Device mounted at IV pole - what about mechanical stability testing?

#1
Hello everyone,

we are developing an infusion pump and are quite happy about our progress. But now it's time to catch up with the regulatory pieces. During the checking for the 60601-1 family we got some questions and would like to reach out if anyone has any experience or can just share their opinion on the topic.

The infusion pump itself must pass the 60601-1, which especially means 9.4 Instability hazards. We want to add a standard claw to attach the device to an IV pole, like many competitor products do. In these manuals there is only the note to attach the device firmly - no additional requirement to the pole other than it must still be stable.
Do we need to test the mechanial (in)stability of the IV pole when our device is attached to it? For obvious reasons we cannot test with every pole on the market and I haven't seen any "only use these poles: ..." up to now in any manual.
Or is it in the responsibility of the personal respectively the manufacturer of the pole that only attachments of up to e.g. 7 kg can be attached?
Or does 16.7 Protection against mechanical hazards for ME systems stand? But would we have to test every or any IV pole?

Thanks in advance
 
Elsmar Forum Sponsor

Peter Selvey

Staff member
Moderator
#2
This is excellent case study on why we need a "third way" on risk management: based on minimizing risk under a risk/resource model rather than a being based on acceptable risk.

When you have a system of devices that are placed on the market independently, the acceptable risk model doesn't give a reasonable (or at least, easy to see) answer as to how to deal with the various risks in the system. This particular case is a classic example - stability: who is responsible? Is it manufacturer A? B? C? Do they all have to assess the system risk? Do they all have to allocate system level risk control measures? Do we have to fix the system in order to be sure about the analysis, testing and results? If so, doesn't that mean a web of contracts between all the parties involved in the system? What happens when there is a design change? And that's just scratching the surface.

A resource based model looks at the system differently and asks: which manufacturer is in the best position (can use the least amount of resources) to minimize the risk? The answer in this particular example is most of the risk can be addressed by the pole manufacturer. They can design the pole to be stable with the typical loads, configurations expected in clinical practice, and also provide labelling on the maximum weights the pole is designed for.

But it's not a matter of other manufacturers closing their eyes: they still need to look at reasonable actions or solutions that can help, especially if there are low cost effective solutions. For example, the pump manufacturer can help by labelling their device clearly with the weight, making it easy for the user to decide if the pole can safely support the device, especially if there are already other devices on the pole.

I keep saying ... one day I'll publish the "minimum risk model" but work keeps getting in the way. The subtitle is planned to be "a mathematical model for common sense in risk management". It's a bit like how most people can catch a ball without knowing the underlying physics and math, but the nerds in society nevertheless want to know it anyway. This model is a nerd's attempts to fit some math behind some of ways we successfully handle risk in the real world, and it turns out that we not only don't need to estimate risk or establish any criteria for acceptable risk, but that the acceptable risk model itself is actually a pretty bad way to make decisions. The math isn't rocket science but it needs to be laid out slowly; one day!
 

Roland chung

Trusted Information Resource
#3
There has never been a very clear requirement as to the responsibility of the accessory manufacturers and the main unit manufacturers. Because there are so many possible scenarios, they can only be analysed on a case-by-case basis.

Sorry, I may be asking a different question here. Regarding the interconnection conditions defined in IEC 60601-2-18, is it the responsibility of the endoscope manufacturers or the manufacturers of the HF surgical accessories used under the endoscope to fulfill?

Merry Christmas!!
 
Thread starter Similar threads Forum Replies Date
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
E Procedure ( SOP) for Device Master Record ( DMR ) and for Device History Record (DHR)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
JoCam Certified QMS for MDR - Class I medical device manufacturers EU Medical Device Regulations 4
J US Device Listing when not the 510(k) owner Medical Device and FDA Regulations and Standards News 1
R Compatibility studies - Medicinal Product and Medical Device Other ISO and International Standards and European Regulations 0
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 2
G Voluntary Class I Device Filing Other US Medical Device Regulations 4
J Medical Device Regulations in Lebanon? Other Medical Device Regulations World-Wide 2
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
N Is this a single integral drug device combination product EU MDR CE Marking (Conformité Européene) / CB Scheme 1
S Medical Device MRI Compatibility EU Medical Device Regulations 3
A ISO 13485 for Class 1 Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 7
R Components to a finished medical device, MDR requirements Other US Medical Device Regulations 1
J Warnings/Cautions in Medical Device IFU Medical Device and FDA Regulations and Standards News 4
L Medical device HIPAA compliance in encryption Medical Information Technology, Medical Software and Health Informatics 1
N Validating Software before getting approved as Class 2 device US Food and Drug Administration (FDA) 5
M Clinical Benefit of device that only aids in a process for managing or treating disease EU Medical Device Regulations 2
M Office and manufacturing site relocation effect on Device Technical File CE Marking (Conformité Européene) / CB Scheme 2
D Can a diagnostic device be class I? EU Medical Device Regulations 1
M V&V phase: Justification of acceptance criteria (statistical method ) - (Medical Device) Design and Development of Products and Processes 2
supadrai Possible to Submit a 510(k) if you have never, do not and cannot produce the device for which you're submitting the application? US Food and Drug Administration (FDA) 4
S Pre-Submission for Breakthrough Device Designation - What's the threshold? US Food and Drug Administration (FDA) 8
M Surgical angle measurement guide device with an application software Medical Device and FDA Regulations and Standards News 1
E Medical Device - CE marking - Local market notifications EU Medical Device Regulations 1
S Medical Device Registration in Qatar Other Medical Device Regulations World-Wide 1
M Medical device substance based-leachables Other Medical Device Related Standards 2
R What information do i need to get from the device manufacturer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
P Anyone have an Idea on UAE Medical device registeration- Class B with FDA only Other Medical Device Regulations World-Wide 0
F Mobile app regulations - Class II medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
C Warehousing two separate components to a finished device while meeting regulations Other US Medical Device Regulations 7
M What are the basics of Medical Device Single Audit Program (MDSAP)? ISO 13485:2016 - Medical Device Quality Management Systems 7
D Legacy Device Review CE Marking (Conformité Européene) / CB Scheme 1
U Medical Device CE Marking - Using a disposable bearing CE Marking (Conformité Européene) / CB Scheme 3
L Medical Device Registration in Macau Other US Medical Device Regulations 1
M 'Active' device? Software/laptop with attached camera 'looking' at passive metal probe EU Medical Device Regulations 3
A 60601-1 understanding of stationary device IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Medical Device Registration in the Dominican Republic Other Medical Device Regulations World-Wide 4
Aymaneh Medical Device Cybersecurity Risk Management IEC 27001 - Information Security Management Systems (ISMS) 2
J Most versatile measurement device General Measurement Device and Calibration Topics 2
C Does an accessory need an IFU if it use is discussed in the Parent device IFU? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
T B2C Medical Device Shipping across the US Other Medical Device Related Standards 0
M Medical Device Registration In Malaysia Other Medical Device Regulations World-Wide 2
N Adding unclassified product to the medical device registration US Food and Drug Administration (FDA) 1
V Sister companies selling same medical device under different names ISO 13485:2016 - Medical Device Quality Management Systems 3
K CE Marking for Class I Medical Device? CE Marking (Conformité Européene) / CB Scheme 7
L Medical device storage conditions ISO 13485:2016 - Medical Device Quality Management Systems 1
F USB powered handheld medical device - Isolation requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L How to determine / validate Medical Device Storage Conditions ISO 13485:2016 - Medical Device Quality Management Systems 1
C Where to draw the line for "sufficient evidence" to verify safety/performance of a device? CE Marking (Conformité Européene) / CB Scheme 2
D Deciding whether or not pre-market clinical investigation is required for low risk device EU Medical Device Regulations 5

Similar threads

Top Bottom