Device mounted at IV pole - what about mechanical stability testing?

#1
Hello everyone,

we are developing an infusion pump and are quite happy about our progress. But now it's time to catch up with the regulatory pieces. During the checking for the 60601-1 family we got some questions and would like to reach out if anyone has any experience or can just share their opinion on the topic.

The infusion pump itself must pass the 60601-1, which especially means 9.4 Instability hazards. We want to add a standard claw to attach the device to an IV pole, like many competitor products do. In these manuals there is only the note to attach the device firmly - no additional requirement to the pole other than it must still be stable.
Do we need to test the mechanial (in)stability of the IV pole when our device is attached to it? For obvious reasons we cannot test with every pole on the market and I haven't seen any "only use these poles: ..." up to now in any manual.
Or is it in the responsibility of the personal respectively the manufacturer of the pole that only attachments of up to e.g. 7 kg can be attached?
Or does 16.7 Protection against mechanical hazards for ME systems stand? But would we have to test every or any IV pole?

Thanks in advance
 
Elsmar Forum Sponsor

Peter Selvey

Staff member
Super Moderator
#2
This is excellent case study on why we need a "third way" on risk management: based on minimizing risk under a risk/resource model rather than a being based on acceptable risk.

When you have a system of devices that are placed on the market independently, the acceptable risk model doesn't give a reasonable (or at least, easy to see) answer as to how to deal with the various risks in the system. This particular case is a classic example - stability: who is responsible? Is it manufacturer A? B? C? Do they all have to assess the system risk? Do they all have to allocate system level risk control measures? Do we have to fix the system in order to be sure about the analysis, testing and results? If so, doesn't that mean a web of contracts between all the parties involved in the system? What happens when there is a design change? And that's just scratching the surface.

A resource based model looks at the system differently and asks: which manufacturer is in the best position (can use the least amount of resources) to minimize the risk? The answer in this particular example is most of the risk can be addressed by the pole manufacturer. They can design the pole to be stable with the typical loads, configurations expected in clinical practice, and also provide labelling on the maximum weights the pole is designed for.

But it's not a matter of other manufacturers closing their eyes: they still need to look at reasonable actions or solutions that can help, especially if there are low cost effective solutions. For example, the pump manufacturer can help by labelling their device clearly with the weight, making it easy for the user to decide if the pole can safely support the device, especially if there are already other devices on the pole.

I keep saying ... one day I'll publish the "minimum risk model" but work keeps getting in the way. The subtitle is planned to be "a mathematical model for common sense in risk management". It's a bit like how most people can catch a ball without knowing the underlying physics and math, but the nerds in society nevertheless want to know it anyway. This model is a nerd's attempts to fit some math behind some of ways we successfully handle risk in the real world, and it turns out that we not only don't need to estimate risk or establish any criteria for acceptable risk, but that the acceptable risk model itself is actually a pretty bad way to make decisions. The math isn't rocket science but it needs to be laid out slowly; one day!
 

Roland chung

Trusted Information Resource
#3
There has never been a very clear requirement as to the responsibility of the accessory manufacturers and the main unit manufacturers. Because there are so many possible scenarios, they can only be analysed on a case-by-case basis.

Sorry, I may be asking a different question here. Regarding the interconnection conditions defined in IEC 60601-2-18, is it the responsibility of the endoscope manufacturers or the manufacturers of the HF surgical accessories used under the endoscope to fulfill?

Merry Christmas!!
 
Thread starter Similar threads Forum Replies Date
M Changing Investigational Device from Non-Sterile to Sterile - what are the implications? Other Medical Device and Orthopedic Related Topics 3
B FDA Breakthrough Device can be overlapped with a designated device? US Food and Drug Administration (FDA) 6
D Add new device to current license Canada Medical Device Regulations 0
S Obsolescence of the Medical Device in Various Countries Other Medical Device Related Standards 0
A MDR - Legacy Device Review Timeframe and Requirements EU Medical Device Regulations 3
A IT-NETWORK in PEMS Sub-Clause 14.13 for Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
G Medical Device - Borderline/Definition EU Medical Device Regulations 0
B Is labeling on the device itself required? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
A Intended Purpose on device label EU Medical Device Regulations 7
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
P European Medical Device Nomenclature (EMDN) and CND EU Medical Device Regulations 3
S Reseller Request to Change UPC on Medical Device via Re-labeling Medical Device and FDA Regulations and Standards News 4
D Hand Held medical device - power supply requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T CE Marking - Bench mark Device from same manufacturer CE Marking (Conformité Européene) / CB Scheme 2
E Medical device applicability to WEEE Other ISO and International Standards and European Regulations 2
Fjalar Spare parts for discontinued MDD compliant class I medical device EU Medical Device Regulations 4
H Medical Device Label Acceptance Criteria Manufacturing and Related Processes 4
J Calling a medical device a medical device (when it might not be one..) UK Medical Device Regulations 29
B Regulatory Affairs Certification (RAC) Book - Fundamentals of Medical Device Regulations Medical Device and FDA Regulations and Standards News 0
A How can we change a device trade name in 510(k) summary? US Food and Drug Administration (FDA) 0
N Medical device name in different countries EU Medical Device Regulations 4
M Issues with leakage current testing for a CLASS II device since no PE and FE is present IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
V Medical Device Literature Translation Software ISO 13485:2016 - Medical Device Quality Management Systems 1
M Are you allowed to reuse the same asset number if the device was replaced? General Measurement Device and Calibration Topics 10
G ISO 13485 Certification - Can we get the ISO 13485 certification prior to shipment of the device? ISO 13485:2016 - Medical Device Quality Management Systems 6
J Med Device Class for a Service Organization ISO 13485:2016 - Medical Device Quality Management Systems 2
Z Over The Air (OTA) updates for medical device Other US Medical Device Regulations 1
H Tukery Medical Device Regulstion Other Medical Device Regulations World-Wide 0
J UDI-DI how should we interpret Device version or model to determine if a new UDI-DI is needed? EU Medical Device Regulations 0
T Device & Accessory, Components in device, or System EU Medical Device Regulations 0
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 3
P UDI-PI requirements on reusable surgical device, do we need serialisation? ISO 13485:2016 - Medical Device Quality Management Systems 3
M Medical device certificate in Australia - ARTG certificate Other Medical Device Regulations World-Wide 0
Q Software as a medical device vs software not sold as medical device: local regulations for sale? EU Medical Device Regulations 4
H Medical device Product Registration Registrars and Notified Bodies 2
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K Technical file for MHRA - looking for assistance with self declaration Class I device CE Marking (Conformité Européene) / CB Scheme 7
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
M Sample record for verification performed by importers before placing a device on the market EU Medical Device Regulations 0
S IVD Device History Record ISO 13485:2016 - Medical Device Quality Management Systems 3
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
K How to handle GTINs for different configurations of one device with one part number? Other Medical Device Related Standards 0
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0

Similar threads

Top Bottom