Differences between BS EN IEC 60601-1:1990 and 2007 standards
U
uksmiler
I wonder if anyone can help me on this? Seems like the most appropriate place to ask....
Ok, so if a device has passed the BS EN 60601 - 1 : 1990 electrical safety test standard, is there any reason why this same device would not pass the BS EN 60601 -1 : 2007 standard? Were there any significant changes to the safety test requirements during these years? Or have I missed something more obvious?
Any detail of any changes would be greatly appreciated.
Stijloor
Welcome to The Cove Forums! :bigwave: :bigwave:
OK, I am NOT an electrical/medical devices guy....:nope:
But, could this be of help?
Stijloor.
Marcelo
A device complying to the former edition of EN 60601-1 does not comply to the new edition.
There are two major new modifications in the third edition (IEC 60601-1:2005, the new EN standard is based on this edition). and some new requirements.
The first major modification is the requirement of a ISO 14971 compliant risk management system to claim compliance with IEC 60601-1. More that just a requirement to have a risk management system, there´s the need to have a risk management system integrated with the requirements of IEC 60601-1 and other safety standards (meaning that the requirements of IEC 60601-1 - and other safety standards - must be used as risk control measures in the risk management system). This alone would mean that a device complying with the former edition would not comply with the new.
The second major change is the introdution of essential perfornace. As essential performance is a new conept and determined thru risk management, this also means that devices compying with the second edition does not address this.
There´s also a LOT of new requirements (the standards almost doubled in size) such as vibration, noise, and new and modified mechanical requirements.
Finally, a lot of requirements now have to be determined thur risk management (for example, the test conditions or the test limists have to be determined thur risk management) meaning that you can´t even PRFORM the tests if your risk management does not address the test.
Scottep
I would like to post a correction to the information provided on ISO 14971 being required. Conformance to ISO 14971 is not required by IEC 60601-1 3rd edition, it is one of two options to choose from regarding how you deal with risk management.
IEC 60601-1 3rd Edition does allow for a "question based approach" to risk assessment. In this case, you have the option of modifying your risk plan to meet the risk requirements set forth in the 3rd Edition. There are two sides to this decision.
One: Using the question based approach will save money up front. Allowing you to meet the risk requirements of the 3rd Edition without having to spend the money to get certified to ISO 14971
Two: If you do go ahead with certification to ISO 14971, you automatically meeting the risk assessment requirements of the 3rd edition. This would go for any product you develop under the 14971 system that you would have in place. Opposed to performing an evaluation of the risk assessment plan for each product.
Marcelo
Hello Scottep, and welcome to the Cove!
I did not say that conformance with ISO 14971 is required (not implying that it needs a audit for exemple)... i just said that a "ISO 14971 compliant risk management system" is required, as is written in clause 4.1 of the third edition.....these are two different things :)
Anyway, the standard is being ammended to clarify what is really required (a risk management process complying with ISO 14971 but not the whole process, for example, it does not require production and post-production information).
I´ve never heard about this, where does this come from?
This is really not required by the third edition, and it´s one of the reasons that a clarification in the amendment is needed.
Not really.
As IEC 60601 is for type testing, you need to evaluate it for each device. So you would need to evaluate the components of the ISO 14971 process required by the third edition for each device. For example, you need to evaluate if the risks presented by the product are acceptable. This acceptability is seen by comparing the results with the risk acdeptability criteria, which is device-specific, an not related to the system as a whole.
Martin IT
Another difference between 2nd edition and 3rd edition is the introduction new paragraphs due to absorption by the 3rd edition of some side rules, including:
IEC 60601-1-4 standard on programmable devices (para. 14)
IEC 60601-1-1 standard on medical systems (para. 16)
Rule of usability IEC 60601-1-6 (para. 12)
IEC 60601-1-8 standard for alarms (para. 12)
This implies being aligned with a larger number of constraints and the need to
review the analysis of the electrical safety of the product.
In any case I would put emphasis on paragraph 4 and the need to perform a process of risk management according to ISO 14971. This is, in my opinion, the most delicate step of the 3rd edition.
Scottep
Mmantunes, I misunderstood what you were saying about ISO 14971. I took what you said as saying that someone needs to be compliant with 14971, my apologies.
As far as the "question based approach" comment that I made. This is wording that I have heard from a number of different safety consultants (UL, Intertek, etc). Basically, you just meet the risk requirements in the 3rd Edition, opposed to having an all inclusive ISO 14971 certified risk management program, which would be the "audit based approach."
I didn't said that you had to be ISO 14971 certified. When you say:
I 100% agree with you, it is not required.
Would you agree that being certified to ISO 14971 would ultimately make it easier for a company to consistently meet 3rd Edition risk management requirements for the products that they develop?
E Differences between IEC 60601-1-11:2015 and 2010 Other US Medical Device Regulations 0
L Differences in the requirements between IEC 60601-1 3rd ed. and EN 60601-1:2006 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N R&R for Differences between 2 measurements Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
DuncanGibbons Clear differences between ISO 13485 and AS 9100D requirements ISO 13485:2016 - Medical Device Quality Management Systems 10
K Differences between Manufacturing Process Audit & Layered Process Audit Lean in Manufacturing and Service Industries 4
S Differences in CB expectations between the US and Europe Registrars and Notified Bodies 3
M What are the differences between "Process Flowchart form" and "Process Flow Diagram form"? IATF 16949 - Automotive Quality Systems Standard 4
K Differences between EN ISO 11607-1:2009 vs. 2017 Other Medical Device Related Standards 1
M What are the differences between TIR 13004:2013 and TIR 33? Other Medical Device Related Standards 2
H Differences between CQI 15 and CQI 17 IATF 16949 - Automotive Quality Systems Standard 0
A ISO 13485:2003 vs ISO 13485:2016 - What are the differences between the two? ISO 13485:2016 - Medical Device Quality Management Systems 1
C Differences between IATF 16949 and AS9100 IATF 16949 - Automotive Quality Systems Standard 13
T PPAP differences between TS16949 and VDA APQP and PPAP 1
J Differences between a Risk Management Plan vs. Production Part Approval Process AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
Y Differences between the Classic 7 QC Tools and New QC Tools Quality Tools, Improvement and Analysis 3
C Differences between IATF 16949:2016 vs ISO/TS 16949:2009 IATF 16949 - Automotive Quality Systems Standard 13
Q What are the differences between ISO 13485 vs. AS9100C ISO 13485:2016 - Medical Device Quality Management Systems 3
L AQL ANSI z1.4 & z1.9 - What are the differences between these 2 standards? AQL - Acceptable Quality Level 5
V Differences between ISO 13485 & FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
Q Differences between Clauses 7.5, 8.2.3 and 8.2.4 of ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
V Statistics - Designs - Differences between Orthogonality and Confounding Using Minitab Software 5
K Differences between ISO 9001 and API Q1. Other ISO and International Standards and European Regulations 1
M Differences between international and regional standards (EN, AAMI, CSA...) Other Medical Device Related Standards 6
C Control of Records - Differences between Controlled Documents and Controlled Records Records and Data - Quality, Legal and Other Evidence 18
S What are the differences between Traditional 510(k) and a PMA to meet 21 CFR 814.20 ? Other US Medical Device Regulations 3
O Differences between EN 61326-1:2006 and EN 61326-1:2013 CE Marking (Conformité Européene) / CB Scheme 2
K The differences between ISO 9001 vs a QMS (Quality Management System) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Y Differences between Top Management, Management & Organization ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
W Differences between VDA 6.3 and ISO 9001 VDA Standards - Germany's Automotive Standards 2
G Differences between Quality Objectives and Business KPI's ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
P Differences between an NCR and CAR Internal Auditing 9
M Differences between High, Medium, Low Risk Suppliers Supplier Quality Assurance and other Supplier Issues 5
G Differences between Design, Development, and Production Misc. Quality Assurance and Business Systems Related Topics 4
C Differences between ISO 9001:2008 and AS9100 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
S Differences between Conductive vs. Dissipative ESD Flooring Manufacturing and Related Processes 3
D Quality Management Systems - Differences between ISO 13485, 21 CFR Part 820 & ICH Q10 Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
D Differences between (c)GMP for Drugs and Medical Devices Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
A EN 60601-2-51 - Differences between the 2003 and the 2005 edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C What are the differences between Medical and Non Medical Grade Adhesives 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
V Differences between Business Process Reengineering (BPR) and Total Quality Management Quality Manager and Management Related Issues 3
V Differences between Quality Tools and Quality Concepts? Quality Tools, Improvement and Analysis 4
G What are the differences between US and EU Pharmacopeias? Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 11
N Differences between a Safety Nonconformance and a Quality Nonconformance Nonconformance and Corrective Action 14
F Differences between ISO 9001 and ISO 9004 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
A Defining the differences between Prototype vs. Production Guidelines? Contract Review Process 5
S Differences between the SAE AS9100C and BS EN AS 9100:2009 versions of the standard AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
V Similarities and Differences between EU & US GMPs (Good Manufacturng Processes) Document Control Systems, Procedures, Forms and Templates 8
J Definition Servicing circuit boards - Differences between Rework and Repair Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 9
M Differences between ISO 9004:2000, Baldrige National Quality Progr ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Q Differences Between TS16949 and 13485 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1