When you identify critical components based on the Risk analysis, what is the further action? Do you have more strict incoming inspection or something similar?
 

Ronen E

Problem Solver
Moderator
Maybe the following (from IEC 60601-1 Ed. 3.1 Annex A) can provide some clues:
[...] These capacitors are safety critical since they provide a leakage path to earth or across a barrier. So they must be certified and monitored by a recognised test house to IEC 60384-14, which serves to control their manufacture.
The standard doesn't use the term "critical" systematically, except for software, and "critical components" or similar doesn't appear in it at all.
 

Kumarsc

Registered
Hello Peter Selvey,
Thanks a lot for the detailed explanation. It looks very plausible.

So, to sum up, I can distinguish two critical components lists (ССL): "live" CCL and CCL "for a testing agency".

"Live" CCL is based on risk analysis and involves evaluating risks of appearance HAZAROUS SITATIONS i.e. if malfunction of a component can lead to electrical shock, fire, violation of EMI requirements or deviations from ESSENTIAL PERFORMANCE etc., so, this component should be recognized as Critical component. It's important to make your own "live" CCL to ensure your device is really safe and reliable.

CCL "for a testing agency" is a formalized table, which depends on a particular agency and is based rather on an established practice than on genuine risk analysis. It can be reasonable. CB test (at least ours) doesn't involve evaluation of a risk analysis file, therefore it's easier to use forms "to fill in" than try to evaluate reasonability of including each component to CCL and actually to accomplish risk analysis one more time.

So, an algorithm of creating CCL "for a testing agency" can be like this:
1. To accomplish risk analysis and create your own "live" CCL.
2. Basing on information from the Internet (special thanks to the members of this thread) about "typical" critical components create rough CCL "for a testing agency".
3. In cooperation with the particular test agency clarify CCL "for a testing agency"

I think, it's will be correct to use both CCLs for a production process.
Hello,
Thanks for the detailed explanation.
Generally, whether applied parts that need bio-compatibility to be considered as critical components for testing Lab?
Also what about secondary batteries, whether they are considered as critical components, just because they are secondary batteries?
 