Design and Development Requirement - MDSAP Audit Finding

XRAY_3121

Wholesale Distributor Compliance and Regulatory
#1
Hi all,
I work for a company that has no design and development files (they've been making same product with no changes for 50+ years - Lead aprons, dental xray film, etc.). We just had MDSAP audit under Annex II of MDD, so these files were expected. In my initial response I said we would remove the requirement from the scope of the audit - but was told:
"The company decision to exclude the device is not sustained based on the regulatory requirements about design process => do all countries regulations allow to exclude design for all potential device classifications. What impact will this decision have on the actual or expected certificates ?"

My updated response proposal:
Design and development will be excluded from the scope of the quality management system and we request to receive CE certification under MDD Annex V (not II) which should have been initially requested. The design and development is excluded because there are no design activities performed on the devices. The devices are grandfathered/ legacy devices and there are no design changes.
Brazil and Australia will be removed from scope and we will not sell to these countries.

All of our devices in US are class I and only the intraoral film is a Class IIa in EU (and Australia I believe).

My question is does this seem acceptable, does only Brazil and Australia REQUIRE this??
Am I missing any type of exception or technicality based on the type of product?

Thanks I really appreciate it!

Matthew
 
Elsmar Forum Sponsor

somashekar

Staff member
Super Moderator
#2
Impact on certificate is No Certificate.
Grandfathered or Legacy devices, no issues., but you are responsible for the design and must have your design development records. Make one.
Your updated response will not be sustained.
 

Ronen E

Problem Solver
Staff member
Moderator
#3
Technical files are required for all devices (including class I) under the MDD, regardless of conformity assessment route (Annex II, V etc.). See Annex VII.
Some class I device types require design control in the US (refer 21 CFR 820.30). All devices require a DMR.
 

XRAY_3121

Wholesale Distributor Compliance and Regulatory
#4
Yes good points, I see Annex VII and what is required for DoC, I think we have them all. I'm not certain it's so cut and dry though.

We do have a DMR basically, it's just not called a "DMR". Easy fix on name. We do have technical files as we know this is required (Product descriptions, specs, IFUs, materials, etc., all of the required items.

I guess what we are missing is: Design Input / Output, Design Control, Design Verification, Design Validation type data specifically.
And design control as we don't change them. Product History is missing too as the data for that is very old.

I guess it's doable to create this and use it going forward as the corrective action.
As far as 21 CFR 820.30 I did see this and we are OK without design.

And as far as Annex V...
"Annex V has similar quality management system requirements to Annex II, but the main difference is that Annex II includes design requirements and Annex V does not. For Annex V conformity assessments, the manufacturer needs to have a quality management system for the manufacturing and final inspection for the devices, but there is no requirement that a design system is included. As a result, there may be no design history file to include in the manufacturer’s technical documentation.
However, even if there is no DHF, this conformity assessment includes a technical documentation review."

That is from GMED. Which would imply we would be good for EU (as technical documentation is our TF).

Is it not conceivable to ask for our CE marking under Annex V then?

Annex VII (which is in regards to declaration of conformity) says:
Class I
This route is self-declaration or self-certification and is described in Annex VII Module A, EC Declaration of Conformity. The manufacturer ensures and formally declares, via a written statement, that the products meet the applicable provisions of the Directive.
Class IIa
The manufacturer declares conformity with the provisions of the Directive and Regulations (Annex VII) and ensures that the products comply with relevant essential requirements. However, for Class IIa products, this declaration must be backed up in all cases with conformity assessment by a NB using Annex II, IV, V or VI. WE HAVE THIS

These are the only 2 classes we manufacture.

Thanks for input somashekar and Ronen E. I appreciate it!
 

Ronen E

Problem Solver
Staff member
Moderator
#5
Is it not conceivable to ask for our CE marking under Annex V then?
You can definitely try your luck with your NB and see what comes back.
I wonder though how you'd be able to demonstrate full compliance with the ERs without having documented verification and validation on file.
Do you have RMFs?
 
Thread starter Similar threads Forum Replies Date
smryan Satisfying the ISO 9001:2008 7.3.5 Design & Development Verification requirement ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
C D&D (Design and Development) Requirement for Semi-Finished Products Design and Development of Products and Processes 4
R Requirement of Design and Development Schedule ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
A 8.6 Release of products and services, 8.3 Design and development - evidence required ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
A Design and development procedure for API Spec Q2 Oil and Gas Industry Standards and Regulations 6
D Using Laboratory Notebooks in R&D and Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 3
D ISO 13485 - 7.3.6 Design and development verification - Do most folks create a separate SOP? ISO 13485:2016 - Medical Device Quality Management Systems 4
L Design & Development of a SERVICE Service Industry Specific Topics 13
T Design Control Procedures later in the Development Process ISO 13485:2016 - Medical Device Quality Management Systems 6
K Old medical devices -> 7.3.7. Design and development validation ISO 13485:2016 - Medical Device Quality Management Systems 1
M Design Development MDR Design and Development of Products and Processes 0
O How can I justify excluding the R&D group and the design and development clause? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
C IATF 16949 8.3 Exclusion - Manufacturing process design and development IATF 16949 - Automotive Quality Systems Standard 5
A Design and development of products and services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S ISO 9001 Clause 8.3 - Design & Development for training course center ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
S ISO 9001:2015 & ISO 14001:2015 - I need a format for Design & Development planning ISO 14001:2015 Specific Discussions 2
AlienraverX Design and Development Audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
V Who should define and own the Design and Development Plan and how to maintain the updates and revisions. ISO 13485:2016 - Medical Device Quality Management Systems 2
K Design and Development Exclusion Quality Management System (QMS) Manuals 1
V Exclusion of 'Design and Development' from scope of certification ISO 13485:2016 - Medical Device Quality Management Systems 9
P Performance evaluation (IVDD Annex VIII) & Design and Development Validation Studies EU Medical Device Regulations 1
S Waterfall model or v-model design and development ISO 13485:2016 - Medical Device Quality Management Systems 2
JoshuaFroud Design and Development of Software under 13485:2016 or 62304? ISO 13485:2016 - Medical Device Quality Management Systems 6
M IATF Clause 8.3 - Design and Development IATF 16949 - Automotive Quality Systems Standard 5
K Design and Development Exemption/NA confusion Design and Development of Products and Processes 6
C AS9100 8.3.5.e Design and Development Outputs - Key Characteristics / Critical Items AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
L Software Medical Device - 7.3.8 - Design and Development Transfer ISO 13485:2016 - Medical Device Quality Management Systems 4
S Purchasing for Design and Development Organization ISO 13485:2016 - Medical Device Quality Management Systems 3
qualprod Development separated from design? ISO 9001 cl. 8.3 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
G Design and development of user centric audit report visualisation tool Design and Development of Products and Processes 5
D IATF 16949 Design and Development Planning IATF 16949 - Automotive Quality Systems Standard 1
M IATF 16949 Cl. 8.3 - Manufacturing process design and development IATF 16949 - Automotive Quality Systems Standard 1
P Design and Development Clause ISO 9001:2015 Exclusion for Medical Services Design and Development of Products and Processes 3
A ISO 13485:2016 Certification for Design and Development ISO 13485:2016 - Medical Device Quality Management Systems 0
H IATF 16949 Cl. 8.3 - Design and Development Exclusion Question IATF 16949 - Automotive Quality Systems Standard 4
R Development of Design FMEA - Functional Requirements FMEA and Control Plans 1
G ISO 9001:2015 8.3 Design and Development (in Civil Engineering) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
M AS9100D excluding Design and Development - Small Job Shop AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 18
M API Spec Q2 Design and Development Requirements for Service Providers Oil and Gas Industry Standards and Regulations 11
T Design Changes - Updating the Design and Development File EU Medical Device Regulations 5
Y ISO 9001 Cl. 8.3 - Design and Development of Products and Services not applicable ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J ISO 9001:2015 8.3.2 f) - Interfaces between Persons in Design & Development ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
T ISO 9001:2015 - 8.3 Design and Development of Products and Services (Repair Company) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
I Design & Development, Research and Innovation in the Medical Device framework ISO 13485:2016 - Medical Device Quality Management Systems 3
M Question about ISO 9001:2015 8.5.3 - Design and Development of Products and Services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
D Design and Development Procedure example and Template wanted - AS9100 Rev D AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
D Recommended Training for Design Engineers in IATF 16949 Cl.8.3 - Design & Development Design and Development of Products and Processes 0
B IATF 16949 Section Clause 8.3.4.1 - Monitoring - Design and Development Input(s) IATF 16949 - Automotive Quality Systems Standard 3
T FDA Design and Development Planning Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3

Similar threads

Top Bottom