Do MDD (Medical Device Directives) Devices require MDD components?
Ronen E
Problem Solver
Moderator
#11
There's another possible interpretation of a design/manufacturing customer spec that components must be "medically approved".
Many makers of component materials and devices are gunshy about the use of their products in applications that have an adverse history regarding civil lawsuits in the event of less-than-optimal performance. Such component makers frequently impose terms of sale that attempt to establish an agreement-by-purchase not to use their products in those product types--frequently, in medical devices--and to indemnify the component maker against judgments, damage to reputation, costs of staff time, legal counsel costs, etc.
Such conditions and contractural barriers are very common in regard to two use-types...life safety systems, and implantables.
Thus, "medically approved" might refer to a component maker not establishing such constraints against medical-device use.
Sounds more like "not medically restricted"...:)
alimary15
#12
Good morning to everyone,
I am new to this forum and I found this discussion a lot similar to what I am actually tryign to search and understand.
My company want to sell to the EU market a flat panel detector based on X-Ray Imagining. These detectors can have both medical and industrial applications and my company would like to sell it not as a Medical Device, but as a "Component".
By selling it as a "Component", there will then be a third part manufacturer, that is going to assembly this device into a medical device system, thus following Article 12 Declaration, be repsonsbile for its CE requiremtns.
I would like to know what CE mark should have our detector in order to be put into market as a "component" and not as a medical device??
Hope to receive some help about this!!!
Regards,
Mary
Ronen E
Problem Solver
Moderator
#13
Regards,
Mary
Hello Mary and welcome to the Cove :bigwave:
Every product offered for sale in the EC has to comply with all applicable legislative instruments: Directives, Regulations, Decisions etc. It is the legal manufacturer (entity placing the product on the EC market) responsibility to research and identify all such applicable instruments, and then comply with them all as applicable. Start here:
http://ec.europa.eu/enterprise/poli...ods/internal-market-for-products/index_en.htm
Some Directives (not all) lead to CE marking. If one of those applies to your product, you are entitled and expected to CE mark your product once you and it are compliant with all the applicable requirements. In general, the CE mark is the legal manufacturer's declaration that the manufacturer and the product are compliant with all applicable requirements, under all applicable Directives that lead to CE marking (could be more than one in parallel, for a given product).
As part of the initial sorting of applicable instruments / requirements, I recommend that you review your detector as-offered-for-sale, including all labelling and associated promotional / marketing materials, taking account of its intended use as your company sees it (and conducts accordingly), and decide whether or not it falls within the MDD's scope. If it does, you don't have a choice - it can only be CE marked as a medical device, unless you change the product and/or the labelling and/or the intended use, so as to preclude it from the MDD's scope. If it doesn't, then any other applicable CE marking route you follow will qualify it as a "non-medical-device CE marked product". In general, there is no difference in the mark itself.
I hope I managed to clarify things a little...
Cheers,
Ronen.
alimary15
#14
Hello and thanks for your answer...
I am sorry I am very new about EC regulations and CE marking.
My device can have both medical and industrial applications, but my company would like to classify it as a non MDD and so avoiding to have ISO 13485 or going trought a Notify Board.
We have then another company which is developing the software for our device, which it IS already marked CE as MDD. this other company will be classified as an "assembler" and so should be responsible for integrating our device as a "component" into a medical device kit.
We just would like to sell the detector as a "imagining system" without specifying if the application its medical or industrial, because it can be for both!
I just don't know in what kind of EC Directives this panel can be classifed into.
Ronen E
Problem Solver
Moderator
#15
If it can't / not intended to be used as a medical device (see definition in MDD) when you supply it and as you supply it, then it's not a medical device. the fact that it can / will / intended to later be integrated into a medical device doesn't change that.
So maybe a little homework can help:
http://europa.eu/youreurope/business/profiting-from-eu-market/selling-goods/index_en.htm
Cheers,
Ronen.
alimary15
#16
Thank you for your help!
I also suppose that it won't be classifed as a medical device, unless my company decide to sell it as a "medical" purpose device.
However, I am sure there should be some directives that apply to my product so I will be looking into the papers..
I will post back soon if (as I am sure will be ) I will have other questions!
Thanks again!
alimary15
#17
Hello again,
I was reading trought some EU directives....:frust:
so my panel is not a Machinery ( no part moving), is not a Low-Voltage-Device (Since it is used in radiology) ... and it can have multi-purposes (medical yes, but also indstrial,sceintific, etc) .
Let's say I want to classify it as a "general pursose" X-ray detector.
So the only applicable directive I can think of is the EMC directive of electro-magnetic compatibility that can lead me to the CE marking.
Any other suggestion?
Ronen E
Problem Solver
Moderator
#18
Hello again,
Drafting a regulatory clearance strategy for a given product usually takes me more than 4 hours, even when I have access to all required information. Sometimes I even get paid for it...;)
If you have studied the legislation to your satisfaction, and are confident with your findings, then you can proceed. The regulatory system should have enough safety mechanisms to ensure that unsafe / undue conduct on manufacturers part will be identified along the route -- if you are inadvertently doing something wrong someone with authority will tell you, sooner or later. Oversight is typically commensurate with the risk your product encompasses to the public & environment.
Regulatory specialists make their living by helping you ensure you don't get told you're doing something wrong, or at least you get told rather sooner than later (when the consequences are easier to handle).
Cheers,
Ronen.
H MDD Article 12 Labeling for Class IIa Medical Device - Please Advise EU Medical Device Regulations 3
A Class medical device (MDD) - Classification help EU Medical Device Regulations 1
P Adding "new" medical device to existing CER - under MDD CE Marking (Conformité Européene) / CB Scheme 1
T Custom made medical device MDD vs MDR definition CE Marking (Conformité Européene) / CB Scheme 12
W CE Mark for Class III Medical Device - MDR or MDD applies? CE Marking (Conformité Européene) / CB Scheme 10
F MDD Article 12 Labeling Requirements for CE Marked Medical Device - Help CE Marking (Conformité Européene) / CB Scheme 4
S Final Medical Device Regulations replacing MDD 93/42/EEC, etc. EU Medical Device Regulations 15
N MDD Class I Medical Device Self Declaration Articles/Supplementary Documents EU Medical Device Regulations 1
R Class I Medical Device will be Class IIa after new MDD EU Medical Device Regulations 3
rob73 New MDD (European Medical Device Regulations) next steps - 2016 EU Medical Device Regulations 1
J0anne MDD 93/42/EEC Medical Device Classification EU Medical Device Regulations 3
M Do I need to comply with both the MDD and the PED for my Medical Device? EU Medical Device Regulations 8
P EU MDD revisions white paper on the proposals for the Medical Device Regulations EU Medical Device Regulations 1
A MDD 93/42 EEC - Requirements for Detachable Medical Device Components EU Medical Device Regulations 4
S MDD 93/42 Annex I 12.3 - Medical Device Electrical Alarm Requirements EU Medical Device Regulations 3
B Bangladesh Class IIA Medical Device MDD Registration Requirements Other Medical Device Regulations World-Wide 6
C Mobile Medical Device for MDD - Medical Image Viewer EU Medical Device Regulations 2
M Proposed EU MDD (Medical Device Directive) Revision(s) - 2012 - 2013 EU Medical Device Regulations 16
Q Storage Temperature - Medical Device - Objective Evidence - MDD 93/42/EEC Other Medical Device Regulations World-Wide 1
B Certificate of Exportability of Class IIA (EU MDD) Medical Device US Food and Drug Administration (FDA) 1
D MDD (Medical Device Directive) 6.1 Record Retention Requirements question Records and Data - Quality, Legal and Other Evidence 2
L MDD (Medical Device Directive) Compliance Review checklist wanted EU Medical Device Regulations 9
G IVD (In Vitro Diagnostic) MDD (Medical Device Directive) Borderline Issues EU Medical Device Regulations 3
T MDD (Medical Device Directive) and ISO 13485 and a lot of Questions EU Medical Device Regulations 3
M MDD (Medical Device Directive) Online Training Recommendations and Suggestions EU Medical Device Regulations 1
xcanals_tecno-med.es Revised List of Harmonized Standards for Medical Device Directives (MDD) in the EU EU Medical Device Regulations 48
R Declaration of Conformity - MDD (Medical Device Directive) 2007/47/EC or 93/42/ECC CE Marking (Conformité Européene) / CB Scheme 9
ScottK MDD (Medical Device Directive) and Secondary Packaging EU Medical Device Regulations 3
E Does MDD (Medical Device Directive) Require a CAPA Procedure? Document Control Systems, Procedures, Forms and Templates 3
B What MDD (Medical Device Directive) Annex to follow for Class I device? EU Medical Device Regulations 14
D Requirement of "State Of The Art" in MDD (Medical Device Directive) EU Medical Device Regulations 42
P Medical Device Vigilance - MDR or MDD? EU Medical Device Regulations 9
E MDD 2007/47/EEC - Medical Device Directive Labeling Changes EU Medical Device Regulations 44
J Medical Device Directive (MDD) 2007/47/EC and 2006/42/EC EHSR checklist EU Medical Device Regulations 7
L Containment Device Classification - Class I Medical Device via Rule 1 MDD Annex IX EU Medical Device Regulations 1
L MDD Definition of Address as it applies to Medical Device Labeling EU Medical Device Regulations 20
S Standards used for Medical Device MDD Technical Files ISO 13485:2016 - Medical Device Quality Management Systems 25
W MDD (Medical Device Directive) Essential Requirements Checklist ISO 13485:2016 - Medical Device Quality Management Systems 18
echo7tango MDD (Medical Device Directive) Update? The copy I have is 14 June 1993 (93/42/EEC) Other ISO and International Standards and European Regulations 13
S MDD (European Medical Device Directive) requires subset of ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 10
W New List of Medical Device Harmonized Standards - MDD CE Marking (Conformité Européene) / CB Scheme 5
chris1price The status of the 2005 ammendments to the MDD (Medical Device Directive)? EU Medical Device Regulations 4
T Calculating Basic Risk/ Basic on Risk Analysis - Medical device MDD Class II and up ISO 13485:2016 - Medical Device Quality Management Systems 8
Q Medical Device Directive - Anyone know of any good MDD training courses Training - Internal, External, Online and Distance Learning 11
V CE Marking for Power Supply for Medical Device - Class IIb device per MDD, Rule 11 EU Medical Device Regulations 3
H Regulation in Australia - Importing medical device (class IIa as per MDD 93/42/EEC) ISO 13485:2016 - Medical Device Quality Management Systems 3
C Medical Device QMS manual - MDD, AIMD, 21 CFR part 820, CMDCS, and the new ISO 13485 Quality Management System (QMS) Manuals 3