Dear all,

We are developing a "new" medical device which shares the majority of its design with another of our products, currently CE marked and marketed. That new product will be intended for the same clinical indication.

We are now at the stage of defining the proper route for its clinical evaluation and we think that the demonstration of equivalence might be challenged by our NB. Altough the principle of operation and performance requirements are similar, the new product will achieve them with a different intensity of energy and will consequently induce a biological response slightly different. In a nutshell, almost equivalent but additional clinical data may be required to assess its conformity.

My question is, would it be possible to benefit somehow from these similarities to justify (based on stats or not) a clinical trial sample size lower than what is usually calculated with classical statistical methods ? The idea is to tell our NB: yes it is equivalent, and to be even more confident, here some "extra" clinical data.

The only content I've found related to that is the use of bayesian statistics which seems far from obvious and quite disruptive.

Thank you for your help,

Louis L.-
 
Elsmar Forum Sponsor
#2
...the principle of operation and performance requirements are similar, the new product will achieve them with a different intensity of energy and will consequently induce a biological response slightly different.
Are you introducing a new clinical risk? It seems like you are. Can you derive the risk-benefit profile of the new device using the previous device's clinical data?
 

iVivien

Starting to get Involved
#3
Thanks for your reply.

No new clinical risks, but the current ones from the previous product are applicable. Therefore, the design mitigations have been reviewed and adapted to be implemented on the new product. Pre-clinical evaluations have been conducted to verify these mitigations.

The different pre-clinical results have highlighted some expected differencies between both products. Although we could argue and justify that these differencies do not raise raise new questions of safety and effectiveness, we think that adding new clinical performance data would streamline the review.
 

dgrainger

Trusted Information Resource
#6
the new product will achieve them with a different intensity of energy and will consequently induce a biological response slightly different.
You are saying that the Technical and Biological characteristics are different - It isn't equivalent unless....

"The characteristics listed in the first paragraph shall be similar to the extent that there would be no clinically
significant difference in the safety and clinical performance of the device. Considerations of equivalence shall be
based on proper scientific justification. It shall be clearly demonstrated that manufacturers have sufficient levels of
access to the data relating to devices with which they are claiming equivalence in order to justify their claims of
equivalence."
 
Thread starter Similar threads Forum Replies Date
G Device Registration Ammendments - Modifications to Medical Devices Other Medical Device Regulations World-Wide 4
P Approval process for modifications to a medical device with a CE Mark CE Marking (Conformité Européene) / CB Scheme 5
B Implications for CE Marking due to Device Modifications by the Purchaser CE Marking (Conformité Européene) / CB Scheme 1
E Tablet integrated into medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S New indication for a third party device US Medical Device Regulations 0
F Philippines: Old Certificate of exemption for Class B device and new LTO. How to proceed? Other Medical Device Regulations World-Wide 0
Ed Panek Is this a medical device? US Medical Device Regulations 2
D Are there Controlled Documents in a DMR (Device Master Record)? ISO 13485:2016 - Medical Device Quality Management Systems 3
P Medical device testing lab/service standard Other Medical Device Related Standards 1
K Medical Device file and "Component" Manufacturers? Other Medical Device and Orthopedic Related Topics 6
R CE MARKED DEVICE REGISTER SEPARATELY (LOCALLY) IN EU COUNTRIES? CE Marking (Conformité Européene) / CB Scheme 2
A Class 2b medical device - best pathway to obtain CE mark? (2022, MDR 2017/745) CE Marking (Conformité Européene) / CB Scheme 2
C Legacy device under Class A sterile devices EU Medical Device Regulations 0
C Statistics/sampling plan for medical device Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
T IVD Device Software - Risk Classification IEC 62304 - Medical Device Software Life Cycle Processes 11
J Can you assign two EMDN codes to a single device for EUDAMED? EU Medical Device Regulations 3
A Non-medical device testing in the medical system IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
H Is testing according to IEC 60601-1 & IEC 60601-1-2 mandatory for a device being registered under MDR? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S Training in D&D for engineers with no medical device experience Design and Development of Products and Processes 8
P Classification of Active non-implantable diagnostic device EU Medical Device Regulations 7
M Software as Medical Device import activities for Chile and Mexico Other Medical Device Regulations World-Wide 0
M Medical Device Indications for Use Health Canada Canada Medical Device Regulations 6
I Where to find MDR compliance device? EU Medical Device Regulations 6
F Can this device without tolerances be used as a predicate for our 510(k)? US Food and Drug Administration (FDA) 4
J Average number of Nonconformances during internal quality audit for Medical Device Manufacturers Internal Auditing 3
D Exclusion 4.2.3 Medical Device File ISO 13485:2016 - Medical Device Quality Management Systems 5
M Lifetime of medical device EU Medical Device Regulations 16
M UDI IVD Medical Device Class III US Food and Drug Administration (FDA) 1
E How to risk assess tooling? For a medical device and is it needed??? Manufacturing and Related Processes 2
P Definition of "Standalone" medical device Other Medical Device Regulations World-Wide 5
B Spare parts of medical device need CE mark? EU Medical Device Regulations 0
R EU MDR Article 22 applicability of medical device with components. EU Medical Device Regulations 4
M Medical Device AE Reporting SOP Medical Device and FDA Regulations and Standards News 0
M Class Exempt Medical Device Recall Question US Food and Drug Administration (FDA) 3
B Technical Documentation for a medical device accessory Medical Device and FDA Regulations and Standards News 6
U Change to the manufacturing Date for a device in the field US Medical Device Regulations 0
S Assembling technical file for a medical device ISO 13485:2016 - Medical Device Quality Management Systems 1
N FDA class 2 Device QS Requirements US Food and Drug Administration (FDA) 2
R Device with WiFi communication - EU MDR vs. EU RED EU Medical Device Regulations 2
M Proprietary name of a device US Medical Device Regulations 1
M Proprietary name of device US Food and Drug Administration (FDA) 3
K Class I Medical Device Registration EU Medical Device Regulations 11
M Drug-Device Combination product CE Marking (Conformité Européene) / CB Scheme 1
R DIGITAL MEASUREMENT DEVICE Manufacturing and Related Processes 2
S BASIC UDI-DI for a class IIa device marketed in a Procedure kit EU Medical Device Regulations 1
B Telemedicine as a Medical Device EU Medical Device Regulations 4
S Definition of disease as per medical device defintion EU Medical Device Regulations 14
B X-Ray Sensor Device Symbols for IFU Medical Device and FDA Regulations and Standards News 1
A How would a candidate for Quality Systems Manager role answer questions in a med device manufacturing org ISO 13485:2016 - Medical Device Quality Management Systems 11
Q Class I Medical Device with measuring function (UK MDR 2002 / MDD) UK Medical Device Regulations 3

Similar threads

Top Bottom