Our ME product plugs into a card in a standard computer which is mounted on a crash cart. The cart has an isolation transformer. We purchase the cart and it has fuse holders which are loaded with 3.15A fuses and marked as such. I add a label where the power cord plugs in telling VAC/Hz/A along with the type B "little man" symbol. The current on this label is what the whole system draws. Our distributor is telling me having both these values for current on the cart makes it confusing. I am trying to look through 60601-1 to see if there is language to support the labeling either way. Do I need to include the actual current draw on a label, or is the fuse rating all that is needed?
Second part, is having a fuse rating of 2X our current draw acceptable, or is there more to it than that?
Elsmar Forum Sponsor
Pads38
Trusted Information Resource
#3
Clause 7.2.7 (Ed. 3)
The rated input shall be given in amperes or volt-amperes or where the power factor exceeds 0.9 in watts.
Clause 7.2.12
Where the fuse-holder is an accessible part the type and full rating of the fuse (voltage, current, operating speed and breaking capacity) shall be marked adjacent to the fuse-holder.
So both markings are required.
As to whether 2x current is the right rating for the fuse (seems high to me); well the ratings of the fuses must be such that hazards are prevented during single fault condition tests (motor stalls etc.).
jradford
#4
So following 7.2.6 & 7.2.7 it makes sense that our label for input power looks like:
~210-230V / 50-60Hz / 1.5A
I will point our distributor to those sections of 60601-1.
Thread starter Similar threads Forum Replies Date
D ISO 9001:2015 4.3 Determining the Scope of the QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Determining SPC tolerance Statistical Analysis Tools, Techniques and SPC 21
B Determining SAT Offsets vs TUS Offsets per SAE AMS 2750E AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 0
B Determining sample size for device sterility Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
D Determining of sample size for 'Operational Qualification' AQL - Acceptable Quality Level 3
R Question on determining defective units - I am not recording fixture to part rejected Statistical Analysis Tools, Techniques and SPC 5
S Clause 8.2.2 Determining the requirements for products and services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
D Determining Calibration Frequency schedule for items used in production Manufacturing and Related Processes 2
C Determining if Maintenance Contractor is an External Service subject to ISO 9001 Clause 8.4 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 43
S AS9100D PEAR - Examples for organization's method for determining process results? AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 5
I Determining Calibration Tolerance of a Measurement Device General Measurement Device and Calibration Topics 2
J ISO 17025 Documented Procedure for 6.2.5 - Determining competency ISO 17025 related Discussions 4
V Determining FDA 820 (registration) vs ISO 13485 - Supplier gives us the kit ISO 13485:2016 - Medical Device Quality Management Systems 1
J ISO 9001 8.4.1 - Determining controls applied to externally provided processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
E Determining what is good and what is bad can be subjective - when is it a quality issue? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
F Determining what type of scrap to include in my internal PPM calculation Quality Manager and Management Related Issues 5
M Determining number of employees within the "Scope" of the QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
A Determining the Scope of the QMS during Stage 1? Registrars and Notified Bodies 11
W Minor Audit Nonconformance Against Determining the scope of QMS IATF 16949 - Automotive Quality Systems Standard 12
D Determining Critical Components for conformity with IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 21
Q ISO 9001, section 4.3 Determining the scope of our QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
R Determining Sample Size for Medical Device Component Validation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
A What does 8.2.2.1, Determining the requirements related to products and services,mean IATF 16949 - Automotive Quality Systems Standard 1
A Determining Retention Period for Medical Device QMS documents Document Control Systems, Procedures, Forms and Templates 5
S Surveillance Sampling Test - Determining Sample Size Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
F ISO 9001:2015 4.3 - Determining the scope of the quality management system ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
B Standards Needed In House - Determining what standards are applicable Various Other Specifications, Standards, and related Requirements 3
W Determining Medical Device Classification in Mexico Other Medical Device Regulations World-Wide 5
K Determining Effect of Failure without a DFMEA (Design FMEA) FMEA and Control Plans 1
W Determining the Status and Importance of the Processes and Areas to be Audited Internal Auditing 7
T Determining Customer Requirements for the US Postal Service (USPS) IATF 16949 - Automotive Quality Systems Standard 4
Crimpshrine13 Rules of achieving and maintaining IATF recognition - Determining audit days IATF 16949 - Automotive Quality Systems Standard 2
R Developing procedure for Determining Company's Context And Interested Parties ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 53
W Determining Asset Life or Depreciation Life for M&TE General Measurement Device and Calibration Topics 1
N Reason for determining no adverse effect on reworked product ISO 13485:2016 - Medical Device Quality Management Systems 7
N Procedure for determining pinhole position in condom Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
alonFAI Determining PCBA Xray Test Sampling Size Statistical Analysis Tools, Techniques and SPC 1
S Determining sample size for inspection to achieve x% confidence re defects Misc. Quality Assurance and Business Systems Related Topics 10
S Determining Sample Size - AQL & LQ AQL - Acceptable Quality Level 10
R Determining the feasibility/scope for FDA IQ/OQ against a very simple application US Food and Drug Administration (FDA) 2
B Determining the "storage volume" (ISO 8536-8:2004 clause 6.14) Other Medical Device Related Standards 4
N Procedure/calculation for determining the foot pressure of dial thickness gauge General Measurement Device and Calibration Topics 10
B Bayes Success-Run on Attribute Data - Determining Sample Size Statistical Analysis Tools, Techniques and SPC 2
H Determining Takt Time in Metal Stamping Lean in Manufacturing and Service Industries 9
J Determining if there is Tin Plating on an Aluminum Bar Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
C Determining PFMEA Severity for Armored Parts FMEA and Control Plans 5
E Correlated data or not? Determining the significance of differences in means Statistical Analysis Tools, Techniques and SPC 3
T Determining What to Test - Prop 65 Testing RoHS, REACH, ELV, IMDS and Restricted Substances 2
M Determining c=0 AQL numbers Misc. Quality Assurance and Business Systems Related Topics 3
M AMS2770 Aluminum Heat Treat - Determining Soak Time Various Other Specifications, Standards, and related Requirements 1
Similar threads
Top Bottom