DFMEA Severity 9-10 with Occurrence 3 and Detection 3 - Actions

Rendral

Registered
If you end up with a failure mode with Severity 9-10 with Occurrence 3 and Detection 3 what approach shall we have?

Would it be ok to state "no action" if we can not bring O & D down without design change ? We understand that a design change can solve the dilemma but if that's not the way out .....

How shall the FMEA manual be understood in this area? How much action is needed ? the goal is of course 10-1-1 but that is not always feasible.

Thanks in advance for your feedback on this topic
 

Chennaiite

Never-say-die
Trusted Information Resource
It is not unusual to have a Severity rating of 9/10 without much opportunities to mitigate Risk. Following could be done under the circumstance;

- Explain the Risk to the Top Management; Just to keep informed

- Indicate the relevant Characteristics as Critical/Safety/Legal/Etc. in the Drawing (Some may not agree to designate Special Characteristics, nevertheless, it remains a good message to Process guys)

- Share your DFMEA or outcome of your DFMEA to Process guys so that they will analyze the Causal factors in PFMEA too and enforce strong controls such as 100% check, Cpk controls, etc.
 

reynald

Quite Involved in Discussions
I will push for a better detection, to the point of receiving false alarms.
I rather have this than false security.
Sometimes an indirect measurement can be used as a detecting system (i.e. if this occurs, then this (my primary concern) will probably occur too). So I add that secondary detection to my current controls.
 
Thread starter Similar threads Forum Replies Date
M Inconsistency of VDA-AIAG severity table for DFMEA with the manual text FMEA and Control Plans 0
R DFMEA/PFMEA mitigation of high severity (9-10) in low volume products IATF 16949 - Automotive Quality Systems Standard 1
M Severity Reduction in a Design FMEA (DFMEA) FMEA and Control Plans 26
H DFMEA Severity Ranking of Non-Compliance to EMC Standards FMEA and Control Plans 2
C Does the PFMEA Severity have to match the DFMEA Severity? FMEA and Control Plans 3
J DFMEA SOD Rating Scale - Severity, Occurance, and Detection FMEA and Control Plans 1
D Redundant Component DFMEA Severity & Occurrance FMEA and Control Plans 2
Ron Rompen Development of the PFMEA - SEVERITY scores not derived from the DFMEA - Ford FMEA and Control Plans 13
O Process Flow Diagram, DFMEA, PFMEA, Control Plan integrated softwares FMEA and Control Plans 7
Y DFMEA Sample in Semiconductor FMEA and Control Plans 2
M AIAG-VDA DFMEA 2019 - it is compulsory? FMEA and Control Plans 7
T PFMEA vs MFMEA or DFMEA on equipment/machinery FMEA and Control Plans 19
M Should there be another column in the Optimization section (step 6) of AIAG-VDA DFMEA form? FMEA and Control Plans 1
M How to record the effects of adjacent systems (shown in boundary diagram) in the AIAG-VDA DFMEA form FMEA and Control Plans 0
OpExPro The note at the bottom of DFMEA occurrence tables in AIAG VDA FMEA FMEA and Control Plans 0
OpExPro AIAG VDA DFMEA Template Required FMEA and Control Plans 6
P DFMEA - Machinery Design Best Practices FMEA and Control Plans 0
P Equipment URS, pFMEA and dFMEA ISO 14971 - Medical Device Risk Management 3
J Looking for a training organization that can provide in house DFMEA training FMEA and Control Plans 4
P Understanding DFMEA and PFMEA - Supplier Related IATF 16949 - Automotive Quality Systems Standard 21
B Why the Greek god Hephaestus should have done a design FMEA (DFMEA) on his giant robot APQP and PPAP 1
DuncanGibbons Best practice for identifying "items" of parts for DFMEA analysis AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
Q Is that any difficulty to do software DFMEA and PFMEA in ISO 13485? ISO 13485:2016 - Medical Device Quality Management Systems 5
M How to complete structure Analysis and Function analysis sections' columns of AIAG-VDA DFMEA form FMEA and Control Plans 0
Jimmy123 Handling nonfunctional requirements in DFMEA FMEA and Control Plans 5
Jimmy123 P-Diagram vs DFMEA FMEA and Control Plans 4
C Function definition for DFMEA FMEA and Control Plans 1
G DFMEA buy-in strategies for Design Engineers APQP and PPAP 22
optomist1 DFMEA - Real Life Drama FMEA and Control Plans 25
optomist1 Share a DFMEA template please Document Control Systems, Procedures, Forms and Templates 2
P Detection rating for a user (surgeon) related failure mode in DFMEA APQP and PPAP 3
Marc Definition DFMEA (Design FMEA) - Definition Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 0
S DFMEA when the supplier doesn't have design responsibility APQP and PPAP 15
T DFMEA/PFMEA and SCCAF alignment FMEA and Control Plans 19
C Design for Assembly in DFMEA - Failure Effect of Sub-System(s) FMEA and Control Plans 5
C Resource allocation problems in Software DFMEA FMEA and Control Plans 4
I Making a process PFMEA, no DFMEA (non-automotive toll manufacturer) FMEA and Control Plans 1
Proud Liberal DFMEA / PFMEA linkage - Where in the standards do I find it? FMEA and Control Plans 5
D Introducing an existing product into automotive context - DFMEA needed? IATF 16949 - Automotive Quality Systems Standard 8
N DFMEA, Drawings and PFMEA of an Assembly Process FMEA and Control Plans 12
N YS Class in DFMEA & Process Control Measures FMEA and Control Plans 1
G Is it compulsory to make DFMEA for proto parts making organisation APQP and PPAP 1
K Determining Effect of Failure without a DFMEA (Design FMEA) FMEA and Control Plans 1
Stijloor DFMEA & DVP&R Training Providers FMEA and Control Plans 1
P Does anyone have a DFMEA process Interface Matrix .xls template FMEA and Control Plans 7
C DFMEA vs. PFMEA Relationship: The cause in DFMEA becomes Failure in the PFMEA FMEA and Control Plans 2
G How to identify Key Characteristics (KC) in a Design FMEA (DFMEA) FMEA and Control Plans 2
W 'Classification' column in DFMEA FMEA and Control Plans 3
N What is the relationship between a DFMEA and PFMEA? FMEA and Control Plans 1
S DFMEA countermeasures in PFMEA FMEA and Control Plans 18

Similar threads

Top Bottom