Design FMEA, Process FMEA, Component FMEA --> FDA requirements?

A

andymo13

#1
Hello everyone,

we conducted a design FMEA and a component FMEA but no seperate process FMEA. The component (or equippment FMEA) includes the process risks. Now I wonder ist the FDA requires to conduct both FMEAs separately? If so, where in the CFR ?
Can anything be said against doing both (equipment and process risk analysis) in one FMEA? :confused:

thanks for your answers in advance
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
The FDA wants you to have an effective and comprehensive risk analysis and management process. They don't yet have detailed rules as to exactly how you must proceed. If your process is rigorous and you can coherently explain its effectiveness to a reviewer or inspector, it should be accepted.
 

v9991

Trusted Information Resource
#4
Hello everyone,

we conducted a design FMEA and a component FMEA but no seperate process FMEA. The component (or equippment FMEA) includes the process risks. Now I wonder ist the FDA requires to conduct both FMEAs separately? If so, where in the CFR ?
Can anything be said against doing both (equipment and process risk analysis) in one FMEA? :confused:

thanks for your answers in advance
I agree with MIREGMGR; Effectiveness matters the most;

but but would like to emphasize the SOP (Systems-adequacy & completeness)
viz.,Since you are covering the 'process' as part of equipment/component FMEA which most probably occurs at "later-part/before" of operations.; arn't u missing the opportunity in early design/development phases to identification-evaluation of process risks to the product!
 
A

andymo13

#5
Hi v9991,

thanks for your respond.

During developing we do conduct a seperate process and design FMEA but after design transfer into production there is only one FMEA which evaluate both.
I´ve looked into SOP which states that you ´ve to check each process step and each system part for potential risks.
In the FMEAs the boundaries between the process and the design are not clearly defined they rather merge.
At times it can be confusing/ easily to miss something
 
G

gholland

#6
As part of your risk analysis you should be noting what risk control measures your process or design has. For example if you have a sensor to detect an out-of-tolerance condition then you should track this as a risk control measure. All of your risk control measures should be evaluated for effectiveness.

If you have done a proper job with your effectiveness checks you should easily be able to demonstrate you are properly controlling the process risk. Your Design Verifcations should take the same approach for design risk control measures, i.e. an alarm properly sounds when a failure is detected.


:2cents:
 

Ronen E

Problem Solver
Staff member
Moderator
#8
As part of your risk analysis you should be noting what risk control measures your process or design has... All of your risk control measures should be evaluated for effectiveness.
Could you please kindly state the specific regulation(s) that drive these "should"s?

Many thanks.
 
M

MIREGMGR

#9
Can you please kindly point out the regulation that this is based upon?
You might have noted my comment above:

They don't yet have detailed rules as to exactly how you must proceed.
However, it's easy to discern their risk expectation by paying attention to the stream of topical information from the agency and its managers, and by discussing the topic with submission examiners.

As one data point among many, see the strategic white paper at http://www.fda.gov/AboutFDA/ReportsManualsForms/Reports/ucm183673.htm, which discusses how FDA's fulfillment of its communication duty to the public is based on manufacturers' comprehensive awareness and communication of their products' risks.

We first noted that examiners were pushing for inclusion of a risk analysis section in 510(k)s about two years ago. I don't think that requirement appears in any of the how-to-submit-a-510(k) guidances yet, but that doesn't make their expectation any less real for submission preparers.

As I'm sure you're aware, it's hard to get along with the FDA if your stance is that you're only required to comply with legally defined regulations, and not their interpretative guidances and policies.
 

Sam Lazzara

Trusted Information Resource
#10
Here is my attempt to address Ronen's queries.

FDA is probably kicking themselves for saying so little about risk management in Part 820 (see excerpt below), but that has not prevented FDA CDRH ODE reviewers and field investigators from expecting manufacturers to apply risk management principles.

820.30(g) Design validation shall include software validation and risk analysis, where appropriate.

When the QSR was released in 1996, ISO 14971:2000 did not exist yet, and EN 1441 (risk analysis) was the current risk standard. Hence, the term "risk analysis" instead of the broader term "risk management".

Although 21 CFR 820 (FDA Quality System Regulation) does not have "risk management" requirements that are as broad as in ISO 13485, FDA has in practice been expecting manufacturers to have a life-cycle risk management process per ISO 14971: 2007, which is an FDA recognized standard.

http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfStandards/detail.cfm?id=19639

As stated at the above link "A declaration of conformity to ISO 14971 may be used to satisfy the risk management needs for a Special 510(k)". This demonstrates FDA's focus on risk management.

The preamble for 21 CFR 820 (written in 1996) says more about risk (called risk analysis at the time) than the regulation itself. The current day risk management concepts per ISO 14971 are broader than just "risk analysis" but the concepts are the same.

http://www.fda.gov/downloads/Medica...dicalDeviceQualitySystemsManual/UCM122806.pdf

Here are some excerpts:

"Risk analysis must be conducted for the majority of devices subject to design controls and is considered to be an essential requirement for medical devices under this regulation, as well as under ISO/CD 13485 and EN 46001."

"FDA agrees that the degree of corrective and preventive action taken to eliminate or minimize actual or potential nonconformities must be appropriate to the magnitude of the problem and commensurate with the risks encountered. FDA cannot dictate in a regulation the degree of action that should be taken because each circumstance will be different, but FDA
does expect the manufacturer to develop procedures for assessing the risk, the actions that need to be taken for different levels of risk, and how to correct or prevent the problem from recurring, depending on that risk assessment."

"FDA has deleted the term ‘‘hazard analysis’’ and replaced it with the term ‘‘risk analysis.’’ FDA’s involvement with the ISO TC 210 made it clear that ‘‘risk analysis’’ is the comprehensive and appropriate term. When conducting a risk analysis, manufacturers are
expected to identify possible hazards associated with the design in both normal and fault conditions. The risks associated with the hazards, including those resulting from user error, should then be calculated in both normal and fault conditions. If any risk is judged unacceptable, it should be reduced to acceptable levels by the appropriate means, for example, by redesign or
warnings. An important part of risk analysis is ensuring that changes made to eliminate or minimize hazards do not introduce new hazards. Tools for conducting such analyses include Failure Mode Effect Analysis and Fault Tree Analysis, among others."
 
Last edited:
Thread starter Similar threads Forum Replies Date
V Examples of Design and Process FMEA for Pharmaceutical Product Development ISO 14971 - Medical Device Risk Management 7
R Use of Design and Process FMEA tools - Kitchen Cabinets Manufacturing Quality Tools, Improvement and Analysis 6
T DFMEA (Design FMEA) - We already have a PFMEA (Process FMEA) FMEA and Control Plans 5
C DFMEA (Design FMEA) and PFMEA (Process FMEA) - Who should lead? FMEA and Control Plans 14
D Would Design & Process FMEA (PCBA) be sufficient to fulfill requirement of Cl 7.1? ISO 13485:2016 - Medical Device Quality Management Systems 2
N How to Calculate RPN in Design FMEA (DFMEA) vs. Process FMEA (PFMEA) FMEA and Control Plans 9
S FMEA Severity Rating - Will a Process Design Change also Change the Severity Rating? FMEA and Control Plans 21
Y Design (DFMEA) and Process (PFMEA) FMEA Failure Mode Identification Design and Development of Products and Processes 8
B Why the Greek god Hephaestus should have done a design FMEA (DFMEA) on his giant robot APQP and PPAP 1
P Design FMEA - Detection Rating criteria ISO 14971 - Medical Device Risk Management 3
I Does anybody use Detection in medical device Design FMEA? ISO 14971 - Medical Device Risk Management 18
P Design FMEA for Industrial Machinery FMEA and Control Plans 3
Marc Definition DFMEA (Design FMEA) - Definition Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 0
D Design FMEA for a component - Should I make the following assumptions? FMEA and Control Plans 7
D Do you need requirements for a Design FMEA FMEA and Control Plans 1
R Development of Design FMEA - Functional Requirements FMEA and Control Plans 1
K Determining Effect of Failure without a DFMEA (Design FMEA) FMEA and Control Plans 1
Moncia Design FMEA - Where to Start FMEA and Control Plans 20
T Would a Leak be a Failure Mode on a Design FMEA? Design and Development of Products and Processes 2
M Severity Reduction in a Design FMEA (DFMEA) FMEA and Control Plans 25
G How to identify Key Characteristics (KC) in a Design FMEA (DFMEA) FMEA and Control Plans 2
V Linking the FMEA Causes & Action Plan with DOE Experiments or Experimentation Design FMEA and Control Plans 1
C AIAG FMEA 4th Ed "Design Assumption" - I Need Examples FMEA and Control Plans 15
R DFMEA (Design FMEA) on Printed Circuit Boards (PCB's) FMEA and Control Plans 2
R Rear Axle Gear Group DFMEA (Design FMEA) - Example wanted FMEA and Control Plans 2
M ISO 14971 Risk Analysis and use of a DFMEA (Design FMEA) ISO 14971 - Medical Device Risk Management 8
A AIAG Design FMEA Severity Rankings - Noncompliance Government Regulation? FMEA and Control Plans 9
Chennaiite DFMEA (Design FMEA) at Vehicle Level - Some inputs needed. FMEA and Control Plans 4
V Biggest bottleneck/hurdle in Implementation of Effective Design FMEA FMEA and Control Plans 4
P Design FMEA (DFMEA) Basics - Help needed IATF 16949 - Automotive Quality Systems Standard 3
S DFMEA (Design FMEA) for a New Line Operation of the Underground Operation FMEA and Control Plans 10
T Electronic Assembly Design FMEA (DFMEA) FMEA and Control Plans 6
V In Design FMEA, how to give Occurrence Rating? FMEA and Control Plans 4
michellemmm Reverse Engineering a Design FMEA using aerospace DFMEA guidelines FMEA and Control Plans 2
M What are Comfort and Convenience Items on a Design FMEA FMEA and Control Plans 4
K Design FMEA Detection Scoring - FMEA 3rd edition manual FMEA and Control Plans 7
O Bulk Material PPAP - Design Matrix instead of a normal Design, Production FMEA APQP and PPAP 3
P Design FMEA (DFMEA) required on all products and their parts? Design and Development of Products and Processes 5
R Can a DFMEA (Design FMEA) effectively analyse design modifications? FMEA and Control Plans 12
A DFMEA (Design FMEA) - Occurance and Detection Ratings FMEA and Control Plans 2
B How to fill s/c and c/c in DFMEA (Design FMEA)? Design and Development of Products and Processes 3
S DFMEA - Is a Design FMEA required? FMEA and Control Plans 3
M DFMEA (Design FMEA) for Trailers or any heavy vehicle design Design and Development of Products and Processes 1
D Bulk Suppliers - Design FMEA Required? Aluminium ingots FMEA and Control Plans 10
M When to start a DFMEA (Design FMEA)? APQP and PPAP 2
K Could a good Design FMEA have prevented this? Death inside centrally locked car FMEA and Control Plans 6
B What is the meaning of the Occurrence in the DFMEA (Design FMEA)? FMEA and Control Plans 8
R DFMEA (Design FMEA) - Document or Record? Document Control Systems, Procedures, Forms and Templates 3
S Do you have any information about Design FMEA? FMEA and Control Plans 1
W DFMEA (Design FMEA) for Bulk Material - Resin for Plastic Injection Moulding FMEA and Control Plans 1