As always, thanks Bev! You are amazingly helpful for those statistically-challenged such as myself!
For the benefit of other Covers, and to be sure I've got this right, I'd like to maybe go step-by-step through a concrete, hypothetical, example.
Suppose specification is 20 +/- 2
Step 1: Estimate Variance (standard deviation, SD)
We could just "guesstimate" the variance. We are confident the process produces consistent outputs so the variance should be relatively small (SD<1).
Or, alternatively, we make a rough estimate by taking STDEV from, say, the first 10 units (
aside: would this need to be justified?)
STDEV(20,20,21,20,19,18,20,21,20,19) = 0.92
So, for this example,
SD = 0.92
Step 2: What is the allowable tolerance (delta, d)
This is taken from the specification (20 +/- 2). So the tolerance is:
d = 2
Step 3: What is the sought confidence interval (t = 1 - alpha risk)?
We will choose 95%. Looking up on the t-table, this is a value of:
t = 1.96
Step 4: Calculate the sample size
We use the formula for continuous data: n = [t*SD / d]^2
= [(1.96)*(0.92) / 2]^2 = 0.81
Rounding to the nearest integer, gives us
n = 1
Hence, according to the calculations above, only 1 sample is required in this example.
Do I have this correct?