Hello - quick and to the point, thank you for any help you can give.

Does MDR 2017/745 have a similar exception as 21 CFR 801.30(d); ie. can the UDI indicated on the label of a class 1 device only contain the UDI-DI?

Does MDR 2017/745 Annex VI s 4.10 require the UDI-DI and UDI-PI to be directly marked on the device, or can the UDI carrier on the device only include the UDI-DI, while the UDI carrier on the label includes the UDI-DI and UDI-PI?

Can the exception in MDR 2017/745 Annex VI s 4.10(b) be used to justify only marking the UDI-DI portion of the UDI carrier where the size of the full UDI carrier would make direct marking it not technologically feasible?

The device in question, a class 1 reusable device that requires cleaning and sterilization before reuse, has a UDI-DI, a lot number, and manufacturing date on the label. It is sold in Europe and the USA therefore, in accordance with 21 CFR 801.40(b), each device's full UDI carrier is the UDI-DI, the lot number, and the manufacturing date. We use the exception in 21 CFR 801.30(d) to only mark the UDI-DI on the device, although we include a full UDI, comprised of UDI-DI and UDI-PI, on the label.

The MDR, under Annex VI s 3.5, says that the manufacturing date does not need to be in the UDI-PI but come now, are we expected to issue different lots with different UDI formats for different regions because of direct marking?

Thanks for any help!

Derek
 
Elsmar Forum Sponsor

Junn1992

Quite Involved in Discussions
#2
I would say no. The UDI carrier must include both UDI-DI and UDI-PI under the MDR requirements, there is no exception clause similar to 21 CFR 801.

If you can't fit the UDI carrier, meaning AIDC + HRI on the device, then don't direct mark it at all.

The MDR, under Annex VI s 3.5, says that the manufacturing date does not need to be in the UDI-PI but come now, are we expected to issue different lots with different UDI formats for different regions because of direct marking?
You can still include it in your UDI-PI according to the UDI issuer format. Nothing wrong here.
 
#3
Thank you for your reply.

I have the same interpretation regarding the UDI carrier under the MDR.

I'm not sure how your idea of not marking at all would fly under 21 CFR 801. The UDI-DI in AIDC-only format fits fine so how could we justify not direct-marking the device?

Once it is direct-marked, the slippery slope of including the UDI-PI & the manufacturing date component come in to play.

Am I out to lunch on this? The inconsistent exceptions & requirements really seem to create a monster here. It's there something I'm missing or a different perspective?

Thanks again, I really appreciate the community's help with this and all of the other great questions here.
 

Junn1992

Quite Involved in Discussions
#4
The UDI-DI in AIDC-only format fits fine so how could we justify not direct-marking the device?
The tricky thing about the EU is that there are many notified bodies, whereas in the US you only deal with the FDA. I would advise checking with your notified body on this, since labeling is also part of technical documentation they will review.

Yes the ADIC fits fine, but not the HRI. And UDI carrier = AIDC + HRI. If we look at Annex VI, clause 6.2 reusable devices, it is implicit that the UDI = UDI-DI + UDI-PI.

Once it is direct-marked, the slippery slope of including the UDI-PI & the manufacturing date component come in to play.
You can always bundle the MFG Date into your UDI-PI.

HTH, and this was a great exercise!
 
Thread starter Similar threads Forum Replies Date
Z Direct Part Marking (DPM) requirements for Class II Medical Devices Manufacturing and Related Processes 0
J UKCA direct marking implementation UK Medical Device Regulations 0
M Medical Device News FDA Guidance - UDI: Policy Regarding Compliance Dates for Class I and Unclassified Devices and Certain Devices Requiring Direct Marking Other US Medical Device Regulations 0
C Data Matrix and DPM (direct part marking) UDI Standards - ISO/IEC TR 29158 Other US Medical Device Regulations 2
K UDI Direct Marking Compliance Date Clarification and one other UDI question Other US Medical Device Regulations 0
M UDI - Direct Marking and Reprocessing Clarification Other US Medical Device Regulations 12
A Direct vs. Post-License Investigational Test Authorization (ITA) Canada Medical Device Regulations 0
T Is there a direct relation between surface roughness and gloss? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
R Class 1 Convenience kits for home teeth impression, NOT smile direct or other for alignment purposes Canada Medical Device Regulations 0
W Direct to customer export of medical device (class I: prescription lenses + frame) US Food and Drug Administration (FDA) 2
J COC (Certificate of Conformity) when NOT purchasing medical devices direct from the Manufacturer Other Medical Device Related Standards 2
Ahang "DIRECT MODE" or "ADJUSTED MODE" according to ISO 80601-2-56 Other Medical Device Related Standards 1
T Type BF or CF (Direct conductive contact with the heart) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
lyobovnik Matters being audited, independence, participation and direct responsibility 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
JoCam UDI - Direct Labelling EU Medical Device Regulations 9
N PFMEA for Non-Manufacturing Processes such as audits, direct procurement, etc. FMEA and Control Plans 5
G Interpretation of phrase - "Direct Authority" EASA and JAA Aviation Standards and Requirements 5
C Meaning of "direct diagnosis" (Annex IX, Rule 10) EU Medical Device Regulations 6
S Difference between Direct & Indirect Elisa (Enzyme-Linked ImmunoSorbent Assay) Food Safety - ISO 22000, HACCP (21 CFR 120) 1
L Shipping Medical Devices and Pharmaceuticals Direct to Physicians - Puerto Rico US Food and Drug Administration (FDA) 4
T Duties and Responsibilities as a Direct Supervisor of a Pregnant Employee Occupational Health & Safety Management Standards 12
S ISO 10993 Cytotox Testing - Direct Contact rather than Extractables Other Medical Device Related Standards 4
A Medical Devices - Direct vs. Indirect Body Contact - Definitions Other US Medical Device Regulations 4
M AS9100 Requirements for Direct and Indirect Procurement AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
I Magnehelic Calibration by Direct Comparison with a Shortridge Airdata ADM-870C Measurement Uncertainty (MU) 1
E Material Review Board - We will get FAA Direct Ship Authority Quality Manager and Management Related Issues 4
I Does anyone have direct experience with BVQI (Brazil)? Registrars and Notified Bodies 3
G Environmental Issues - Indirect (influence)/ Direct Aspects Miscellaneous Environmental Standards and EMS Related Discussions 6
K Definitions of direct & indirect cause that needs to be answered in a 5 WHY analysis? Quality Tools, Improvement and Analysis 11
H What is 'Direct Run First Time Quality Result' - IATF Guide, Para 7.4.3.2? IATF 16949 - Automotive Quality Systems Standard 6
W ISO 13485 Certification - No Sterilization, Packing nor Direct Shipping to End User. ISO 13485:2016 - Medical Device Quality Management Systems 6
R Tool Kits to Reduce Direct Time on Milling Machine Manufacturing and Related Processes 15
SteelMaiden Competency evaluation records - Direct correlation to EMS-related topics Miscellaneous Environmental Standards and EMS Related Discussions 7
M SPC is not a direct requirement of ISO 9001:2000, correct? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
T PPAP supplied to our customer Direct from Heat Treater supplier Supplier Quality Assurance and other Supplier Issues 3
Marc Micro direct methanol fuel cell system for IBM ThinkPad notebooks After Work and Weekend Discussion Topics 0
E Definition Management - Definition of - Coordinates activities to direct and control Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 5
D Over-Worked Instruction - Quality Technician for a direct mail printing company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
L ISO 14001 Aspect and Impact Assessment - Direct and Indirect Environmental Impacts ISO 14001:2015 Specific Discussions 18
S Suppliers - How "direct" is "direct"? Services DIRECTLY to Chrysler, Ford, GM Supplier Quality Assurance and other Supplier Issues 7
K What is the difference between Direct and Indirect product? Other ISO and International Standards and European Regulations 21
CarolX Self Direct Training General Auditing Discussions 4
R Is a complete Internal Audit prior to registration a direct requirement of QS9000? Internal Auditing 3
H Marking of Resin Identification Code CE Marking (Conformité Européene) / CB Scheme 0
G UKCA and EU MDR marking UK Medical Device Regulations 5
A Distributors with stock of expired CE marking EU Medical Device Regulations 2
R CE marking of products not sold into the EU CE Marking (Conformité Européene) / CB Scheme 4
Y CE marking with NB identification number EU Medical Device Regulations 0
B ASTM A53 Marking of length on recut pipes by processor Manufacturing and Related Processes 11
M FDA Syringe Marking requirements US Food and Drug Administration (FDA) 4

Similar threads

Top Bottom