Determining a New Product Bioburden Limit

H

Hondo812

#1
Hi,

I have a few questions.

I have a fairly new product and need to determine the bioburden limit. Most of what I have read indicates I am supposed to use historical data to set these limits (ala SPC?). But what if I don't have any historical data for product "X"? How do I specify a limit in that case?

Also, how do I co-relate the data from my sterilization qual that indicates "populations" to bioburden data in "CFU's"? Is there some sort of conversion factor at play here?

All answers are appreciated, correct ones more so.
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
Re: Determining a product bioburden limit

A product that is to be sterilized before use, must have a bioburden that is not excessive to permit the intended sterilization method to work effectively.

A number of standards address the requirements for effective implementation of various sterilization methods. You asked in an ISO 13485 context, so you might want to take a look at http://ec.europa.eu/enterprise/newapproach/standardization/harmstds/reflist/meddevic.html, which lists harmonized standards that may be applicable to medical devices marketed under the MDD.
 
H

Hondo812

#3
Re: Determining a product bioburden limit

Products produced in our cleanroom have had a high of 9 CFU's with an overall average of 3 CFU's. In contrast, we qualify our sterilization methods with spore strips (aka BI's) that have 10^6 spores on them. In some cases we will place 2 or more per device being tested.

All I am looking for is a direct relationship between the spore count and the bioburden CFU count. Keep in mind the spores are known bacteria that is selected because of its resistance to sterilization versus the bioburden "bugs" which are usually easily killed. It's a little bit like comparing apples to oranges yet it is an FDA requirement. I have yet to come across a guidance that makes the comparison between the two clear.
 
M

MIREGMGR

#4
Re: Determining a product bioburden limit

All I am looking for is a direct relationship between the spore count and the bioburden CFU count.
I'm not aware that there's any relationship between the two numbers. What would such a relationship mean...how would you use it?

The conceptual standard for BIs is that they are populated with at least 10^6 spores. BIs are not available with a particular number of spores, that you could obtain as a specified multiple of your number of CFUs.

I reckon the absence of such a guidance may be related to the absence of such a relationship.

...it is an FDA requirement.
Where is that requirement stated?
 

xcanals_tecno-med.es

Involved In Discussions
#5
Re: Determining a product bioburden limit

Hi
in Europe the bioburden reference levels for the product are not established by any standard and must be validated the sterilization process which indicate the bioburden of the products before sterilization and the processing conditions (clean rooms, laminar flow cabines, ...)

For the clean rooms usally the medical devices industry uses the reference levels of the annex 1 of the EU GMP for medicinal products

http://ec.europa.eu/enterprise/pharmaceuticals/eudralex/vol-4/2008_11_25_gmp-an1.pdf

stated in the paragraph:

19. Recommended limits for microbiological monitoring of clean areas during operation:


Recommended limits for microbial contamination (a)

Grade ..........air sample..... settle plates
...................cfu/m3........ (diameter 90 mm)
.................................... cfu/4 hours (b)
-----------------------------------------------
A (ISO4)........ < 1 ............. < 1
B (ISO5) ....... 10 ............. 5 -- implants
C (ISO7) ....... 100 .............50 -- non pirogenic products
D (ISO8) ....... 200 .............100 -- general products

Notes
(a) These are average values.
(b) Individual settle plates may be exposed for less than 4 hours.

So as a rule the level on the product must be 2 to 10 times lower than the maximum allowed for the clean area.

Regards
 
Last edited:

bio_subbu

Super Moderator
#6
Re: Determining a product bioburden limit

Hi,

I have a few questions.

I have a fairly new product and need to determine the bioburden limit. Most of what I have read indicates I am supposed to use historical data to set these limits (ala SPC?). But what if I don't have any historical data for product "X"? How do I specify a limit in that case?

Also, how do I co-relate the data from my sterilization qual that indicates "populations" to bioburden data in "CFU's"? Is there some sort of conversion factor at play here?

All answers are appreciated, correct ones more so.
Refer ISO 11737-1:2006 “Sterilization of medical devices- Microbiological methods – Part 1: Determination of a population of microorganisms on products”.
 
H

Hondo812

#7
Re: Determining a product bioburden limit

Hi
in Europe the bioburden reference levels for the product are not established by any standard and must be validated the sterilization process which indicate the bioburden of the products before sterilization and the processing conditions (clean rooms, laminar flow cabines, ...)

For the clean rooms usally the medical devices industry uses the reference levels of the annex 1 of the EU GMP for medicinal products

http://ec.europa.eu/enterprise/pharmaceuticals/eudralex/vol-4/2008_11_25_gmp-an1.pdf

stated in the paragraph:

19. Recommended limits for microbiological monitoring of clean areas during operation:


Recommended limits for microbial contamination (a)

Grade ..........air sample..... settle plates
...................cfu/m3........ (diameter 90 mm)
.................................... cfu/4 hours (b)
-----------------------------------------------
A (ISO4)........ < 1 ............. < 1
B (ISO5) ....... 10 ............. 5 -- implants
C (ISO7) ....... 100 .............50 -- non pirogenic products
D (ISO8) ....... 200 .............100 -- general products

Notes
(a) These are average values.
(b) Individual settle plates may be exposed for less than 4 hours.

So as a rule the level on the product must be 2 to 10 times lower than the maximum allowed for the clean area.

Regards
Thank you for your informative post. Where did you get "implants" = Class B cleanroom? I did not see that in the EU document. Also, where is the "2 to 10 times lower" specification documented?
 

xcanals_tecno-med.es

Involved In Discussions
#8
Re: Determining a product bioburden limit

Thank you for your informative post. Where did you get "implants" = Class B cleanroom? I did not see that in the EU document. Also, where is the "2 to 10 times lower" specification documented?

The annex 1 of the GMP which are directed to medicinal products establishes
"...
Terminally sterilised products
28. Preparation of components and most products should be done in at least a grade D environment in order to give low risk of microbial and particulate contamination, suitable for filtration and sterilisation. Where the product is at a high or unusual risk of microbial contamination, (for example, because the product actively supports microbial growth or must be held for a long period before sterilisation or is necessarily processed not mainly in closed
vessels), then preparation should be carried out in a grade C environment.
29. Filling of products for terminal sterilisation should be carried out in at least a grade C environment.
..."

The indications of use are based on my experience, and for implants usually the operations are made in a laminar flow cabinet ISO 5 or lower in a cleanroom ISO 7.
Please other experts in the cove indicate if corresponds with your facilities and experiece.


For a new product and need to determine the bioburden limit.
But what if I don't have any historical data for product "X"? How do I specify a limit in that case?
The better is estimate this limit with some measurements of the first batches manufactured and remember to use the method of the standard ISO 11737-1:2006 (which include validation tests) as bio_subbu previously says.

Xavier
 
Last edited:
J

Jimmy the Brit

#9
Sorry so late to join the discussion - I've away travelling for a while.

I have a fairly new product and need to determine the bioburden limit. Most of what I have read indicates I am supposed to use historical data to set these limits (ala SPC?). But what if I don't have any historical data for product "X"? How do I specify a limit in that case?
You need to be pragmatic in determining your bioburden level (not limit - assuming you are using an overkill approach to terminal sterilization). I would advocate doing three reps from each batch of product that you make and apply some common sense. You are looking for significant differences, which microbiologically means a log10 change in bioburden level, seasonal trends of microflora with highly variant process resistance (significantly spore formers for steam or EO, Micrococci & Staphs for radiation) or high recovery variability that needs to be traced to the causative factor (water system sanitization or cleaning schedule, that kind of thing)

Please note the contamination rates stated in the earlier response are nothing to do with product bioburden - just control of the manufacturing environment.

Also, how do I co-relate the data from my sterilization qual that indicates "populations" to bioburden data in "CFU's"? Is there some sort of conversion factor at play here?
I think I see what you are trying to do here. If your sterilization cycle is based on the product bioburden, how do you assure that the BI represents a worst case bioburden when you do not have a long history of data to back it up? As answered by others, the BI organism, with its population in exces of 6 log10, it D value above 1.5 minutes (for steam) is significantly more resistant than normal bioburden.

I demonstrate this by identifying spore formers from product bioburden then boil-testing them for 10 minutes at 100 degrees C - this equates to ~5 seconds at 121 degrees (assuming a Z value of 10 degrees) - we have never had a survivor, which, taken with bioburden levels of <100 fu per unit, assures us that the BI represents a vastly tougher process challenge than our bioburden in our overkill sterilization cycles, both in terms of population and resistance to process.

In an overkill situation the relationship you describe is therefore an overwhelming shift in favour of the BI, and hence patient safety. This still means you have to measure bioburden as above, but in terms of relating the two there should be little in common.

In a bioburden based cycle however the relationship is much more meaningful. The sterilization process is usually detrimental to the product and thus needs to be managed to do the least product harm, either by controlling time, temperature, residuals, pressure, or absorbed dose, whilst delivering the required sterility assurance level. In these instances the BI tends to be a suspension painted onto a product surface at a relevant population level, perhaps with a deliberately manipulated D value to represent product bioburden. As a sterilization nerd I find this all very exciting, but it is intensely expensive, hard to explain to a non-specialist auditor and fraught with failure modes. If this is the approach you are using let me know and I will explain the approach I have used to co-relate BI and bioburden data in a gamma sterilization process.

Hope this helps,

Jimmy
 

somashekar

Staff member
Super Moderator
#10
Hi,

I have a few questions.

I have a fairly new product and need to determine the bioburden limit. Most of what I have read indicates I am supposed to use historical data to set these limits (ala SPC?). But what if I don't have any historical data for product "X"? How do I specify a limit in that case?

Also, how do I co-relate the data from my sterilization qual that indicates "populations" to bioburden data in "CFU's"? Is there some sort of conversion factor at play here?

All answers are appreciated, correct ones more so.
Is your new product a drug or a medical device ? Any information of which USP class in comes under if it is medical device. Ex <20 EU/device
 
Thread starter Similar threads Forum Replies Date
N Reason for determining no adverse effect on reworked product ISO 13485:2016 - Medical Device Quality Management Systems 8
M 7.2.1.c Product requirements - Determining product, enviroment, health, safety, etc IATF 16949 - Automotive Quality Systems Standard 4
S Determining the requirements for the products and services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Q Determining Adverse Effects of Corrective/Preventive Actions ISO 13485:2016 - Medical Device Quality Management Systems 2
S Determining sample sizes for PQ Qualification and Validation (including 21 CFR Part 11) 4
M Determining if an Insulin Pen Testing Machine is a Medical Device? EU Medical Device Regulations 4
M Determining a tolerance value for Measuring devices in-house inspection General Measurement Device and Calibration Topics 12
D Determining the the maximum number of reprocessing cycles of attachments CE Marking (Conformité Européene) / CB Scheme 2
R Determining Uncertainty from Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
A IATF 16949 4.3.1 - Determining the scope of the quality management system - supplemental IATF 16949 - Automotive Quality Systems Standard 9
D ISO 9001:2015 4.3 Determining the Scope of the QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Determining SPC tolerance Statistical Analysis Tools, Techniques and SPC 21
B Determining SAT Offsets vs TUS Offsets per SAE AMS 2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
B Determining sample size for device sterility Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
D Determining of sample size for 'Operational Qualification' AQL - Acceptable Quality Level 5
R Question on determining defective units - I am not recording fixture to part rejected Statistical Analysis Tools, Techniques and SPC 5
S Clause 8.2.2 Determining the requirements for products and services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
D Determining Calibration Frequency schedule for items used in production Manufacturing and Related Processes 2
C Determining if Maintenance Contractor is an External Service subject to ISO 9001 Clause 8.4 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 43
S AS9100D PEAR - Examples for organization's method for determining process results? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
I Determining Calibration Tolerance of a Measurement Device General Measurement Device and Calibration Topics 2
J ISO 17025 Documented Procedure for 6.2.5 - Determining competency ISO 17025 related Discussions 4
V Determining FDA 820 (registration) vs ISO 13485 - Supplier gives us the kit ISO 13485:2016 - Medical Device Quality Management Systems 1
J ISO 9001 8.4.1 - Determining controls applied to externally provided processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
E Determining what is good and what is bad can be subjective - when is it a quality issue? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
F Determining what type of scrap to include in my internal PPM calculation Quality Manager and Management Related Issues 5
M Determining number of employees within the "Scope" of the QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
A Determining the Scope of the QMS during Stage 1? Registrars and Notified Bodies 11
W Minor Audit Nonconformance Against Determining the scope of QMS IATF 16949 - Automotive Quality Systems Standard 12
D Determining Critical Components for conformity with IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 21
Q ISO 9001, section 4.3 Determining the scope of our QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
R Determining Sample Size for Medical Device Component Validation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
A What does 8.2.2.1, Determining the requirements related to products and services,mean IATF 16949 - Automotive Quality Systems Standard 1
A Determining Retention Period for Medical Device QMS documents Document Control Systems, Procedures, Forms and Templates 5
S Surveillance Sampling Test - Determining Sample Size Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
F ISO 9001:2015 4.3 - Determining the scope of the quality management system ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
B Standards Needed In House - Determining what standards are applicable Various Other Specifications, Standards, and related Requirements 3
W Determining Medical Device Classification in Mexico Other Medical Device Regulations World-Wide 5
K Determining Effect of Failure without a DFMEA (Design FMEA) FMEA and Control Plans 1
W Determining the Status and Importance of the Processes and Areas to be Audited Internal Auditing 7
T Determining Customer Requirements for the US Postal Service (USPS) IATF 16949 - Automotive Quality Systems Standard 4
Crimpshrine13 Rules of achieving and maintaining IATF recognition - Determining audit days IATF 16949 - Automotive Quality Systems Standard 2
R Developing procedure for Determining Company's Context And Interested Parties ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 53
W Determining Asset Life or Depreciation Life for M&TE General Measurement Device and Calibration Topics 1
N Procedure for determining pinhole position in condom Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
alonFAI Determining PCBA Xray Test Sampling Size Statistical Analysis Tools, Techniques and SPC 1
S Determining sample size for inspection to achieve x% confidence re defects Misc. Quality Assurance and Business Systems Related Topics 10
S Determining Sample Size - AQL & LQ AQL - Acceptable Quality Level 10
R Determining the feasibility/scope for FDA IQ/OQ against a very simple application US Food and Drug Administration (FDA) 2
B Determining the "storage volume" (ISO 8536-8:2004 clause 6.14) Other Medical Device Related Standards 4