Determining sample sizes for PQ
Can you give us more context and detail? What are you trying to do? This appears to be unrelated to the original thread topic, so we don't have much to go on.
Johnnymo62
Haste Makes Waste
I ended up doing capability on each cavity individually because the cavities varied too much, compared to each other, to be capable on many of our injection molds.
SylvieS
Sorry - I agree that my question was oddly phrased.
If I have a multi-cavity tool - let's say 2 cavities, to do the PQ I will evaluate using the Binomial Law the quantity n of sample required for a specific confidence and reliability level I wish to achieve.
My doubt is about the unit for n. Should the unit be 1 shot/run of my tool (1 unit / n=1 / 2 cavities => 2 products) or n=1 is 1 product?
I hope it clarifies.
This is a common scenario, but interestingly enough, it is never addressed in texts/manuals on statistics. This may have a significant impact on the concept of rational subgroups and the warning against mixing process streams, but no guidance is given on the appropriate way to handle it.
So here are my thoughts on the subject. To keep it simple, we will continue your 2-cavity mold scenario. The two cavities will not be identical, and therefore may represent two process streams. If they are indeed different, they should not be included in the same subgroup and should not be combined into the same capability study.
My recommendation would be to collect samples from each cavity and perform a 2-sample t-test on the cavity means. If the t-test is statistically significant, treat the cavities as different process streams. If the test is NOT statistically significant, treat them as a single process stream. You may safely include them in the same subgroups.
The next thing that I would consider for your specific question is, what is the major source of variation? Is it the variation between the two cavities? Or is it the variation from heat to heat? The samples should focus on the major source of variation that has the potential to vary over time. The difference between cavities will likely not vary until a long time passes, while heat to heat variation may occur over a much shorter time frame.
H Determining the Correct Sample Size when AQL points to two Sample Sizes AQL - Acceptable Quality Level 7
R Determining Sample Sizes for Testing (Chemistry or Microbiology) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
U Determining population defect rate from sample defect rate Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
S Determining Sample Size and Method Quality Tools, Improvement and Analysis 12
B Bayes Success-Run on Attribute Data - Determining Sample Size Statistical Analysis Tools, Techniques and SPC 2
A Determining Sample Size in Design V&V activities 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 44
G Determining Sample Size for Equipment Qualification Studies Qualification and Validation (including 21 CFR Part 11) 10
D Determining Sample Size for Capability in Six Sigma Project Six Sigma 5
L Determining Cpk based on a given Sample Size, Confidence Level, & Reliability Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
M Determining Sample Size needed for Operational Qualification (OQ) Inspection, Prints (Drawings), Testing, Sampling and Related Topics 24
T Determining the appropriate Sample Size using an AQL Table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 31
J Functional Test Sampling - Determining Sample Size to eliminate 100% Testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
W Sampling Plan problem - Determining Sample Size for both Attribute and Variables Data ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Determining the Total Sample Size using AQL - AQL Lot Quantity question AQL - Acceptable Quality Level 8
1 Determining AQL and Sample Plan - Batch size is from 1-500 AQL - Acceptable Quality Level 24
E Determining Sample Size for Design Verification and Design Validation Other Medical Device and Orthopedic Related Topics 20
N Auditing Proof of Delivery signed by the customer - Determining Sample Size General Auditing Discussions 6
P Determining Sample Size - Filling a bag with three different types of candy Inspection, Prints (Drawings), Testing, Sampling and Related Topics 31
Q Determining Sample Size for FDA Verification and Validation Activities Inspection, Prints (Drawings), Testing, Sampling and Related Topics 42
D Determining sample size required based on a 95% confidence level Reliability Analysis - Predictions, Testing and Standards 7
B Determining sample size for drop test and waterproof testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
B Selecting a Sample and Determining the Reference Value for a Bias Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
T Determining Receiving Inspection Sample Size Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
N Determining Sample Size for Attribute Data Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
C Internal Audit Sampling Plans - Determining Internal Audit Sample Size Internal Auditing 45
Q Determining the Appropriate Sample Size for Process Validation/Qualification Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
T Using Cpk for the purpose of determining sample size/frequencies Capability, Accuracy and Stability - Processes, Machines, etc. 2
G Determining Sample Size and Frequency for Features on Control Plans FMEA and Control Plans 8
L Determining effectiveness and efficiency Process Maps, Process Mapping and Turtle Diagrams 6
C Determining an as received OOT condition during calibration does not affect measurements taken prior to calibration. General Measurement Device and Calibration Topics 16
C Need help in determining applicable clause for an audit finding (based on AS9120B) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Q Kap.4.3 Determining the scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
T Determining FIPS 140-2 compliance in a medical device Medical Information Technology, Medical Software and Health Informatics 2
W Strategy for determining which components from a system should be "ME EQUIPMENT" -- home healthcare environment IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
J Process Capability - Determining the process capability of certain equipment Lean in Manufacturing and Service Industries 6
S Determining a device category according to the MDR EU Medical Device Regulations 3