Details of Operational Qualification (OQ) test design for plastic extrusion processes, effect of material property noise, and GHTF/SG3/N99-10

Stoic

Starting to get Involved
#1
I've been trying to identify the industry standard approach to OQ of plastic extrusion processes, and am instead finding a wide range of opinions and practices among medical device OEMs and extrusion suppliers. It's a complicated process with a high number of inputs and a significant noise factor in the resin material property (melt flow index), so some confusion is to be expected. Some individuals recognize the importance of extrusion process validation, while others are strongly disdainful of its engineering or business value or even feasibility. Others grudgingly acknowledge the reality of interpretation and enforcement: extrusion is similar enough to injection molding that most auditors regard validation as mandatory under 21 CFR 820 / MDSAP. However, the implementation can differ significantly from an OQ of a simple process with just one set of time, temperature, and pressure variables.

Practitioners I've encountered generally understand that they have to identify the most sensitive inputs as CTQ and focus the OQ on those. Some use final (as close to the die as possible) melt temperature and haul-off (puller) speed, others may also use screw speed. Some use final melt pressure instead of temperature to compensate for variation in resin MFI. Some of these answers may be better than others, but perhaps the best answer may be different for various combinations of equipment and material.

The compromises begin to creep in when interpreting OQ limits relative to process control limits. Even simpler processes are not immune to misinterpretation. I've encountered scenarios for a variety of process types in which:
  • OQ is regarded as unnecessary because the process has just one "set point" for each CTQ control variable
  • Calibration tolerances of the gauges are not taken into account when determining OQ limits relative to process control limits
In the most extreme cases of extrusion PV, OQ limits are regarded as just starting windows within which the operators will hunt for the combination of input variables that will produce conforming product during a given run. The common defense is that MFI variation (and possibly other noise sources) make it impossible to set OQ limits that guarantee conforming output. This seems to me to conflict with the very definition of OQ as presented in GHTF/SG3/N99-10. While the exercise appears on the surface to check the box for the OQ requirement, it actually represents an implicit statement that the process cannot feasibly be validated and must be actively "tuned" around the noise during every production run, or non-conformances will result.

A less extreme but still debatable approach involves OQ test runs that only vary one CTQ control parameter at a time. While it's a significant improvement over the previous school of thought, it's still far less rigorous than running at the worst-case high/low input combinations. (Imagine proposing this method for a sterile pouch sealing process.) As before, the defense is typically that noise factors make it impossible to run worst-case combinations and get conforming output. I.e., the process can't really be validated at the level of control the equipment provides, the level of precision & accuracy required in the output, and the level of noise that must be compensated.

What have you observed in practice? What are the most capable and effective organizations doing? Are they treating extrusion OQ the same as other processes and running worst-case combinations of the CTQ controls? Are they making compromises relative to validation of other processes? If so, how do they rationalize them?
 
Elsmar Forum Sponsor

Tidge

Trusted Information Resource
#2
I've never done a OQ for an extrusion process, but your thinking seems correct to me. One thing that you report does jump out at me (with alarums similar to your own) is:
In the most extreme cases of extrusion PV, OQ limits are regarded as just starting windows within which the operators will hunt for the combination of input variables that will produce conforming product during a given run. The common defense is that MFI variation (and possibly other noise sources) make it impossible to set OQ limits that guarantee conforming output. This seems to me to conflict with the very definition of OQ as presented in GHTF/SG3/N99-10. While the exercise appears on the surface to check the box for the OQ requirement, it actually represents an implicit statement that the process cannot feasibly be validated and must be actively "tuned" around the noise during every production run, or non-conformances will result.
Any process with operational limits that can't guarantee a conforming output is not one that I would say is "validated", except possibly with some extreme caveats. If my back was against the wall, and I had to do something like a variable challenge and yet I knew there was non-conforming output... I think the best option is to do enough OQ studies to be able to quantify the fraction of non-conforming product so that meaningful downstream controls could be implemented.
 

Stoic

Starting to get Involved
#3
Thanks for your response, Tidge. Agreed that it's better to perform OQ and accept less than 100% conforming process output with sufficient post-process inspection rather than just declare OQ impossible, but only if all of the identified CTQs "can be fully verified". I don't see how this works if you can't pass a destructive OQ test. But often extrusions have only dimensional CTQs, and this is the hardest concept for engineers to accept about the regulations: even when all the (identified) CTQs can be fully verified, validation is still required for thermoplastic processes. (Whether this should be the rule is debatable, but the enforcement pattern since the QSR preamble was published is clear, and the language for the U.S. note about injection molding in the MDSAP documents makes it even clearer.)

In the case of only dimensional CTQs, let's assume that OQ performed at worst-case combinations of critical process control variables +/- calibration tolerance shows a capability that's less than desirable but not catastrophic. Should it not be an acceptable course of action to impose a higher sampling criteria on each lot, possibly up to 100%? You've met the requirement to challenge the process's sensitivity to inputs, your best efforts have shown that there is reason to be concerned, and you adjust your inspection requirements to compensate. As long as your company's quality system incorporates this option, the auditor should be OK with it.
 
Thread starter Similar threads Forum Replies Date
D Audit Report details when ISO 13485:2016 and cGMP 21 CFR 820 are applicable ISO 13485:2016 - Medical Device Quality Management Systems 6
L En 62368 How to complete with incomplete parts details CE Marking (Conformité Européene) / CB Scheme 4
CharlieUK EU representative details on Class I device CE Marking (Conformité Européene) / CB Scheme 3
I PFD (Process Flow Diagram) approach. Setup details as part of PFD? APQP and PPAP 5
Ed Panek Immediate need for 80601-2-56 Consulting expert. PM me for details Career and Occupation Discussions 0
N Authorised Representative details on product label? EU Medical Device Regulations 1
J IMDS submission details for aluminium wrought alloys APQP and PPAP 5
adir88 Amendment 1 of IEC 62366-1 - Details yet? IEC 62366 - Medical Device Usability Engineering 2
I Importer Address in addition to Manufacturer details on label EU Medical Device Regulations 13
A Level of details required for Class IIb Product Verification and Validation CE Marking (Conformité Européene) / CB Scheme 1
optomist1 Recommend Primer Details of Non-Contact White/Blue Light Scanning General Measurement Device and Calibration Topics 0
S ISO 9001:2015 clauses and clause details in Excel format Document Control Systems, Procedures, Forms and Templates 11
R Help on providing Pesticide Details US Food and Drug Administration (FDA) 1
C Document Control - Details to include on Documents Document Control Systems, Procedures, Forms and Templates 2
J FAI for Heat Treated Details - Inco Parts Shrink AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
I Medical Device Registration details for the following Countries ISO 13485:2016 - Medical Device Quality Management Systems 5
Proud Liberal Minitab - Organizing a large number of graphs using file details Using Minitab Software 5
5 Counterfeit Parts Mitigation Plan/Procedure Details Requirement(s) Document Control Systems, Procedures, Forms and Templates 5
C How to Write Procedures without too many Details ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
S Organization Chart - Level of Details? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
L ISO 17021:2011 - The Application of Annex A - Work Instruction Details General Auditing Discussions 9
Z Control Plan & FMEA Specifications - Level of Details FMEA and Control Plans 4
M PSE Mark Japan - Any Details Japan Medical Device Regulations 1
somashekar Label Details and Contents for Reusable Surgical Instruments EU Medical Device Regulations 7
D The best way to Document Process Details in Accounting ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
V Where & How to Capture Details of Periodic Reviews of Change Control & CAPAs Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 8
S Control Plan Details vs. Operator Instruction Sheets - Document Contents IATF 16949 - Automotive Quality Systems Standard 9
Q Content and Details of a Client (Customer) Claim Procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
H Customer Requirements and Request for all Process Details and Documents APQP and PPAP 8
S Details to be covered under Equipment Qualification for Pharmaceutical Products? Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
G Certificate details ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Ajit Basrur FDA to disclose more details on Drug, Food Safety US Food and Drug Administration (FDA) 0
P Must show Sub-Supplier details? Supplying a Tier 2 company IATF 16949 - Automotive Quality Systems Standard 7
M APQP Timing Plan Details and APQP Package Contents (e.g.: floor plan) - Batch Mfg. APQP and PPAP 5
A Clutch Plate Blackening Process - Help with process flow and details Manufacturing and Related Processes 1
Stijloor GM to take back 5 Delphi plants, contract details show World News 0
W QMS Details for an e-services company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
N Details of Clinical Trials of Medical Devices in China ISO 13485:2016 - Medical Device Quality Management Systems 5
B Sloping control chart to take care of tool wear - Details needed Statistical Analysis Tools, Techniques and SPC 5
P Traceability (Details to assembly/sealant) - AS9100 7.5.3 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
T Automotive Software Integrity Level (ASIL) - What are the Requirements & Details? Software Quality Assurance 2
C Gutting the Control Plan of all details and and getting it accepted by the big 3 FMEA and Control Plans 11
B Planekator Calibration Method Details - Calibration of a Surface Plate General Measurement Device and Calibration Topics 6
A AS 9100 Certification Details AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
G TSCA (Toxic Substances Control Act) Inventory List - Details and Samples? Various Other Specifications, Standards, and related Requirements 7
N Details of Nissan (Renault-Nissan aliance) Supplier Quality Assurance expectations Customer and Company Specific Requirements 11
S Design of Experiments - Details of Shainin technique and a training source - India Quality Tools, Improvement and Analysis 5
C Present Employer Contact Details - May we contact your present employer? Career and Occupation Discussions 14
C Toyota Firewall Process - TPS - Does anyone know what this is? Details? FMEA and Control Plans 14
RoxaneB Sarbanes-Oxley : Details requested on environmental connections Miscellaneous Environmental Standards and EMS Related Discussions 2

Similar threads

Top Bottom