Designing an Ethylene Oxide Sterilization Process

A

ahmed.bpharm

#1
Dear forum Members,

Ethylene Oxide Sterilization Process is complex process and not very simple as mentioned in guidelines and standards. We have validated our process on following parameters;

1) Temperature of Chamber: 65.
2) Humidity of Chamber: 60 -70 %.
3) Exposure Time: 8 hrs.
4) EO gas Admission Temp: 30-40.
5) EO gas Conc: 400 mg / L.
6) Pre-Vacuum: -40 KPa.
7) De-gassing: -40 KPa.
8) Aeration Cycle: 1 hr.

We are using EO-Cross Check chemical indicators to monitor sterilization parameters and SGM Bacillus atropheus 106 for microbiological monitoring.

We are facing now-a-days problem in achieving color change in chemical indicators. Is there any seasonal effect on EtO sterilization Process as our problem started in rainy days.

Expertise in EtO sterilization validation are required for help.


Ahmed Zafar.
 
Elsmar Forum Sponsor

Stijloor

Staff member
Super Moderator
#2
Dear forum Members,

Ethylene Oxide Sterilization Process is complex process and not very simple as mentioned in guidelines and standards. We have validated our process on following parameters;

1) Temperature of Chamber: 65.
2) Humidity of Chamber: 60 -70 %.
3) Exposure Time: 8 hrs.
4) EO gas Admission Temp: 30-40.
5) EO gas Conc: 400 mg / L.
6) Pre-Vacuum: -40 KPa.
7) De-gassing: -40 KPa.
8) Aeration Cycle: 1 hr.

We are using EO-Cross Check chemical indicators to monitor sterilization parameters and SGM Bacillus atropheus 106 for microbiological monitoring.

We are facing now-a-days problem in achieving color change in chemical indicators. Is there any seasonal effect on EtO sterilization Process as our problem started in rainy days.

Expertise in EtO sterilization validation are required for help.


Ahmed Zafar.
Any comments/suggestions for Ahmed?

Thank you!

Stijloor.
 

somashekar

Staff member
Super Moderator
#4
Dear forum Members,

Ethylene Oxide Sterilization Process is complex process and not very simple as mentioned in guidelines and standards. We have validated our process on following parameters;

1) Temperature of Chamber: 65.
2) Humidity of Chamber: 60 -70 %.
3) Exposure Time: 8 hrs.
4) EO gas Admission Temp: 30-40.
5) EO gas Conc: 400 mg / L.
6) Pre-Vacuum: -40 KPa.
7) De-gassing: -40 KPa.
8) Aeration Cycle: 1 hr.

We are using EO-Cross Check chemical indicators to monitor sterilization parameters and SGM Bacillus atropheus 106 for microbiological monitoring.

We are facing now-a-days problem in achieving color change in chemical indicators. Is there any seasonal effect on EtO sterilization Process as our problem started in rainy days.

Expertise in EtO sterilization validation are required for help.


Ahmed Zafar.
At the same time what has been your result from the microbiological monitoring ?
That it started in your rainy days is an observation but this system being a tightly closed loop, I do not envisage an effect due to weather. There could be problems in your EO cross check chemical indicator, its shelflife or your storage conditions of them.
Please do come back with more information and try making a change in your EO cross check chemical indicator to find if the indicator itself is an issue.
 
A

ahmed.bpharm

#5
Dear Somashekar,

First of all thanks for reply, I confirmed the EO cross check chemical indicators by sending them to another competitor and in their sterilizer, all indicators changed their colors according to the standard with less EO concentration in less exposure time.

Recently i have increased my exposure time, results were improved over all but color change is not up to the standard. If temperature of gas admission matters, i have tried both EO / CO2 gas temperature up to 45 degree Celsius and also up to 11 degree Celsius but both have not improved color change.

Apparently EO gas penetration seems the issue, because 30-40 % indicators changes their color but other 70 % remains dull. Before 15 days the chemical indicators were changing their color routinely and accurately but all of sudden in 2 - 3 days the results came poor.

I have tried all the parameters, i.e. applying deep shallow pre-vacuum upto -80 KPa, Controlling flow rate in between 140- 150 L/min, Pre-heating in chamber, Increasing chambers temperature, by increasing and decreasing relative humidity, increasing exposure time and last by decreasing load density and inner spaces between cartons. Nothing is giving any hope for improvement.

Any engineering solution will be highly appreciated after my process parameters.

Thanks,

Ahmed Zafar.
 

Qara123

Involved In Discussions
#6
Apparently EO gas penetration seems the issue, because 30-40 % indicators changes their color but other 70 % remains dull. Before 15 days the chemical indicators were changing their color routinely and accurately but all of sudden in 2 - 3 days the results came poor.
Are the dull indicators randomly distributed across the load or do they tend to be concentrated at a certain location?

Q
 
A

ahmed.bpharm

#7
Dear Q;

Well all the indicators placed at different locations in sterilizer chamber are dull but in some cycles 2-3 locations were identified where indicators changed their color as per standard while other 12 location distributed across whole sterilizer were same dull.

Today we have tried another experiment in which we placed open chemical indicators i.e. not blistered with syringes in unit cartons to check the penetration across two barriers which are outer master carton and inner unit cartons. Whole indicators changed their color equally and upto the standard, which finished my doubt about the penetration of EO /CO2 mixture from large master carton and smaller unit carton. The only thing which remains is now blister paper and poly film the last but not least barrier.
In the light of above experiment, we have decided to place some open blisters of syringes in master carton without unit carton hence reducing one barrier, secondly we pasted some blistered indicators outside the large master cartons to verify the penetration through blister paper and poly film.

hope that the cause of least penetration is due to blister paper and poly film.

Other questions in mind are some process parameters such as ; Chamber temperature, Chamber humidity, Steam pulses, Pressure of Steam, Pre-Heating time, Temperature of EO gas heat exchanger, Flow of EO gas, Pre-Vacuum to evacuate air from the cartons, Exposure time after gas completion, De-gassing and aeration cycles to reduce the EO residues. Some one having command on EO sterilizer will be highly appreciated if guide me through the above said parameters, ?I shall share my parameters as well.

Thanks for all the replies,


Ahmed Zafar.
 

somashekar

Staff member
Super Moderator
#8
Your blister paper is made to permit penetration. However if you resort to printing on this paper any product information, then it will add to reduced levels of penetration. Hope that your printing (if any) is still keeping about 70% clear minus the sealing widths. Are you certain that there is penetration also through polyfilm ?
 
Thread starter Similar threads Forum Replies Date
S Earthing of moving parts, designing tips IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Designing new gauge tracking software IATF 16949 - Automotive Quality Systems Standard 4
A Leadership Role in Designing QMS - Upper Management Support Quality Manager and Management Related Issues 16
S Designing the experiment with minimum run Statistical Analysis Tools, Techniques and SPC 0
S Designing the experiment with minimum run Statistical Analysis Tools, Techniques and SPC 0
B Designing a Sampling Plan - AQL AQL - Acceptable Quality Level 19
F Designing a Hollow Medical Device for Autoclavability Other Medical Device and Orthopedic Related Topics 2
Claes Gefvenberg Designing Supplier Questionnaire Forms Supplier Quality Assurance and other Supplier Issues 25
M Designing a Quality Assurance and Control System Quality Manager and Management Related Issues 3
P Sheet Metal Design: Standard method for designing Electrical Knockouts Design and Development of Products and Processes 7
N Designing an attribute MSA for a new piece of metrology equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
R Line Balance Simulation when Designing a New Production Line Lean in Manufacturing and Service Industries 1
D Designing a Training Course called "Role of the Laboratory Quality Manager" Training - Internal, External, Online and Distance Learning 13
A Designing a Documents Control System from scratch (in a php intranet) Document Control Systems, Procedures, Forms and Templates 5
K Designing a gauge to check the Axis of Snorkel of Axles Design and Development of Products and Processes 5
A Designing a Closed Loop Water System using Adiabatic Cooling Towers Manufacturing and Related Processes 5
R AS9100C Clause 7.3 Design & Development - Designing a System and Procedures AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S API Std 620 vs. API 650 standard? Which is the least restrictive in designing a tank? Manufacturing and Related Processes 1
M Designing a Nonconformance Management Process Nonconformance and Corrective Action 35
Q Are there any threads specifically concerned with designing with AutoCAD? Design and Development of Products and Processes 1
P Designing a "part-specific" Receiving Gauge - Gauge first or job? Design and Development of Products and Processes 4
C Designing Control Charts - Variable with Small Variation and Tight Control Limit Statistical Analysis Tools, Techniques and SPC 20
C Designing control chart for non-normal variables Statistical Analysis Tools, Techniques and SPC 3
A Is the Assembly of the Prototype after Designing considered a Production Process? Design and Development of Products and Processes 11
M Designing a scope for APQP Design and Development of Products and Processes 2
B Designing a Test Part to Assess CMM Operator's Knowledge Design and Development of Products and Processes 9
D Designing an ISO 9001 Internal Audit Training Course - Need Help Internal Auditing 6
S Project Management QMS - Manufacturing and Designing of Special Purposed Machines Design and Development of Products and Processes 5
L Categories for data collection - Designing a form for when errors occur Nonconformance and Corrective Action 6
J Designing a Mil-Std-105 or ANSI/ASQ Z1.4 Sampling Plan Inspection, Prints (Drawings), Testing, Sampling and Related Topics 49
R Designing and building our own machines Design and Development of Products and Processes 5
V Designing an integrated ISO 9000 - ISO 14000 System in a Big Marine Port Design and Development of Products and Processes 3
S US Success in Designing & Manufacturing World-Class Goods Design and Development of Products and Processes 2
Claes Gefvenberg Document control applications - Be very wary about designing your own Document Control Systems, Procedures, Forms and Templates 4
F 8.2.1 Customer Satisfaction - Designing & Building a Solid waste disposal facility ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 36
O Need documentation - Designing a data acquisition system for the enviromental chamber Design and Development of Products and Processes 3
M Informational USFDA Ethylene Oxide Sterilization for Medical Devices website Medical Device and FDA Regulations and Standards News 0
M Informational FDA Posts Information about Ethylene Oxide Sterilization and Medical Devices Medical Device and FDA Regulations and Standards News 0
T ISO 10993-7 - Ethylene oxide sterilization residues Other Medical Device Related Standards 1
E Ethylene Oxide Sterilization Validation-Reference Load Other Medical Device Related Standards 2
D Ethylene Oxide Sterilization aeration Manufacturing and Related Processes 1
D Ethylene Oxide Sterile Implantable Medical Device Shelf Life Extension Other Medical Device Related Standards 1
B Product definition for Ethylene Oxide Sterilization Validation Other Medical Device Related Standards 2
D Adding an extra Label on an Ethylene Oxide Sterilized Box ISO 13485:2016 - Medical Device Quality Management Systems 3
D Switching Master Product in EO (Ethylene Oxide) Other Medical Device Related Standards 5
S ECH (ethylene chlorohydrin) residues - Plastic catheter packed in a sterile bag Other Medical Device Related Standards 5
N Decision to file a new 510(k) - ETO (Ethylene Oxide) Sterilization Other US Medical Device Regulations 4
V Ethylene Oxide Testing Requirements as per US-pharmacopoeia Other US Medical Device Regulations 2
C ETO (Ethylene Oxide) Sterilization Audit Question Other Medical Device Related Standards 9
Sam Lazzara Ethylene Oxide Sterilization Validation question Other Medical Device Related Standards 11