Do MDD (Medical Device Directives) Devices require MDD components?
P
ptaillie
I am reviewing a DR Imaging system w/ a Notified Body. The DR detector is CE marked, but not to Class IIa. The auditor says this component of the system must be Class IIa certified before the system can be certified as Class IIa. Alternatively, if all of the Essential Requirements are shown to be met by the detector, the Class IIa cert is not required for the detector. The problem is that the detector vendor has not done all of the Class IIa testing. Is this a real requirement? Was it introduced in the 2007/47/EC change? I know I have sold detectors in EU as part of an imaging system w/o Class IIa cert before, under the mindful eyes of a Notified Body. Thanks for any help you can give.
MIREGMGR
It's fundamental that if a system-product is to be type-approved based on existing conformance determinations for its subsidiary critical components, those existing conformance determinations must be for at least the class to which the system-product is to be type-approved; or, that the system-product must be determined to be conformant on its own, without reference to existing conformance determinations for subsidiary components.
As far as I know, conformance determination for systems has always been properly handled via one of these paths or the other.
I would expect a type approval authority to exhibit flexibility regarding required classification for non-critical components, but obviously the detector of an imaging system wouldn't be subject to that exception.
P
ptaillie
Thank you both for your responses. I agree with your assessment. I am just confused as to how the two companies I worked at in the past did not have to provide all of this proof for non-Class IIa detectors (when included in Class IIa devices). Further, I don't know why big detector companies are not getting Class IIa approval, and how they have sold product in EU without it, or the underlying proof. This is putting a monkey wrench in my present product introduction plans, and I want to be sure I really need this to be done before I put a lot of time into it. Thanks again.
Ronen E
Problem Solver
Moderator
Hello and welcome to the cove :bigwave:
It is very important not to confuse between medical devices and components thereof, wherever the latter are not medical devices in their own right.
I am no expert on DR, so I'm referring to the generic principles; you can then apply them to your field.
Speaking in general terms, there is no generic requirement for medical device components to qualify for CE-marking in order for a system (or device) containing them to qualify as well. As far as I know, this is true regardless of class and a given component criticality. There might be (easier) qualification routes that are based on such hierarchy, but typically there would be more than one path to CE-marking and it is up to you to consider and conclude which are accessible to your company / product, and which are preferred.
Of course, there could be specific requirements applicable to DR - as I stated above I'm no expert in this field. If this is actually the case, you can politely ask your NB to point the exact regulation section they refer to, and continue from there.
Having said all that, if the component (in this case, the detector) can be considered a medical device on its own, then yes, it would be required to comply with the MDD in order to be sold in the EC (even if the buyer esentially uses it as a component). The definition of a medical device can be found in the MDD and is quite clear in my opinion; you could try and read it on the detector as-received, and see if it holds. regardless of that, the detector manufacturers must comply with all (other) applicable EC directives before they sell these detectors in the EC, and if any of theses directives is one that leads to CE-marking then the detector should be marked accordingly (not necessarily as a medical device).
Cheers,
Ronen.
MIREGMGR
I 100% agree with Ronen's comment, and would note that this underlies the alternate path of a full Essential Requirements analysis of the imaging system without reference to any pre-existing MDD conformance establishment for the detector module.
AMIGOS
This is my first post....so bear with.
I have a customer who wants the components that we use (for a device we are designing) to be medically approved - I am right in presuming medically approved is CE marked or is there another method? Also, what implications would this have on the device? Can I presuade them against this as the device will go through CE marking approval/clinical investigations etc prior to release (the device is Class IIb)
AM
somashekar
AM
Thanks to Ronen E.
Please make a clear distinction between component and accessory to a medical device. An accessory is a medical device in its own capacity and can be placed in the market as a medical device and this will have to be CE marked per its class. Take the case of a pulseoximeter and its probe. The pulseoximeter would be class IIb while the probe can be designed and assessed to meet class IIa or class IIb.
Component on the other hand by itself cannot be a medical device in its own capacity. I have seen several components where in the data sheet mentions meeting several safety standards and its typical application areas including medical devices, with a note that when used in medical device, the fitness for its use must be assessed by the medical device designers.
AMIGOS
Thank u!
Yep by component I mean the actual components e.g. resistors/ICs etc that make up the device, not an accessories.
Ronen E
Problem Solver
Moderator
Thank u!
In this case, the closest thing I know to "medically approved" is compliance with applicable parts of IEC 60601 series, though IMHO it makes little sense to apply this standard to resistors and such (as it is intended for electrical medical equipment).
I would strongly urge you to try and explain to your customer that "medically" qualifying the components has little added value on top of the high-level MDD compliance (CE marking) of the final device. If they seek extra confidence in the quality of raw ingredients, then I believe there's a host of more appropriate industry standards to ensure resistors and such are of high quality. Having said all that, it's always important in business to remember that "the client has the right to be the client":cool:, meaning, if the client is not happy with you / the product, they may seek happiness somewhere else :(
Cheers,
Ronen.
MIREGMGR
There's another possible interpretation of a design/manufacturing customer spec that components must be "medically approved".
Many makers of component materials and devices are gunshy about the use of their products in applications that have an adverse history regarding civil lawsuits in the event of less-than-optimal performance. Such component makers frequently impose terms of sale that attempt to establish an agreement-by-purchase not to use their products in those product types--frequently, in medical devices--and to indemnify the component maker against judgments, damage to reputation, costs of staff time, legal counsel costs, etc.
Such conditions and contractural barriers are very common in regard to two use-types...life safety systems, and implantables.
Thus, "medically approved" might refer to a component maker not establishing such constraints against medical-device use.
H MDD Article 12 Labeling for Class IIa Medical Device - Please Advise EU Medical Device Regulations 3
A Class medical device (MDD) - Classification help EU Medical Device Regulations 1
P Adding "new" medical device to existing CER - under MDD CE Marking (Conformité Européene) / CB Scheme 1
T Custom made medical device MDD vs MDR definition CE Marking (Conformité Européene) / CB Scheme 12
W CE Mark for Class III Medical Device - MDR or MDD applies? CE Marking (Conformité Européene) / CB Scheme 10
F MDD Article 12 Labeling Requirements for CE Marked Medical Device - Help CE Marking (Conformité Européene) / CB Scheme 4
S Final Medical Device Regulations replacing MDD 93/42/EEC, etc. EU Medical Device Regulations 15
N MDD Class I Medical Device Self Declaration Articles/Supplementary Documents EU Medical Device Regulations 1
R Class I Medical Device will be Class IIa after new MDD EU Medical Device Regulations 3
rob73 New MDD (European Medical Device Regulations) next steps - 2016 EU Medical Device Regulations 1
J0anne MDD 93/42/EEC Medical Device Classification EU Medical Device Regulations 3
M Do I need to comply with both the MDD and the PED for my Medical Device? EU Medical Device Regulations 8
P EU MDD revisions white paper on the proposals for the Medical Device Regulations EU Medical Device Regulations 1
A MDD 93/42 EEC - Requirements for Detachable Medical Device Components EU Medical Device Regulations 4
S MDD 93/42 Annex I 12.3 - Medical Device Electrical Alarm Requirements EU Medical Device Regulations 3
B Bangladesh Class IIA Medical Device MDD Registration Requirements Other Medical Device Regulations World-Wide 6
C Mobile Medical Device for MDD - Medical Image Viewer EU Medical Device Regulations 2
M Proposed EU MDD (Medical Device Directive) Revision(s) - 2012 - 2013 EU Medical Device Regulations 16
Q Storage Temperature - Medical Device - Objective Evidence - MDD 93/42/EEC Other Medical Device Regulations World-Wide 1
B Certificate of Exportability of Class IIA (EU MDD) Medical Device US Food and Drug Administration (FDA) 1
D MDD (Medical Device Directive) 6.1 Record Retention Requirements question Records and Data - Quality, Legal and Other Evidence 2
L MDD (Medical Device Directive) Compliance Review checklist wanted EU Medical Device Regulations 9
G IVD (In Vitro Diagnostic) MDD (Medical Device Directive) Borderline Issues EU Medical Device Regulations 3
T MDD (Medical Device Directive) and ISO 13485 and a lot of Questions EU Medical Device Regulations 3
M MDD (Medical Device Directive) Online Training Recommendations and Suggestions EU Medical Device Regulations 1
xcanals_tecno-med.es Revised List of Harmonized Standards for Medical Device Directives (MDD) in the EU EU Medical Device Regulations 48
R Declaration of Conformity - MDD (Medical Device Directive) 2007/47/EC or 93/42/ECC CE Marking (Conformité Européene) / CB Scheme 9
ScottK MDD (Medical Device Directive) and Secondary Packaging EU Medical Device Regulations 3
E Does MDD (Medical Device Directive) Require a CAPA Procedure? Document Control Systems, Procedures, Forms and Templates 3
B What MDD (Medical Device Directive) Annex to follow for Class I device? EU Medical Device Regulations 14
D Requirement of "State Of The Art" in MDD (Medical Device Directive) EU Medical Device Regulations 42
P Medical Device Vigilance - MDR or MDD? EU Medical Device Regulations 9
E MDD 2007/47/EEC - Medical Device Directive Labeling Changes EU Medical Device Regulations 44
J Medical Device Directive (MDD) 2007/47/EC and 2006/42/EC EHSR checklist EU Medical Device Regulations 7
L Containment Device Classification - Class I Medical Device via Rule 1 MDD Annex IX EU Medical Device Regulations 1
L MDD Definition of Address as it applies to Medical Device Labeling EU Medical Device Regulations 20
S Standards used for Medical Device MDD Technical Files ISO 13485:2016 - Medical Device Quality Management Systems 25
W MDD (Medical Device Directive) Essential Requirements Checklist ISO 13485:2016 - Medical Device Quality Management Systems 18
echo7tango MDD (Medical Device Directive) Update? The copy I have is 14 June 1993 (93/42/EEC) Other ISO and International Standards and European Regulations 13
S MDD (European Medical Device Directive) requires subset of ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 10
W New List of Medical Device Harmonized Standards - MDD CE Marking (Conformité Européene) / CB Scheme 5
chris1price The status of the 2005 ammendments to the MDD (Medical Device Directive)? EU Medical Device Regulations 4
T Calculating Basic Risk/ Basic on Risk Analysis - Medical device MDD Class II and up ISO 13485:2016 - Medical Device Quality Management Systems 8
Q Medical Device Directive - Anyone know of any good MDD training courses Training - Internal, External, Online and Distance Learning 11
V CE Marking for Power Supply for Medical Device - Class IIb device per MDD, Rule 11 EU Medical Device Regulations 3
H Regulation in Australia - Importing medical device (class IIa as per MDD 93/42/EEC) ISO 13485:2016 - Medical Device Quality Management Systems 3
C Medical Device QMS manual - MDD, AIMD, 21 CFR part 820, CMDCS, and the new ISO 13485 Quality Management System (QMS) Manuals 3
D Getting Started - ISO13485:2003 and MDD (Medical Device Directive) ISO 13485:2016 - Medical Device Quality Management Systems 9
N When is "placing on the market"? MDD certified medical devices EU Medical Device Regulations 32
R Identification of Medical Devices in MDD 93/42 Certificate EU Medical Device Regulations 2