Hello Everyone!
What is the difference between ME Equipment and ME system?
IEC 60601-1, 3.1 edition – Terms and Definitions
* MEDICAL ELECTRICAL SYSTEM
ME SYSTEM
combination, as specified by its MANUFACTURER, of items of equipment, at least one of which is ME EQUIPMENT to be inter-connected by FUNCTIONAL CONNECTION or by use of a MULTIPLE SOCKET-OUTLET
3.33
* FUNCTIONAL CONNECTION
connection, electrical or otherwise, including those intended to transfer signals, data, power or substances
Examples for ME equipment’s
- High Frequency Surgical Equipment.
- Cardiac Defibrillators.
- Patient Monitors.
- Therapeutic and Diagnostic Ultrasound Equipment.
- Medical Lasers.
- Patient Ventilators.
- Infant Incubators and Warmers
Particular standard takes priority over the general standard IEC 60601-1. In that regard wrt IEC 60601-2-1, states that the requirements of clause 16 of the general standard do not apply.
Requirements of clause 16 of 60601-1 are not applicable to electron accelerators in the scope of 60601-2-1?
Medical LINAC does not falls under ME System?
Can anyone help in understand the difference between ME Equipment and ME System