I've finally had a chance to compare the two to see what exactly has changed.

The one thing that I found in Management Requirements is the addition of "Improvement", section 4.10 which used to be "Corrective Actions".

Wouldn't it have made more sense to add Improvement as section 4.15? It's only a three sentence paragraph for continual improvement. Instead of making ones work easier we now have to change sections 4.10 - 4.15:confused:

Have you seen anything that has left you scratching your head wondering why?
 
Elsmar Forum Sponsor

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#2
I disagree.....however, I do think that it would have been better placed under Preventive Action, Clause 4.11, as continuous improvement is what preventive action tires to achieve. Creating a whole new Clause would have - in my opinion - been an error.

It should also be included as part of the preventive action provisions required during management review, Clause 4.14.1.....

Hershal
 
K

Ken K

#3
I must also disagree Hershal since the clause takes more into account than just PA. Continual improvement is echoed, IMO, too often through the changes that have been added to 2005. Clause 4.10 should have been adequate.


4.10 Improvement

The laboratory shall continually improve the effectiveness of its management system through the use of the quality policy, quality objectives, audit results, analysis of data, corrective and preventive actions and management review.


Clause 4.11 is now corrective action
Clause 4.12 is now preventive action
Clause 4.13 is now control of records
Clause 4.14 is now internal audits
Clause 4.15 is now management review

My point is that improvement should have been designated 4.15 instead of shifting each clause up.
 
G

Graeme

#4
There's the right way, the wrong way and the Marine Corps way ...

Ken K said:
My point is that improvement should have been designated 4.15 instead of shifting each clause up.
Ah, Ken, now you're trying to apply rational thinking and logical planning to a bureaucratic multinational and multi-organizational working group of volunteers ... that's how we all may have done it, but with "them" I think all bets are off. :D
 
K

Ken K

#6
Isn't that the truth Graeme:bonk:

I'm not sure if our auditor worked on this revision or not, but you can bet there will be some discussion concerning the changes with her come February.

I'm sort of curious on finding out how their minds work...but that feeling shall also pass:lol:
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#7
Ken K said:
I'm not sure if our auditor worked on this revision or not, but you can bet there will be some discussion concerning the changes with her come February.

I'm sort of curious on finding out how their minds work...but that feeling shall also pass:lol:
Each AB has their own policy, although APLAC/ILAC have an implementation cycle, I am not sure how long it is though.

By "their" do you mean the accreditation assessors? If so, then certainly talk to your assessor and find out the AB's policy/approach, and timetable.

Hershal
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#8
I have just been doing a bit more reading on the 2005 revision of 17025.....besides 4.10 (and the subsequent shifting up of other section four clauses), there is also changes in clauses 4.1.5, specifically the sub-clause j, and in 4.2.3 through 4.2.7, and boosting a note to a requirement in 4.7.2

Accredited laboratories should discuss with their accrediting body what the changes mean to them.

Hershal
 

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#10
Thanks Sidney,

That provides a good thumnail for the differences.

There are actually two hard requirements in the 2005 revision.....4.7.2 and 5.9.2.

Clause 4.7.2 takes a note from the 1999 revision and elevates it to an actual requirement. For this particular clause this is the right move. Labs (both test and cal) are now REQUIRED to seek customer feedback, both positive and negative, analyze it to figure out how to become better, and then actually put steps in place to become better, where possible.

Clause 5.9.2 requires the lab to take information obtained from Assurance of Quality steps that the lab has taken, then analyze that information and assure that any issues don't recur. For labs accredited by ILAC/APLAC/EA/IAAC, they are required to participate in proficiency testing or inter-laboratory comparison (PT/ILC) if it exists. In the case where there is PT/ILC then an "outlier" as it is known will require the lab to employ its corrective and preventive action system to investigate, resolve, and prevent recurrence of the outlier. An outlier means that the results of the information plus the measurement uncertainty is outside the limits of the reference norm.

The other clauses that change (e.g. 4.10, 4.1.5.k) are taken primarily to lign 17025 with 9K. These clauses may - at least in short term - create issues with both implementation and assessment. The reason is that they involve a substantial degree of subjectivity Time will tell how these clause shake out.

Hershal
 
Thread starter Similar threads Forum Replies Date
N R&R for Differences between 2 measurements Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
A VDA2 - 6th edition June 2020 | Differences for the previous VDA2 VDA Standards - Germany's Automotive Standards 0
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S SAE AS13004 vs. SAE J1739 - What are the main differences? FMEA and Control Plans 0
M Differences in post market safety reporting for Combination Product Applicants Medical Device and FDA Regulations and Standards News 1
A MSA results differences - Supplier results vs. My results Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
DuncanGibbons Clear differences between ISO 13485 and AS 9100D requirements ISO 13485:2016 - Medical Device Quality Management Systems 10
K Differences between Manufacturing Process Audit & Layered Process Audit Lean in Manufacturing and Service Industries 4
S Differences in CB expectations between the US and Europe Registrars and Notified Bodies 3
M Authorized Rep (AR), competent authority, NCA and notified body - Differences EU Medical Device Regulations 9
D USFDA vs NRTL/IEC 17025 Differences - Compliance testing lab ISO 17025 related Discussions 0
D USFDA vs NRTL/IEC 17025 Differences - Compliance testing lab US Food and Drug Administration (FDA) 6
P HVAC System vs Air Handling Unit (AHU) - Differences Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
G Differences - Nonconformances vs CAPA, Corrections vs Corrective Actions ISO 13485:2016 - Medical Device Quality Management Systems 16
M What are the differences between "Process Flowchart form" and "Process Flow Diagram form"? IATF 16949 - Automotive Quality Systems Standard 4
K Differences between EN ISO 11607-1:2009 vs. 2017 Other Medical Device Related Standards 1
M What are the differences between TIR 13004:2013 and TIR 33? Other Medical Device Related Standards 2
I "Method" vs "Process" - Differences Document Control Systems, Procedures, Forms and Templates 33
S The key differences in 510k types - Traditional vs. Abbreviated vs. Special US Food and Drug Administration (FDA) 0
M Differences in CQE, CSTE, CSQA, CQA Professional Certifications and Degrees 3
H Differences between CQI 15 and CQI 17 IATF 16949 - Automotive Quality Systems Standard 0
L GMDN code differences, Manufacturer and Sponsor Other Medical Device Regulations World-Wide 3
F UMDN vs GMDN - What are the differences? EU Medical Device Regulations 1
D AS9100D vs AS9120B differences AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
A ISO 13485:2003 vs ISO 13485:2016 - What are the differences between the two? ISO 13485:2016 - Medical Device Quality Management Systems 1
F EU IVD registrations - CFG or CFS - Differences EU Medical Device Regulations 1
C Differences between IATF 16949 and AS9100 IATF 16949 - Automotive Quality Systems Standard 13
P Differences CSQA vs CQA ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 1
T PPAP differences between TS16949 and VDA APQP and PPAP 1
E Differences between IEC 60601-1-11:2015 and 2010 Other US Medical Device Regulations 0
J Differences between a Risk Management Plan vs. Production Part Approval Process AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
Y Differences between the Classic 7 QC Tools and New QC Tools Quality Tools, Improvement and Analysis 3
M Standard work, standardized work instruction & work instruction: What are differences IATF 16949 - Automotive Quality Systems Standard 7
M Barrier Free vs. Wheelchair Accessible - Differences Occupational Health & Safety Management Standards 1
F IEC 60601-1-2: Differences 2007 and 2015 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R 510(k) Executive Summary vs 510(k) Summary - Differences 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
C Differences between IATF 16949:2016 vs ISO/TS 16949:2009 IATF 16949 - Automotive Quality Systems Standard 13
A ISO 13485:2003 vs. ISO 13485:2016 Differences Comparison ISO 13485:2016 - Medical Device Quality Management Systems 4
L Differences in the requirements between IEC 60601-1 3rd ed. and EN 60601-1:2006 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
V Application and differences of OC Curve and Acceptance Sampling Statistical Analysis Tools, Techniques and SPC 2
Q What are the differences between ISO 13485 vs. AS9100C ISO 13485:2016 - Medical Device Quality Management Systems 3
L AQL ANSI z1.4 & z1.9 - What are the differences between these 2 standards? AQL - Acceptable Quality Level 5
V Differences between ISO 13485 & FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
J EMI/EMC Test differences for Radiography Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Q Differences between Clauses 7.5, 8.2.3 and 8.2.4 of ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
V Statistics - Designs - Differences between Orthogonality and Confounding Using Minitab Software 5
K Differences between ISO 9001 and API Q1. Other ISO and International Standards and European Regulations 1
Q RUO (Research Use Only) vs. IVD (In Vitro Diagnostic) - Differences EU Medical Device Regulations 6
M Safe Work Practice vs Safe Work Procedures - Differences Occupational Health & Safety Management Standards 6
M Differences between international and regional standards (EN, AAMI, CSA...) Other Medical Device Related Standards 6

Similar threads

Top Bottom