BS 6468.
 

Michael_M

Trusted Information Resource
The following is our instructions on depth micrometers:

  • Micrometer, Depth
  • ±.0005” from gage actual measurement. These gage typically only read to .XX1. Tolerance on measurement is ½ distance between lines.
  • The range of the device will be measured, as applicable:
  • Using a single rod, measure minimum to maximum 4 points. At least one measurement to be at a non-standard increment (i.e., measuring .0625)
  • Measure 1 point for each additional rod in set.
  • Inspect each rod for bend and/or damage.
 