Determining Adverse Effects of Corrective/Preventive Actions

#1
Per ISO 13485:2016, I am trying to incorporate a process for determination of adverse effects incurred from Corrective/Preventive Actions. I have two questions.

1) Should determination of adverse effects be documented while determining which corrective/preventive actions to take, or should it be documented after the actions are in place? Maybe both? I could see an auditor wanting me to consider and document determination of adverse effects before corrective/preventive actions are decided on, but I could also see an auditor wanting to see documented determination of adverse effects during effectiveness review when the corrective/preventive actions are already in place. Should I do both?

2) Adverse effects are always on the mind of our Quality team so I am not really worried about us forgetting to think about adverse effects when determining actions to take. Obviously though, determination of adverse effects needs documented somewhere for CAPA's. With that being said, does anyone have any suggestions or examples of a simple standard for reviewing and documenting adverse effects? I want something that won't take a lot of time to complete, but will address and provide documentation for determining adverse effects from corrective/preventive actions.
 
Elsmar Forum Sponsor

Tagin

Trusted Information Resource
#2
1) Strictly speaking, it says "verifying that the action does not adversely...", which would probably indicate -after-the-fact. However, I think it should also be considered (and mentioned) as part of "planning and documenting action needed..."

2) Maybe if you can adjust your form accordingly, you could explicitly list the 3 areas related to adverse effects. E.g.
Will the planned corrective/preventative action adversely affect the medical device's:
  • Regulatory requirements
  • Safety
  • Performance
and in the verification/effectiveness:

Did the implemented corrective/preventative action adversely affect the medical device's:
  • Regulatory requirements
  • Safety
  • Performance
 

somashekar

Staff member
Super Moderator
#3
Per ISO 13485:2016, I am trying to incorporate a process for determination of adverse effects incurred from Corrective/Preventive Actions. I have two questions.

1) Should determination of adverse effects be documented while determining which corrective/preventive actions to take, or should it be documented after the actions are in place? Maybe both? I could see an auditor wanting me to consider and document determination of adverse effects before corrective/preventive actions are decided on, but I could also see an auditor wanting to see documented determination of adverse effects during effectiveness review when the corrective/preventive actions are already in place. Should I do both?

2) Adverse effects are always on the mind of our Quality team so I am not really worried about us forgetting to think about adverse effects when determining actions to take. Obviously though, determination of adverse effects needs documented somewhere for CAPA's. With that being said, does anyone have any suggestions or examples of a simple standard for reviewing and documenting adverse effects? I want something that won't take a lot of time to complete, but will address and provide documentation for determining adverse effects from corrective/preventive actions.
Are you over thinking .....perhaps yes.
When you determine the corrective action after your root cause analysis, your focus will be on eliminating the root cause. At this instance, you are required to also see outward if any adverse affects could occur due to your corrective action changes upon the safety and performance of your medical device, as well as on the meeting of the applicable regulatory requirements.
With experience and with suitable team work you will begin to evalute your corrective action concurently addressing these areas of your medical device. Its a alert here to make sure that your corrective actions are not taken in isolation without looking into the regulatory requirements, safety requirements and performance requirements of your medical device.
Do this for the purpose of your medical device safety and performance as a discipline. Involve your regulatory affairs team / lead to assess from the regulatory angle. Do not look at any situation from the auditor view point, but approch from the standards intent. You will be good.
 
Thread starter Similar threads Forum Replies Date
N Reason for determining no adverse effect on reworked product ISO 13485:2016 - Medical Device Quality Management Systems 8
S Determining the requirements for the products and services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Determining sample sizes for PQ Qualification and Validation (including 21 CFR Part 11) 4
M Determining if an Insulin Pen Testing Machine is a Medical Device? EU Medical Device Regulations 4
M Determining a tolerance value for Measuring devices in-house inspection General Measurement Device and Calibration Topics 12
D Determining the the maximum number of reprocessing cycles of attachments CE Marking (Conformité Européene) / CB Scheme 2
R Determining Uncertainty from Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
A IATF 16949 4.3.1 - Determining the scope of the quality management system - supplemental IATF 16949 - Automotive Quality Systems Standard 9
D ISO 9001:2015 4.3 Determining the Scope of the QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Determining SPC tolerance Statistical Analysis Tools, Techniques and SPC 21
B Determining SAT Offsets vs TUS Offsets per SAE AMS 2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
B Determining sample size for device sterility Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
D Determining of sample size for 'Operational Qualification' AQL - Acceptable Quality Level 5
R Question on determining defective units - I am not recording fixture to part rejected Statistical Analysis Tools, Techniques and SPC 5
S Clause 8.2.2 Determining the requirements for products and services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
D Determining Calibration Frequency schedule for items used in production Manufacturing and Related Processes 2
C Determining if Maintenance Contractor is an External Service subject to ISO 9001 Clause 8.4 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 43
S AS9100D PEAR - Examples for organization's method for determining process results? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
I Determining Calibration Tolerance of a Measurement Device General Measurement Device and Calibration Topics 2
J ISO 17025 Documented Procedure for 6.2.5 - Determining competency ISO 17025 related Discussions 4
V Determining FDA 820 (registration) vs ISO 13485 - Supplier gives us the kit ISO 13485:2016 - Medical Device Quality Management Systems 1
J ISO 9001 8.4.1 - Determining controls applied to externally provided processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
E Determining what is good and what is bad can be subjective - when is it a quality issue? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
F Determining what type of scrap to include in my internal PPM calculation Quality Manager and Management Related Issues 5
M Determining number of employees within the "Scope" of the QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
A Determining the Scope of the QMS during Stage 1? Registrars and Notified Bodies 11
W Minor Audit Nonconformance Against Determining the scope of QMS IATF 16949 - Automotive Quality Systems Standard 12
D Determining Critical Components for conformity with IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 21
Q ISO 9001, section 4.3 Determining the scope of our QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
R Determining Sample Size for Medical Device Component Validation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
A What does 8.2.2.1, Determining the requirements related to products and services,mean IATF 16949 - Automotive Quality Systems Standard 1
A Determining Retention Period for Medical Device QMS documents Document Control Systems, Procedures, Forms and Templates 5
S Surveillance Sampling Test - Determining Sample Size Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
F ISO 9001:2015 4.3 - Determining the scope of the quality management system ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
B Standards Needed In House - Determining what standards are applicable Various Other Specifications, Standards, and related Requirements 3
W Determining Medical Device Classification in Mexico Other Medical Device Regulations World-Wide 5
K Determining Effect of Failure without a DFMEA (Design FMEA) FMEA and Control Plans 1
W Determining the Status and Importance of the Processes and Areas to be Audited Internal Auditing 7
T Determining Customer Requirements for the US Postal Service (USPS) IATF 16949 - Automotive Quality Systems Standard 4
Crimpshrine13 Rules of achieving and maintaining IATF recognition - Determining audit days IATF 16949 - Automotive Quality Systems Standard 2
R Developing procedure for Determining Company's Context And Interested Parties ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 53
W Determining Asset Life or Depreciation Life for M&TE General Measurement Device and Calibration Topics 1
N Procedure for determining pinhole position in condom Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
alonFAI Determining PCBA Xray Test Sampling Size Statistical Analysis Tools, Techniques and SPC 1
S Determining sample size for inspection to achieve x% confidence re defects Misc. Quality Assurance and Business Systems Related Topics 10
S Determining Sample Size - AQL & LQ AQL - Acceptable Quality Level 10
R Determining the feasibility/scope for FDA IQ/OQ against a very simple application US Food and Drug Administration (FDA) 2
B Determining the "storage volume" (ISO 8536-8:2004 clause 6.14) Other Medical Device Related Standards 4
N Procedure/calculation for determining the foot pressure of dial thickness gauge General Measurement Device and Calibration Topics 11
B Bayes Success-Run on Attribute Data - Determining Sample Size Statistical Analysis Tools, Techniques and SPC 2

Similar threads

Top Bottom