Design and Development of Software under 13485:2016 or 62304?

JoshuaFroud

Involved In Discussions
#1
Good Morning

I'm currently working through our transition materials/gap analysis for ISO 13485:2016. I'm looking at the design and development requirements and I'm a little lost. Previously I have only worked with physical goods, whereas now I work for a company produces both a physical device and the software to run it. So have little experience in the area of software.

We currently have in place a Software development Procedure based on IEC 62304 and a Design and Development procedure based in ISO 13485:2003/2012.

My question is does our software development procedure/process need to meet the requirements of 13485 in addition to 62304. I.e. at present our software development procedure makes no reference to Design Review.

Thanks in advance!
 
Elsmar Forum Sponsor

Marcelo

Inactive Registered Visitor
#2
My question is does our software development procedure/process need to meet the requirements of 13485 in addition to 62304. I.e. at present our software development procedure makes no reference to Design Review.
Yes. Basically, you have to map the design controls of ISO 13485 to your software development lifecycle model (in the same way you have to map the requirements of IEC 62304 to your software development lifecycle model).
 

yodon

Staff member
Super Moderator
#3
Marcelo is correct, as usual.

The design reviews in 13485 are typically laid out as "stage gate" type reviews; i.e., we're at a particular point in the process and we do a review to make sure we're where we think we need to be and that the design is progressing correctly.

In 62304, the reviews are more "artifact oriented." For example, the step to "verify software architecture" is a (design) review of the architecture.

Because software is more evolutionary than other aspects of the design (i.e., it's easier to change the software than to re-design a board), the "stage gate" approach isn't all that effective with software - especially in a more agile environment. However, as your 13485-based design reviews occur, you should be including a review of the relevant software aspects to ensure things are tracking as planned.
 

JoshuaFroud

Involved In Discussions
#4
Thank you both for your comments. That was how I was reading the requirements, I doubted myself as the previous Quality person had only written the Software procedure to comply with 62304 and they, (in years at least), had significantly more experience than me.



Looks like I have a procedure to re-write this week!
 
#5
Hello,
I am currently working in AI-driven software for Ophthalmology and in process of implementation of ISO 13485:2016. According to the above conversation, I will have to consider ISO 13485 & IEC 62304 standards while preparing the procedure and other design related documents.
So, can I consider the same documents as a part of Medical device file (Clause 4.2.3 of ISO 13485:2016) and also preparing the design documents as per IEC 62304 would fulfill the requirements of Design and development clause of ISO 13485?

Also, for getting ISO 13485 certification do I have to define the class of the medical device? I assume that the classification differs as per country's regulations and thus it is not required now as we are not applying for any market registration. However, I have to consider the safety classification of the software as per IEC 62304.

Please advise.

Thank you
 

yodon

Staff member
Super Moderator
#6
Just to reset the baseline a bit, standards are always voluntary. They are, however, the most direct path to market clearance. Complying with a harmonized standard (like 13485 and 62304) gives you a presumption of conformity in the EU (and similarly 62304 is a recognized consensus standard in the US). So it's probably in your best interest to follow these.

Do note that both 13485 and 62304 reference ISO 14971 for overall risk management and risk is getting a lot of attention.

Oh, and usability is also getting considerable attention. Compliance to IEC 62366 is recommended.

In at least the US and EU, you do have to establish the device class. There are decision trees to help guide. 62304 requires that you define the software safety class. Especially for a "Software as a Medical Device" (SaMD), the device class and software safety class will probably parallel (i.e., if your software device class is III, you'll likely have a software safety class of C). If you have plans to obtain market clearance in the future, you'd be well advised to do all the design and development work at the level indicated by the software safety class. It's very difficult to retrospectively develop the required artifacts.

Following 62304 gets you pretty close to 13485 7.3 but there are some differences. For example, 62304 doesn't really address design verification and validation in the same manner. (And 62366 can certainly fill in some gaps here). Another difference is in planning. 13485 requires you establish the competence of personnel required. So look closely at the requirements and see what maps and where the gaps are.

A few more notes.

Every jurisdiction now is very concerned about cybersecurity. Look to see what country expectations are early. Some of the requirements drive design decisions so you'll want to build this in; it cannot be bolted on after the design is done. I know that UL has some standards (which are recognized consensus standards in the US).

At least the FDA has some working papers on AI / machine learning systems. You'll probably want to get familiar with the framework outlined since it's really an emerging field in medical devices. Much of it is about how you ensure that the learning is improving and not regressing.
 
#7
Just to reset the baseline a bit, standards are always voluntary. They are, however, the most direct path to market clearance. Complying with a harmonized standard (like 13485 and 62304) gives you a presumption of conformity in the EU (and similarly 62304 is a recognized consensus standard in the US). So it's probably in your best interest to follow these.

Do note that both 13485 and 62304 reference ISO 14971 for overall risk management and risk is getting a lot of attention.

Oh, and usability is also getting considerable attention. Compliance to IEC 62366 is recommended.

In at least the US and EU, you do have to establish the device class. There are decision trees to help guide. 62304 requires that you define the software safety class. Especially for a "Software as a Medical Device" (SaMD), the device class and software safety class will probably parallel (i.e., if your software device class is III, you'll likely have a software safety class of C). If you have plans to obtain market clearance in the future, you'd be well advised to do all the design and development work at the level indicated by the software safety class. It's very difficult to retrospectively develop the required artifacts.

Following 62304 gets you pretty close to 13485 7.3 but there are some differences. For example, 62304 doesn't really address design verification and validation in the same manner. (And 62366 can certainly fill in some gaps here). Another difference is in planning. 13485 requires you establish the competence of personnel required. So look closely at the requirements and see what maps and where the gaps are.

A few more notes.

Every jurisdiction now is very concerned about cybersecurity. Look to see what country expectations are early. Some of the requirements drive design decisions so you'll want to build this in; it cannot be bolted on after the design is done. I know that UL has some standards (which are recognized consensus standards in the US).

At least the FDA has some working papers on AI / machine learning systems. You'll probably want to get familiar with the framework outlined since it's really an emerging field in medical devices. Much of it is about how you ensure that the learning is improving and not regressing.
Thank you very much for this detailed information.
It is very informative and surely I will go through all the said requirements for QMS implementation.
Thanks again.
 
Thread starter Similar threads Forum Replies Date
L Software Medical Device - 7.3.8 - Design and Development Transfer ISO 13485:2016 - Medical Device Quality Management Systems 4
M Design and Development Outputs for Software Development (7.3.3) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
A Design Review OR Design Change - Software development company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
P Software to help meet ISO & AS9100 Design & Development Rqmts Design and Development of Products and Processes 2
J Help! Need sample procedures for Product & Software Design and Development Design and Development of Products and Processes 4
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
A 8.6 Release of products and services, 8.3 Design and development - evidence required ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
A Design and development procedure for API Spec Q2 Oil and Gas Industry Standards and Regulations 6
D Using Laboratory Notebooks in R&D and Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 - 7.3.6 Design and development verification - Do most folks create a separate SOP? ISO 13485:2016 - Medical Device Quality Management Systems 4
L Design & Development of a SERVICE Service Industry Specific Topics 13
T Design Control Procedures later in the Development Process ISO 13485:2016 - Medical Device Quality Management Systems 6
K Old medical devices -> 7.3.7. Design and development validation ISO 13485:2016 - Medical Device Quality Management Systems 1
M Design Development MDR Design and Development of Products and Processes 0
O How can I justify excluding the R&D group and the design and development clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
C IATF 16949 8.3 Exclusion - Manufacturing process design and development IATF 16949 - Automotive Quality Systems Standard 5
A Design and development of products and services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S ISO 9001 Clause 8.3 - Design & Development for training course center ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
S ISO 9001:2015 & ISO 14001:2015 - I need a format for Design & Development planning ISO 14001:2015 Specific Discussions 2
AlienraverX Design and Development Audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
V Who should define and own the Design and Development Plan and how to maintain the updates and revisions. ISO 13485:2016 - Medical Device Quality Management Systems 2
K Design and Development Exclusion Quality Management System (QMS) Manuals 1
V Exclusion of 'Design and Development' from scope of certification ISO 13485:2016 - Medical Device Quality Management Systems 9
XRAY_3121 Design and Development Requirement - MDSAP Audit Finding Other Medical Device Regulations World-Wide 5
P Performance evaluation (IVDD Annex VIII) & Design and Development Validation Studies EU Medical Device Regulations 1
S Waterfall model or v-model design and development ISO 13485:2016 - Medical Device Quality Management Systems 2
M IATF Clause 8.3 - Design and Development IATF 16949 - Automotive Quality Systems Standard 5
K Design and Development Exemption/NA confusion Design and Development of Products and Processes 6
C AS9100 8.3.5.e Design and Development Outputs - Key Characteristics / Critical Items AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S Purchasing for Design and Development Organization ISO 13485:2016 - Medical Device Quality Management Systems 3
qualprod Development separated from design? ISO 9001 cl. 8.3 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
G Design and development of user centric audit report visualisation tool Design and Development of Products and Processes 5
D IATF 16949 Design and Development Planning IATF 16949 - Automotive Quality Systems Standard 1
M IATF 16949 Cl. 8.3 - Manufacturing process design and development IATF 16949 - Automotive Quality Systems Standard 1
P Design and Development Clause ISO 9001:2015 Exclusion for Medical Services Design and Development of Products and Processes 3
A ISO 13485:2016 Certification for Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 0
H IATF 16949 Cl. 8.3 - Design and Development Exclusion Question IATF 16949 - Automotive Quality Systems Standard 4
R Development of Design FMEA - Functional Requirements FMEA and Control Plans 1
G ISO 9001:2015 8.3 Design and Development (in Civil Engineering) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
M AS9100D excluding Design and Development - Small Job Shop AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 18
M API Spec Q2 Design and Development Requirements for Service Providers Oil and Gas Industry Standards and Regulations 11
T Design Changes - Updating the Design and Development File EU Medical Device Regulations 5
Y ISO 9001 Cl. 8.3 - Design and Development of Products and Services not applicable ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J ISO 9001:2015 8.3.2 f) - Interfaces between Persons in Design & Development ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
T ISO 9001:2015 - 8.3 Design and Development of Products and Services (Repair Company) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
I Design & Development, Research and Innovation in the Medical Device framework ISO 13485:2016 - Medical Device Quality Management Systems 3
M Question about ISO 9001:2015 8.5.3 - Design and Development of Products and Services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
D Design and Development Procedure example and Template wanted - AS9100 Rev D AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
D Recommended Training for Design Engineers in IATF 16949 Cl.8.3 - Design & Development Design and Development of Products and Processes 0
B IATF 16949 Section Clause 8.3.4.1 - Monitoring - Design and Development Input(s) IATF 16949 - Automotive Quality Systems Standard 3

Similar threads

Top Bottom