Hi everyone. This is my very first post on this platform, so excuse me if it's not the right section for my question. I'm currently doing my thesis in a company that produces surgical motor systems (not OEM). My task is to determine the the maximum number of reprocessing cycles of the reusable attachments of this motor system. I'd very much appreciate it if you'd give me some information on how to go about it. I'm not a complete beginner, so feel free to use technical terms.

In the IFU it's mentioned: "Frequent reprocessing has minor effects on the instruments (motor, attachments, accessories). The end of the device’s lifespan is normally determined by wear and tear, and damage resulting from usage."


This is confusing; although the IFU clearly states that frequent reprocessing has minor effects on the instruments, including the attachments, I'm still expected to come up with the maximum number of the reprocessing cycles. I assume the trick lies in the word "minor", i.e., the effects are there and my task is to know at which point these minor effects become significant in determining the the durability of the attachments (maximum reprocessing cycles). Is my understanding right? If so, how do I go about determining it?

Your help would be very much appreciated.
 
Elsmar Forum Sponsor

planB

Trusted Information Resource
#2
Assuming that you are talking about a re-usable medical-device accessory, ISO 17664:2017 applies, saying in section 6.3.2
6.3.2 If the service life of the medical device is limited by the number of processing cycles or some other end of life indicator(s) this information shall also be provided.
NOTE For example, this information if applicable, can provide a method to monitor the actual number of processing cycles.
The German ZLG provides more guidance on this topic in RDS007, section 8 (unfortunately, German only)

8 Festlegung des letztmöglichen Verwendungszeitpunktes

8.1 Grundlagen

Entsprechend Nr. 13.6 h) des Anhangs I der Richtlinie 93/42/EWG muss die Gebrauchsanweisung Angaben zu einer „eventuellen zahlenmäßigen Beschränkung der Wiederverwendungen“ enthalten. Grundlage hierfür sind die Anforderungen aus Anhang I Nr. 4 und Nr. 5 der Richtlinie (Erfüllung der Anforderungen bis zum letztmöglichen Verwendungszeitpunkt, dies umfasst auch die Biokompatibilität, wie auch in den Kapiteln 4.7 und 4.8 von ISO 10993-1: 2018 [7] vorgegeben). Bei für die Aufbereitung vorgesehenen Medizinprodukten muss grundsätzlich von Veränderungen ausgegangen werden, die ab einem bestimmten Zeitpunkt die Produktsicherheit beeinträchtigen können, wenn eines der folgenden Kriterien zutrifft; insbesondere wenn
• das Produkt aus mehreren Komponenten zusammengesetzt ist
• das Produkt mit beweglichen Teilen ausgestattet ist
• das Produkt kritische Elemente wie z. B. Kavitäten, Bohrungen, Blindbohrungen, Gelenke, Kugeldruckstücke, Federn enthält
• das Produkt aus Kunststoffen und/oder nicht aus Metallen mit zusammensetzungsbedingt zu erwartender langfristig hoher Korrosionsbeständigkeit besteht
• das Produkt mit poröser Oberflächen ausgestattet ist oder
• stärkere mechanische oder sonstige (z. B. chemische) Beanspruchung bei der Anwendung und ggf. des Aufbereitungsverfahrens auftreten können.

8.2 Nachweiserbringung

Ohne einer dieser Vorgehensweisen den Vorzug zu geben, können die erforderlichen Nachweise erbracht werden, durch z.B.
• Simulation des letztmöglichen Verwendungszeitpunktes und Vergleich mit einem nur einmal aufbereiteten Medizinprodukt
• dokumentierte Anwendungserfahrung (z.B. mit einem Vorgänger- oder ähnlichem Produkt)
• Literaturdaten
• Hierbei sind alle relevanten Aspekte wie Funktion, Aufbereitbarkeit, Anreicherung von Detergentienrückständen (Nichtbeeinträchtigung der Biokompatibilität) mit einzubeziehen und – sofern möglich - „worst case“–Parameter in Bezug auf die Aufbereitungsanweisung zu berücksichtigen. Bei Medizinprodukten mit stärkerer mechanischer oder sonstiger Beanspruchung (z.B. chemisch) bei der Anwendung sollten diese Aspekte zusätzlich berücksichtigt werden.
HTH,
 
#3
Hey, planB. Thank you very much for your quick and helpful reply. I'm doing my project in German, so luckily I don't have trouble with the language.

Yes – the attachments (handpieces), of which the maximum number of reprocessing cycles is to be determined, are reusable. I'm well familiar with the ISO 17664, but as far as I'm concerned, the IFU isn't being clear on this issue; it says (I quote the original, German text for accuracy): "Das Produkt kann 500-mal wiederaufbereitet werden. Dies wird durch Überprüfung aller technischen Akzeptanzkriterien nach 500 Aufbereitungszyklen belegt." the way I understand it is that 500 is not necessarily the limit or the maximum number of the reprocessing cycles, so the reusable attachments could possibly undergo more than 500 cycles and still be safe to use. That's how I personally understand it.

Thanks a lot for the ZLG link! Definitely helpful.
 
Thread starter Similar threads Forum Replies Date
S Determining sample sizes for PQ Qualification and Validation (including 21 CFR Part 11) 4
M Determining if an Insulin Pen Testing Machine is a Medical Device? EU Medical Device Regulations 4
M Determining a tolerance value for Measuring devices in-house inspection General Measurement Device and Calibration Topics 12
R Determining Uncertainty from Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
A IATF 16949 4.3.1 - Determining the scope of the quality management system - supplemental IATF 16949 - Automotive Quality Systems Standard 9
D ISO 9001:2015 4.3 Determining the Scope of the QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Determining SPC tolerance Statistical Analysis Tools, Techniques and SPC 21
B Determining SAT Offsets vs TUS Offsets per SAE AMS 2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
B Determining sample size for device sterility Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
D Determining of sample size for 'Operational Qualification' AQL - Acceptable Quality Level 5
R Question on determining defective units - I am not recording fixture to part rejected Statistical Analysis Tools, Techniques and SPC 5
S Clause 8.2.2 Determining the requirements for products and services ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
D Determining Calibration Frequency schedule for items used in production Manufacturing and Related Processes 2
C Determining if Maintenance Contractor is an External Service subject to ISO 9001 Clause 8.4 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 43
S AS9100D PEAR - Examples for organization's method for determining process results? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
I Determining Calibration Tolerance of a Measurement Device General Measurement Device and Calibration Topics 2
J ISO 17025 Documented Procedure for 6.2.5 - Determining competency ISO 17025 related Discussions 4
V Determining FDA 820 (registration) vs ISO 13485 - Supplier gives us the kit ISO 13485:2016 - Medical Device Quality Management Systems 1
J ISO 9001 8.4.1 - Determining controls applied to externally provided processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
E Determining what is good and what is bad can be subjective - when is it a quality issue? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
F Determining what type of scrap to include in my internal PPM calculation Quality Manager and Management Related Issues 5
M Determining number of employees within the "Scope" of the QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
A Determining the Scope of the QMS during Stage 1? Registrars and Notified Bodies 11
W Minor Audit Nonconformance Against Determining the scope of QMS IATF 16949 - Automotive Quality Systems Standard 12
D Determining Critical Components for conformity with IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 21
Q ISO 9001, section 4.3 Determining the scope of our QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
R Determining Sample Size for Medical Device Component Validation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
A What does 8.2.2.1, Determining the requirements related to products and services,mean IATF 16949 - Automotive Quality Systems Standard 1
A Determining Retention Period for Medical Device QMS documents Document Control Systems, Procedures, Forms and Templates 5
S Surveillance Sampling Test - Determining Sample Size Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
F ISO 9001:2015 4.3 - Determining the scope of the quality management system ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
B Standards Needed In House - Determining what standards are applicable Various Other Specifications, Standards, and related Requirements 3
W Determining Medical Device Classification in Mexico Other Medical Device Regulations World-Wide 5
K Determining Effect of Failure without a DFMEA (Design FMEA) FMEA and Control Plans 1
W Determining the Status and Importance of the Processes and Areas to be Audited Internal Auditing 7
T Determining Customer Requirements for the US Postal Service (USPS) IATF 16949 - Automotive Quality Systems Standard 4
Crimpshrine13 Rules of achieving and maintaining IATF recognition - Determining audit days IATF 16949 - Automotive Quality Systems Standard 2
R Developing procedure for Determining Company's Context And Interested Parties ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 53
W Determining Asset Life or Depreciation Life for M&TE General Measurement Device and Calibration Topics 1
N Reason for determining no adverse effect on reworked product ISO 13485:2016 - Medical Device Quality Management Systems 8
N Procedure for determining pinhole position in condom Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
alonFAI Determining PCBA Xray Test Sampling Size Statistical Analysis Tools, Techniques and SPC 1
S Determining sample size for inspection to achieve x% confidence re defects Misc. Quality Assurance and Business Systems Related Topics 10
S Determining Sample Size - AQL & LQ AQL - Acceptable Quality Level 10
R Determining the feasibility/scope for FDA IQ/OQ against a very simple application US Food and Drug Administration (FDA) 2
B Determining the "storage volume" (ISO 8536-8:2004 clause 6.14) Other Medical Device Related Standards 4
N Procedure/calculation for determining the foot pressure of dial thickness gauge General Measurement Device and Calibration Topics 10
B Bayes Success-Run on Attribute Data - Determining Sample Size Statistical Analysis Tools, Techniques and SPC 2
H Determining Takt Time in Metal Stamping Lean in Manufacturing and Service Industries 9
J Determining if there is Tin Plating on an Aluminum Bar Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6

Similar threads

Top Bottom