Consider TL 9000 requirement 4.11.2.H.1 Identified Equipment - IM&TE that is either inactive or unsuitable for use shall be identified and not used for production. All IM&TE that does not require calibration shall be identified. IOW, if used for product acceptance or control of significant Special process parameters, calibrate (or verify) it. If not, identify it as Calibration Not Required. All equipment should be identified, whether calibrated or not.