Differences - Nonconformances vs CAPA, Corrections vs Corrective Actions

G

gporter

I'd appreciate people's thoughts on the following regarding how/what system our medical device company uses for NMRs, CAPAs, corrections, corrective/preventive actions...

As the head of Quality for my company, I hold regular meetings with influential members of my organization to discuss Quality System philosophies. A belief of mine is that all areas of our Quality System should drive disparate activity and shouldn't have overlapping goals. For example, if a product complaint investigation identifies a non-conformance with the complaint product, then a separate nonconforming report (NMR) should be opened to address it in addition to the complaint to address the field feedback. We shouldn't address the non-conformance through the complaint process - that's what the NMR process is for. These well-defined systems help eliminate confusion.

One conversation that has been a hot topic recently is the differences between requirements for nonconforming material/event reports and CAPAs. My team and I get caught up in the discussion if it is allowable to perform corrective actions in an NMR. My stance is that all corrective actions belong in our CAPA system, as opposed to our NMR system. A proper corrective action requires root cause determination and verification of effectiveness (VoE) of the corrective action. Our NMR process does not include VoE. It does have a provision for assignable cause, but not root cause. For the same reason, I do not require root cause in our NMR system because corrective actions are not required. Corrections are used in the NMR system, corrective (and preventive) actions are in our CAPA system.

NMRs require:
  1. Non-conformance identification
  2. risk assessment (which is requires escalation to CAPA if risk is high enough)
  3. investigation and assignable cause
  4. interim actions (if necessary)
  5. disposition
  6. closure
CAPAs require:
  1. Non-conformance identification
  2. investigation
  3. root cause
  4. corrective/preventive action
  5. VoE
My goal is to not have the ability to chose between using a NMR or CAPA for corrective actions by eliminating grey areas so people are clear on what system would be required. It is possible for "corrective action"-type actions to be taken in an NMR's Interim Actions (update a specification, for example), but without a formal root cause investigation or VoE activity, I hesitate to allow formal corrective actions in the NMR process. If we performed corrective actions in an NMR, couldn't an auditor say "I see you took corrective actions here, don't you have a CAPA process?" or "What was the identified root cause in this case?" or "How did you verify the effectiveness of this corrective action?" I feel in some ways it boils down to semantics, but the words can matter...

What are people's thoughts on this topic? Can "corrective actions" live in both NMRs and CAPAs? Thanks for the input.
 

Sidney Vianna

Post Responsibly
Leader
Admin
Sorry for being a "purist", but it is critical to keep terminology in check. Corrections, such as "update a specification" is not the same as corrective actions.

Actually, in the Medical Device sector, the term CAPA seems to be misaligned with most of the rest of the world. The PA in CAPA seems to refer to prevention of recurrence of a problem and that, in the ISO 9000 definition is deemed to be corrective action (not preventive action).

It seems to me that you are really asking if we need (full blown) corrective actions with root cause analysis for each and every nonconformity and the answer is a resounding NO. We have discussed this many times and maybe someone can point to similar threads. There are many cases where system changes (the essence of corrective actions) are not justified based on severity, criticality and/or frequency of the problem.
 

Jim Wynne

Leader
Admin
This all seems fraught with overcomplication and too much thinking. In your list of the things that NMRs require, #2 says that CAPA is necessary in some high-risk cases. Is there any reason that high-risk NMR findings shouldn't just be handed off to the CAPA process? Is there any good reason for tightly circumscribing the processes? Is the tail wagging the dog?
 

Pancho

wikineer
Super Moderator
Yes, CAPA is a term which stems from an incorrect use of the terms from the FDA, in fact CAPA (corrective action then preventive action) does not exist in practice.

Such a common misconception. I bet it even played a part in getting rid of preventive action in ISO 9001:2015.
 

Ronen E

Problem Solver
Moderator
I'd appreciate people's thoughts on the following regarding how/what system our medical device company uses for NMRs, CAPAs, corrections, corrective/preventive actions...

As the head of Quality for my company, I hold regular meetings with influential members of my organization to discuss Quality System philosophies. A belief of mine is that all areas of our Quality System should drive disparate activity and shouldn't have overlapping goals. For example, if a product complaint investigation identifies a non-conformance with the complaint product, then a separate nonconforming report (NMR) should be opened to address it in addition to the complaint to address the field feedback. We shouldn't address the non-conformance through the complaint process - that's what the NMR process is for. These well-defined systems help eliminate confusion.

One conversation that has been a hot topic recently is the differences between requirements for nonconforming material/event reports and CAPAs. My team and I get caught up in the discussion if it is allowable to perform corrective actions in an NMR. My stance is that all corrective actions belong in our CAPA system, as opposed to our NMR system. A proper corrective action requires root cause determination and verification of effectiveness (VoE) of the corrective action. Our NMR process does not include VoE. It does have a provision for assignable cause, but not root cause. For the same reason, I do not require root cause in our NMR system because corrective actions are not required. Corrections are used in the NMR system, corrective (and preventive) actions are in our CAPA system.

NMRs require:
  1. Non-conformance identification
  2. risk assessment (which is requires escalation to CAPA if risk is high enough)
  3. investigation and assignable cause
  4. interim actions (if necessary)
  5. disposition
  6. closure
CAPAs require:
  1. Non-conformance identification
  2. investigation
  3. root cause
  4. corrective/preventive action
  5. VoE
My goal is to not have the ability to chose between using a NMR or CAPA for corrective actions by eliminating grey areas so people are clear on what system would be required. It is possible for "corrective action"-type actions to be taken in an NMR's Interim Actions (update a specification, for example), but without a formal root cause investigation or VoE activity, I hesitate to allow formal corrective actions in the NMR process. If we performed corrective actions in an NMR, couldn't an auditor say "I see you took corrective actions here, don't you have a CAPA process?" or "What was the identified root cause in this case?" or "How did you verify the effectiveness of this corrective action?" I feel in some ways it boils down to semantics, but the words can matter...

What are people's thoughts on this topic? Can "corrective actions" live in both NMRs and CAPAs? Thanks for the input.
I think that your thinking makes sense.
NC SOPs should be about responding and dispositioning the instance at hand, not about preventing recurrence. NC SOP should be a "down to earth", everyday practical tool.
Wherever your system flags recurrence as an issue (it's not always so), CA is in order with all of its heavy cannons (read: resource hungry process).
PA is not directly related to all that - it should be a proactive, continuous improvement tool not directly driven by day-to-day issues.
 

somashekar

Leader
Admin
NMR .. Nonconforming Material report
is for the identification and control of the nonconforming material to prevent its unintended use or delivery. as in 8.3.2
CAR .. Corrective Action Report
is for the action to eliminate the cause of nonconformity in order to prevent recurrence. as in 8.5.2

If you force the requirement of CAR into NMR then you will begin to see a mountain of open corrective action.
Your study and analysis of the NMR must lead you to decide if you have to make a CAR.
Perhaps several NMR may need one CAR
 
G

gporter

Sorry for being a "purist", but it is critical to keep terminology in check. Corrections, such as "update a specification" is not the same as corrective actions.

Actually, in the Medical Device sector, the term CAPA seems to be misaligned with most of the rest of the world. The PA in CAPA seems to refer to prevention of recurrence of a problem and that, in the ISO 9000 definition is deemed to be corrective action (not preventive action).

It seems to me that you are really asking if we need (full blown) corrective actions with root cause analysis for each and every nonconformity and the answer is a resounding NO. We have discussed this many times and maybe someone can point to similar threads. There are many cases where system changes (the essence of corrective actions) are not justified based on severity, criticality and/or frequency of the problem.
Thanks for your thoughts. Two responses:
  1. In our CAPA system, we have an opportunity for CA and/or PA in a single CAPA file. CA would address the root cause of the nonconformance that occurred (the depth of the nonconformance), and PA would address similar systems that could have a similar issue (the breadth of the issue).
  2. I think you've misunderstood my question slightly by the way you restated it. I'll take your words and modify it: "Is it appropriate to have (full blown) corrective actions with root cause analysis for any basic nonconformity (read as "NMR") in lieu of a CAPA?" I'd like to understand your response to that :)
 
G

gporter

This all seems fraught with overcomplication and too much thinking. In your list of the things that NMRs require, #2 says that CAPA is necessary in some high-risk cases. Is there any reason that high-risk NMR findings shouldn't just be handed off to the CAPA process? Is there any good reason for tightly circumscribing the processes? Is the tail wagging the dog?
Thanks, Jim. We would hand off high risk cases to the CAPA process as a matter of process requirements. However, if the risk does not reach that high level, is it appropriate to perform corrective actions in your NMR system? Again, I could even be sold on the argument that if we chose our words properly, "corrective action"-type activities could happen in an NMR, but called something different as to not step on the toes of formal Corrective Action.
 

Jim Wynne

Leader
Admin
Thanks, Jim. We would hand off high risk cases to the CAPA process as a matter of process requirements. However, if the risk does not reach that high level, is it appropriate to perform corrective actions in your NMR system? Again, I could even be sold on the argument that if we chose our words properly, "corrective action"-type activities could happen in an NMR, but called something different as to not step on the toes of formal Corrective Action.
Disclaimer: My experience with medical devices is practically nil. Having said that, I've never worked in a system where CA isn't a normal part of the process in dealing with NC material. I'm still not clear on the motivation for creating borders and territories. We get too hung up on what things are called and lose sight of what actually needs to be done. I'm often reminded of the two guys being eaten by a carnivorous reptile while arguing about whether it's an alligator or a crocodile. If something is broken, fix it.
 
Thread starter Similar threads Forum Replies Date
L Differences between IEC 60601-1-2 and IEC TR 60601-4-2 US Food and Drug Administration (FDA) 3
K What are the differences between the certificates required for home and medical medical equipment? US Food and Drug Administration (FDA) 1
B For professionals from an amateur: What are the differences between some plans and procedures? EU Medical Device Regulations 5
Derrick Yang 各国医疗器械上市流程的区别 (Differences in the listing process of medical devices in various countries) Service Industry Specific Topics 3
S 21 CFR part 11 version differences Qualification and Validation (including 21 CFR Part 11) 1
U How to pick between ANSI Z1.4 vs. ISO2859-1 for AQL Inspection? What are the major differences? AQL - Acceptable Quality Level 4
N R&R for Differences between 2 measurements Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
A VDA2 - 6th edition June 2020 | Differences for the previous VDA2 VDA Standards - Germany's Automotive Standards 5
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S SAE AS13004 vs. SAE J1739 - What are the main differences? FMEA and Control Plans 1
M Differences in post market safety reporting for Combination Product Applicants Medical Device and FDA Regulations and Standards News 1
A MSA results differences - Supplier results vs. My results Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
DuncanGibbons Clear differences between ISO 13485 and AS 9100D requirements ISO 13485:2016 - Medical Device Quality Management Systems 10
K Differences between Manufacturing Process Audit & Layered Process Audit Lean in Manufacturing and Service Industries 4
S Differences in CB expectations between the US and Europe Registrars and Notified Bodies 3
M Authorized Rep (AR), competent authority, NCA and notified body - Differences EU Medical Device Regulations 13
D USFDA vs NRTL/IEC 17025 Differences - Compliance testing lab ISO 17025 related Discussions 0
D USFDA vs NRTL/IEC 17025 Differences - Compliance testing lab US Food and Drug Administration (FDA) 6
P HVAC System vs Air Handling Unit (AHU) - Differences Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
M What are the differences between "Process Flowchart form" and "Process Flow Diagram form"? IATF 16949 - Automotive Quality Systems Standard 5
K Differences between EN ISO 11607-1:2009 vs. 2017 Other Medical Device Related Standards 1
M What are the differences between TIR 13004:2013 and TIR 33? Other Medical Device Related Standards 2
I "Method" vs "Process" - Differences Document Control Systems, Procedures, Forms and Templates 33
S The key differences in 510k types - Traditional vs. Abbreviated vs. Special US Food and Drug Administration (FDA) 0
M Differences in CQE, CSTE, CSQA, CQA Professional Certifications and Degrees 3
H Differences between CQI 15 and CQI 17 IATF 16949 - Automotive Quality Systems Standard 0
Lancaster GMDN code differences, Manufacturer and Sponsor Other Medical Device Regulations World-Wide 3
F UMDN vs GMDN - What are the differences? EU Medical Device Regulations 1
D AS9100D vs AS9120B differences AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
A ISO 13485:2003 vs ISO 13485:2016 - What are the differences between the two? ISO 13485:2016 - Medical Device Quality Management Systems 1
F EU IVD registrations - CFG or CFS - Differences EU Medical Device Regulations 1
C Differences between IATF 16949 and AS9100 IATF 16949 - Automotive Quality Systems Standard 13
P Differences CSQA vs CQA ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 1
T PPAP differences between TS16949 and VDA APQP and PPAP 1
E Differences between IEC 60601-1-11:2015 and 2010 Other US Medical Device Regulations 0
J Differences between a Risk Management Plan vs. Production Part Approval Process AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
Y Differences between the Classic 7 QC Tools and New QC Tools Quality Tools, Improvement and Analysis 3
M Standard work, standardized work instruction & work instruction: What are differences IATF 16949 - Automotive Quality Systems Standard 7
M Barrier Free vs. Wheelchair Accessible - Differences Occupational Health & Safety Management Standards 1
F IEC 60601-1-2: Differences 2007 and 2015 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R 510(k) Executive Summary vs 510(k) Summary - Differences 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
C Differences between IATF 16949:2016 vs ISO/TS 16949:2009 IATF 16949 - Automotive Quality Systems Standard 13
A ISO 13485:2003 vs. ISO 13485:2016 Differences Comparison ISO 13485:2016 - Medical Device Quality Management Systems 4
L Differences in the requirements between IEC 60601-1 3rd ed. and EN 60601-1:2006 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
V Application and differences of OC Curve and Acceptance Sampling Statistical Analysis Tools, Techniques and SPC 2
Q What are the differences between ISO 13485 vs. AS9100C ISO 13485:2016 - Medical Device Quality Management Systems 3
L AQL ANSI z1.4 & z1.9 - What are the differences between these 2 standards? AQL - Acceptable Quality Level 5
V Differences between ISO 13485 & FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
J EMI/EMC Test differences for Radiography Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Q Differences between Clauses 7.5, 8.2.3 and 8.2.4 of ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8

Similar threads

Top Bottom