Digital Timer Calibration Requirements
#1
Greetings,
Have a question about calibrating digital timers..
In our facility have a lot of digital timers, most of them are used on processes that requires less than 5 seconds, it is really necessary to calibrate them?
I've been reading NIST Stopwatch and Timer Calibrations but still not clear for me.
Elsmar Forum Sponsor
dwperron
Trusted Information Resource
#2
It depends on how your are using them.
If you have a process where the actual time is critical (especially if these is a ± tolerance involved), or where you are reporting an actual time, then the timer should be calibrated.
If you are using it for something like "let it cure for at least 5 seconds" or "soak for about 3 seconds" then it is usually considered a convenience device, not a piece of measuring equipment.
Now the disclaimers.
Not all timers or watches come with specifications, so for calibration purposes how do you determine pass / fail?
And a lot of places just want to see calibration labels on anything that can measure, no matter how it is being used.
Your quality policy will need to decide how to handle these.
#3
If fact that's the point of using those timers.. to let the component be cured for at least 3 seconds.
About how we determine a pass/fail... Since the beginning those timers were set with a 0.1 secs tolerance and to be honest im trying to take them out from the cal system but I haven't found a solid way to justify that
dwperron
#4
If you think about it, you are just using the timer as a convenience item for your operator.
Just to be sure they wait at least 3 seconds.
They could use a clock, a watch, their phone, etc. to do this, and nobody would expect those to be calibrated.
If someone pushes the calibration issue you can try pointing out the uncertainty involved. When the operators have applied the compound to the device they need to either look at the timer or push a button to determine the start time, there is no way to determine how accurately do that. Similarly there is nothing accurate as to how they respond to the timer - it could be instant, or "in a little while".
Calibration doesn't add anything to your process to diminish risk to your process, the timer is not being used for a critical measurement.
Thread starter Similar threads Forum Replies Date
C Digital Timer Inadequate for Process? General Measurement Device and Calibration Topics 6
R DIGITAL MEASUREMENT DEVICE Manufacturing and Related Processes 2
W Pratt & Whitney Model C Supermic Digital Tailstock Meter Upgrade General Measurement Device and Calibration Topics 1
S Nadcap - Heat Treatment Proof of Verification question for Digital results AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
V ISO 9001 or ISO 17065 required for an organization issuing 5 star rating certification of digital experience? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
D Digital Healthcare - QMS requirements Other US Medical Device Regulations 3
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
C Uncertainty budgets and meeting 4:1 TUR for digital calipers and micrometer Measurement Uncertainty (MU) 0
G Calibration of Rotronic probe but not digital readout? General Measurement Device and Calibration Topics 3
DuncanGibbons Technical Data Package vs Digital Product Definition APQP and PPAP 0
D Calibration of Digital thermometer with surface probe General Measurement Device and Calibration Topics 1
A Feedback of typical maintenance problems with LMI digital probes Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
O Mitutoyo Digital Caliper to PC USB question General Measurement Device and Calibration Topics 2
J Integrity of electronic digital records - Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 4
C Teaching Metrology - Vernier vs Digital General Measurement Device and Calibration Topics 8
M Informational TGA presentation: Digital Devices Webinar 3, 20 June 2019 Medical Device and FDA Regulations and Standards News 0
M Informational US FDA – Digital Health Update: Mid-Year Update on Software Precertification Program Medical Device and FDA Regulations and Standards News 0
M Informational USFDA – Digital Health Update: The FDA Announces an Opportunity for Test Case Volunteers for the Test Plan for the Software Precertification Program Medical Device and FDA Regulations and Standards News 0
M Informational COCIR paper – Advancing Cybersecurity Of Health And Digital Technologies Medical Device and FDA Regulations and Standards News 0
P How to Calibrate a HP 3458A Digital Multimeter with Fluke 5080A calibrator General Measurement Device and Calibration Topics 1
M Digital user manuals CE Marking (Conformité Européene) / CB Scheme 0
M FDA News USFDA Digital Health Update – New actions and documents Medical Device and FDA Regulations and Standards News 0
M Informational Webinar: The role of the TGA in digital health Medical Device and FDA Regulations and Standards News 0
M FDA News Digital Health Update: USFDA Report on Non-Device Software Functions Now Available Medical Device and FDA Regulations and Standards News 0
D How to Calibrate any Digital Sphygmomanometer under use for 2 years in a Hospital Forum News and General Information 1
M Medical Device News FDA News - 13-09-18 - FDA works w/ tech industry to promote digital health innovation Other US Medical Device Regulations 0
D Kanomax MODEL 3443 DIGITAL DUST MONITOR Calibration and Metrology Software and Hardware 3
S Repairing Mitutoyo Digital Micrometers General Measurement Device and Calibration Topics 7
D AMS 2750E - Digital Instrument Calibration Accuracy Requirements General Measurement Device and Calibration Topics 1
K 21 CFR Part 11 (and EU) compliant Digital Signatures on a Production Line Other US Medical Device Regulations 5
N Digital vs Dial Caliper Comparison Measurement Uncertainty (MU) 17
rob73 Digital DMF (Device Master File) Records EU Medical Device Regulations 3
A Automated calibration of Digital Multimeter without communication port Calibration and Metrology Software and Hardware 2
Ajit Basrur Do you use Digital Pen or Digital Highlighter during audits? General Auditing Discussions 6
D Hand-held Digital Tension Meter Calibration General Measurement Device and Calibration Topics 1
G Digital Autoclave Accuracy Requirements ISO 13485:2016 - Medical Device Quality Management Systems 2
B Do I need to perform Gage R&R on a Digital Temperature & Humidity Gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 18
R Digital Storage of ITAR Controlled Documentation Other ISO and International Standards and European Regulations 7
B Digital Signature Solutions acceptable to use in an FDA Medical Device Environment 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B Measurement Software for Digital Microscopes Manufacturing and Related Processes 2
J Digital Scale Re-Calibration Interval Frequency General Measurement Device and Calibration Topics 10
W In what format do you convert documents to digital form? Document Control Systems, Procedures, Forms and Templates 10
B Digital Torque Meter Gage R&R General Measurement Device and Calibration Topics 4
R Service Recommendation for Excel Digital Signatures Qualification and Validation (including 21 CFR Part 11) 2
K FAI (First Article Inspection) using Digital Design Data AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
L Remote Control Inspection Digital Indicator with a TIR (range) function Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
M Calibrating Digital Pressure Gages In-House General Measurement Device and Calibration Topics 10
I Should we immediately have recalibrated our Digital Torque Gauge? IATF 16949 - Automotive Quality Systems Standard 5
D Procedure for use of a Digital Indicator and a Height Gage Document Control Systems, Procedures, Forms and Templates 6
M Calibration of Digital Panel Meters General Measurement Device and Calibration Topics 4
Similar threads
Top Bottom