Device Shelf Life Extension to 48 months without involving the FDA
Involved In Discussions
#1
Hi all,
I have small questing related to the shelf life of a device.
Reviewing current and draft guidance documents, its fairly clear that we can change the expiration date of the device by documenting the evidence internally.
Current Guidance: Deciding When to Submit a 510(k) for a Change to an Existing Device (1997)
B9: Change in packaging or expiration dating --> Yes --> Documentation
Draft Guidance: Deciding When to Submit a 510(k) for a Change to an Existing Device (2019)
B4: Is there a change in packaging or expiration dating? --> Yes --> B4.1: Is the same method or protocol, described in previous 510(k), used to support change --> Yes --> Documentation
Therefore, for example, if we gain 510(k) clearance for a device with a 12 month shelf life, if we had appropriate data we could expand the shelf life by e.g. 6 months by documenting this internally and not involving the FDA.
I'm wondering to what extremes this could be taken? Would it be possible to gain clearance for a device with a 1 month shelf life. Then expand the shelf life to 48 months without involving the FDA?
The reason i'm asking is that if we're gaining clearance on a device using accelerated aging data, the 510(k) review time and the shelf life study could be run in parallel and result in us getting the device to market a lot sooner. e.g.
  • gain a data for a short shelf life
  • Submit the 510(k)
  • Wait e.g. 200 days for the completion of the shelf life study and the clearance of the device
  • Document the change for the shelf life internally for the planned long shelf life
  • Launch the device
Thanks for your help!
Elsmar Forum Sponsor
JeantheBigone
Quite Involved in Discussions
#2
My gut feeling is "no."
First of all, change in shelf life / expiration date is a modification which according to FDA may require a new 510(k).
https://www.fda.gov/MedicalDevices/...sions/PremarketNotification510k/ucm134575.htm
With that, the answer to this question is firmly in the grey area.
However, FDA thinking is that there are reasonable shelf lives for different devices.
https://www.fda.gov/downloads/Drugs/DevelopmentApprovalProcess/SmallBusinessAssistance/UCM466475.pdf
So a 510(k) for a device that typically has a shelf life of five years but only claims a shelf life of one month would probably raise the reviewers' eyebrows. :rolleyes: I doubt it would get cleared without a discussion about the unusually short shelf life. FDA is not known for wanting manufacturers to take things to extremes. :)
Once the device is cleared, FDA leaves it to the manufacturer to decide whether a new 510(k) is needed or not. If you can justify and document that safety and effectiveness are not affected, then you can extend the shelf life. The risk you take is that FDA may disagree.
Thread starter Similar threads Forum Replies Date
C Medical Device Shelf life studies 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
shrutisancheti Predicate device accelerated aging - 1 year of shelf life is already over Other Medical Device Related Standards 1
H Frequency of shelf life/sterilization/bench studies conducted on a medical device Other Medical Device Related Standards 1
D Ethylene Oxide Sterile Implantable Medical Device Shelf Life Extension Other Medical Device Related Standards 1
N Sample Size Justification for Medical Device Shelf-Life Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
J EU Medical Device Class I - Stability (Shelf Life) Test EU Medical Device Regulations 2
J Non-Sterile Class 2 Medical Device Shelf Life Other Medical Device Related Standards 3
A Medical Device Shelf Life vs. Warranty Period Canada Medical Device Regulations 6
N Does Class I (exempt) device have requirements of Shelf Life and Expiration Dating ? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
U Shelf Life Requirement for Class II Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 26
S How to validate Device Shelf Life? CE Marking (Conformité Européene) / CB Scheme 8
Q Expiry Dates on a Non-Sterile Class 1 Medical Device (5 Year Shelf Life). Other Medical Device and Orthopedic Related Topics 11
E Reference Standard for Shelf Life Testing of Packaging for Sterilized Medical Device Other US Medical Device Regulations 3
M Shelf Life Testing of a Sterile Medical Device Other Medical Device and Orthopedic Related Topics 40
A Medical Device (Gloves) Stability (Aging) testing and Shelf Life Other Medical Device and Orthopedic Related Topics 15
L Shelf-Life vs. Lifetime of Medical Laser Device Other Medical Device and Orthopedic Related Topics 4
A Shelf Life for Products in Storage - Medical Device company ISO 13485:2016 - Medical Device Quality Management Systems 5
J Moisture sensitive device - New reel of a BGA with expired shelf life Other ISO and International Standards and European Regulations 5
J Shelf life control - Examples: solder paste, moisture sensitive device, epoxy Nonconformance and Corrective Action 3
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M OTS (Off-the-shelf) Medical Device Accessories - Camera EU Medical Device Regulations 1
S Off the Shelf Parts - Serial to USB adapter for a Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
shimonv Clinical evaluation report for class I device EU Medical Device Regulations 1
P UDI-PI requirements on reusable surgical device, do we need serialisation? ISO 13485:2016 - Medical Device Quality Management Systems 0
M Medical device certificate in Australia - ARTG certificate Other Medical Device Regulations World-Wide 0
Q Software as a medical device vs software not sold as medical device: local regulations for sale? EU Medical Device Regulations 4
H Medical device Product Registration Registrars and Notified Bodies 2
A Medical device labelling Date of manufacture US Food and Drug Administration (FDA) 2
W Non Sterile Medical Device Environmental Tests Other Medical Device Related Standards 4
A Clinical assessment sample size - Medical device Class IIb implantable (93/42 directive) EU Medical Device Regulations 2
K 25-year lifetime of medical device - document storage period EU Medical Device Regulations 1
M Sample record for verification performed by importers before placing a device on the market EU Medical Device Regulations 0
S IVD Device History Record ISO 13485:2016 - Medical Device Quality Management Systems 3
K Relabeling an existing medical device in the field? Other US Medical Device Regulations 6
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
A Reliable sources for following EU medical device regulatory EU Medical Device Regulations 0
T IVDR Medical device software CE Marking (Conformité Européene) / CB Scheme 8
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
K How to handle GTINs for different configurations of one device with one part number? Other Medical Device Related Standards 0
J Requirements as a Distributor for Incoming Inspection of Purchased Finished Medical Device Medical Device Related Regulations 0
S Microwave medical waste disinfectant - A medical device or not? Other ISO and International Standards and European Regulations 3
S Registration of Medical Device in Hong Kong - labeling requirements Other Medical Device Regulations World-Wide 0
V Software as medical device (SaMD) replicated for multiple clients through APIs IEC 62304 - Medical Device Software Life Cycle Processes 5
M Is the output of a device a Medical Device? IEC 62304 - Medical Device Software Life Cycle Processes 5
C Multiple owners for one listed device Medical Device and FDA Regulations and Standards News 1
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
J Promoting and marketing of a non approved device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B Russian Device to US US Food and Drug Administration (FDA) 5
L Medical device registration in Iran Other Medical Device Regulations World-Wide 0
S DHF/DMR/MDF for a software-only, cloud-based, single-instance device Medical Information Technology, Medical Software and Health Informatics 2
Similar threads
Top Bottom