Determining Sample Size for FDA Verification and Validation Activities
#41
Oh it make sense to me...but when you have an attribute characteristic, Option 1 is not viable. I agree if comparing levels or vendors during OQ then c=0 is inappropriate. So how do you determine a sample size for a PQ when I am testing if a product can withstand a certain pressure for 30 seconds. The test is basically to apply a fixed pressure to the device for 30 seconds. If it does not break it passes. We have very few if any parts that fail. So my only recourse is a c=0 plan, but if I understand you correctly there might be a better way of generating a sample size. Any help would be appreciated.
For PQ (as stated in 3. in my response) a c=0 plan is OK as PQ is a 'dress rehearsal' - if you will - of standard production prior to launch and release. The goal is to determine if the process will produce X number of sequential lots that met the quality/release criteria - that is what acceptance sampling plans are intended to do.
for validation in general, when faced with an acceptance test on a characteristic that rarely fails I tend towards some type of directed testing(test to failure, accelerated or worst case testing) for comparisons.
my other issue is simply that c=0 plans were a deluded attempt to avoid the appearance accepting lots that knowingly had defects in them. c=0 plans don't guarantee or imply that the lot is defect free; which is what some people want to believe. c=0 plans have fairly flat OC curves leaving a lot of gray area between acceptable and unacceptable defect rates. If the plan is RQL based, the protection can be pretty decent but if the plan is AQL based it is quite deceptive. Acceptance sampling plans that use both an AQL and RQL have much steeper OC curves but have larger sample sizes and they often allow for one or more defects in the sample...
Jojo63
#42
I search some information on sample size to validate the functionality of a product. I read the previous posts and I concluded.
I use the simplest formula for a 99% confidence interval.
I perform a first statement of 30 measures and the results are : average 17.4 N.m and standard deviation : 1.3. I want a accuracy of 0.5. My real average not deviate more than 0.5 from my calculated average.
So I get a sample n=45. Is it correct?
Thanks you for your feedback.
B Bayes Success-Run on Attribute Data - Determining Sample Size Statistical Analysis Tools, Techniques and SPC 2
A Determining Sample Size in Design V&V activities 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 44
G Determining Sample Size for Equipment Qualification Studies Qualification and Validation (including 21 CFR Part 11) 10
D Determining Sample Size for Capability in Six Sigma Project Six Sigma 5
H Determining the Correct Sample Size when AQL points to two Sample Sizes AQL - Acceptable Quality Level 7
L Determining Cpk based on a given Sample Size, Confidence Level, & Reliability Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
M Determining Sample Size needed for Operational Qualification (OQ) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 24
T Determining the appropriate Sample Size using an AQL Table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 29
J Functional Test Sampling - Determining Sample Size to eliminate 100% Testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
W Sampling Plan problem - Determining Sample Size for both Attribute and Variables Data ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Determining the Total Sample Size using AQL - AQL Lot Quantity question AQL - Acceptable Quality Level 8
1 Determining AQL and Sample Plan - Batch size is from 1-500 AQL - Acceptable Quality Level 24
E Determining Sample Size for Design Verification and Design Validation Other Medical Device and Orthopedic Related Topics 20
N Auditing Proof of Delivery signed by the customer - Determining Sample Size General Auditing Discussions 6
P Determining Sample Size - Filling a bag with three different types of candy Inspection, Prints (Drawings), Testing, Sampling and Related Topics 31
D Determining sample size required based on a 95% confidence level Reliability Analysis - Predictions, Testing and Standards 7
B Determining sample size for drop test and waterproof testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
T Determining Receiving Inspection Sample Size Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
N Determining Sample Size for Attribute Data Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
C Internal Audit Sampling Plans - Determining Internal Audit Sample Size Internal Auditing 45
Q Determining the Appropriate Sample Size for Process Validation/Qualification Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
T Using Cpk for the purpose of determining sample size/frequencies Capability, Accuracy and Stability - Processes, Machines, etc. 2
G Determining Sample Size and Frequency for Features on Control Plans FMEA and Control Plans 8
R Determining Sample Sizes for Testing (Chemistry or Microbiology) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
B Selecting a Sample and Determining the Reference Value for a Bias Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
R Question on determining defective units - I am not recording fixture to part rejected Statistical Analysis Tools, Techniques and SPC 5
S Clause 8.2.2 Determining the requirements for products and services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V Determining FDA 820 (registration) vs ISO 13485 - Supplier gives us the kit ISO 13485:2016 - Medical Device Quality Management Systems 1