Sorry if this is posted in the wrong forum. I am trying to determine the sample size needed for Operational Qualification (OQ). I have ran some parts in the past, thus I can calculate my current mean, std, and CpK etc. How do I figure out what the sample size should be to pass a particular tolerance? For example, if one of OQ output is length (1" ± 0.5") and currently my mean is 1.05 with stdv 0.005, how would I calculate the sample size? Any help would be appreciated. I have Excel and Minitab 15 at my disposal.
mdurivage
I would try a tolerance interval. A tolerance interval will not give a sample size directly, but it will give an interval that one can state that you are 95% confident will contain 99% of the product produced. The calculation uses a sample size. So one may work backwords to get the required sample size.
*Normal distribution assumed
*Percent confidence and % contained can vary to the situation
Mark
mipelamo
Mark,
Thanks for the reply. I can calculate tolerance interval in Minitab, but do you know the equation?
May I ask what you think the intent of OQ is?
OQ is defined as validating that a process can meet the output specification when the inputs are varied form their allowed minimum to their allowed maximum.
The use of a tolerance interval (intended to estimate the variation of the individual values) is not a bad way to iterate to a sufficient sample size. Common sample size formulas are based on estimations of the mean so they are not useful for OQ or PQ...the only caveat is that tolerance intervals rely heavily on the assumption of Normality of the data.
mdurivage
Here is a formula that I sometimes use to determine a sample size:
n= ln(1-C)/ln(R)
Where:
n= the required sample size
C = confidence level
R = reliability
For exaplme if you want to be 95% that you are 95% relaible you wood need 59 good (failure free parts).
Mark
can you provide the derivation or a reference?
...not questioning your formula, its just helpful to know where formulas come from so we can all understnad what they are appropriately used for...
mdurivage
Reliability Statistics by Robert Dovich page 41
mipelamo
Thanks for all the replies. I ended up just using the Power and Sample Size for One-Way ANOVA tool from Mini-Tab. I entered a standard deviation, Power value and maximum difference b/t the means, and then MiniTab calculated the sample size.
B Bayes Success-Run on Attribute Data - Determining Sample Size Statistical Analysis Tools, Techniques and SPC 2
A Determining Sample Size in Design V&V activities 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 44
G Determining Sample Size for Equipment Qualification Studies Qualification and Validation (including 21 CFR Part 11) 10
D Determining Sample Size for Capability in Six Sigma Project Six Sigma 5
H Determining the Correct Sample Size when AQL points to two Sample Sizes AQL - Acceptable Quality Level 7
L Determining Cpk based on a given Sample Size, Confidence Level, & Reliability Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
T Determining the appropriate Sample Size using an AQL Table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 29
J Functional Test Sampling - Determining Sample Size to eliminate 100% Testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
W Sampling Plan problem - Determining Sample Size for both Attribute and Variables Data ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Determining the Total Sample Size using AQL - AQL Lot Quantity question AQL - Acceptable Quality Level 8
1 Determining AQL and Sample Plan - Batch size is from 1-500 AQL - Acceptable Quality Level 24
E Determining Sample Size for Design Verification and Design Validation Other Medical Device and Orthopedic Related Topics 20
N Auditing Proof of Delivery signed by the customer - Determining Sample Size General Auditing Discussions 6
P Determining Sample Size - Filling a bag with three different types of candy Inspection, Prints (Drawings), Testing, Sampling and Related Topics 31
Q Determining Sample Size for FDA Verification and Validation Activities Inspection, Prints (Drawings), Testing, Sampling and Related Topics 41
D Determining sample size required based on a 95% confidence level Reliability Analysis - Predictions, Testing and Standards 7
B Determining sample size for drop test and waterproof testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
T Determining Receiving Inspection Sample Size Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
N Determining Sample Size for Attribute Data Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
C Internal Audit Sampling Plans - Determining Internal Audit Sample Size Internal Auditing 45
Q Determining the Appropriate Sample Size for Process Validation/Qualification Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
T Using Cpk for the purpose of determining sample size/frequencies Capability, Accuracy and Stability - Processes, Machines, etc. 2
G Determining Sample Size and Frequency for Features on Control Plans FMEA and Control Plans 8
R Determining Sample Sizes for Testing (Chemistry or Microbiology) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
B Selecting a Sample and Determining the Reference Value for a Bias Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
R Question on determining defective units - I am not recording fixture to part rejected Statistical Analysis Tools, Techniques and SPC 5
S Clause 8.2.2 Determining the requirements for products and services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V Determining FDA 820 (registration) vs ISO 13485 - Supplier gives us the kit ISO 13485:2016 - Medical Device Quality Management Systems 1