Hi All, I have a prototype Class 2 medical device (handheld battery-powered instrument for surgery). My team and I are trying to develop our own test methods and test equipment. I would greatly appreciate any and all advice/resources you can point me towards.
Custom Test Method = custom-made "implants" to allow us to measure post-operative results better in a CT Scan. We don't follow any FDA-recognized consensus standard. At most, we base our test methods on several scientific articles. I only read the FDA Guidance on Bench Testing and it wasn't helpful: https://www.fda.gov/media/113230/download. All it says is to provide "justification for the methods used to perform the testing" and "clinical/scientific/engineering basis for the acceptance criteria".
Custom Test Method = custom-made "implants" to allow us to measure post-operative results better in a CT Scan. We don't follow any FDA-recognized consensus standard. At most, we base our test methods on several scientific articles. I only read the FDA Guidance on Bench Testing and it wasn't helpful: https://www.fda.gov/media/113230/download. All it says is to provide "justification for the methods used to perform the testing" and "clinical/scientific/engineering basis for the acceptance criteria".
- Does our test method need to be 'validated' first? Do we define our own test protocol and test report specifically for validating our test method?
- If so, what sample size should we use for 'validation' of the test method? (Do we need run a test with 3 'test samples' (known dimensions) through the CT Scanner, and confirm that our measurements are accurate?)
- Should we have the CAD prints of this test jig be documented through our change control process?
- How should we document the production info? (Should it be part of our medical device files somehow?)
- Do we need to 'validate' this test jig first? Do we need to develop our own "IQ", "OQ", and "PQ" for this homemade test jig?
- If we need to 'validate' this test jig, then is a simple test with 3 'test samples' OK?