Do Reference materials for IATF need to be 17025?

#1
Do calibration oils and buffers (like viscosity oils and pH buffers) need to come from a 17025 lab to comply with IATF 16949 requirements?

We're slowly transitioning from ISO 9001 to IATF and am trying to figure out the nuances to present to the owner.
 
Elsmar Forum Sponsor

Ninja

Looking for Reality
Trusted Information Resource
#3
Really, @Guest?

No, they don't...but if you are using them as a tool to measure or determine acceptance criteria they must be NIST traceable (calibrated, though I'm not so sure that term can really apply to a buffer solution, I suppose it could), or at least validated as accurate in a manner your customer accepts (documented).
All of my auto customers have required A2LA, none have required 17025...but even then it was only for PPAP submission.

Are you selling the oils and buffers to auto? Or are they just tools you use to measure something else?
 
#4
Really, @Guest?

No, they don't...but if you are using them as a tool to measure or determine acceptance criteria they must be NIST traceable (calibrated, though I'm not so sure that term can really apply to a buffer solution, I suppose it could), or at least validated as accurate in a manner your customer accepts (documented).
All of my auto customers have required A2LA, none have required 17025...but even then it was only for PPAP submission.

Are you selling the oils and buffers to auto? Or are they just tools you use to measure something else?
We're in the paint and coatings industry, specifically for wheels. So we don't make buffers and oils, but rather use them for internal verifications/calibrations of our pH probes and viscosity cups.

7.1.5.3.2 External laboratory in the IATF standard says that "facilities used for inspection, test, or calibration services" needs to be done by a 17025 lab (or equivalent/ approved by supplier). So I was wondering if this requirement extended to oils and buffers.
 

Ninja

Looking for Reality
Trusted Information Resource
#6
I think he was snarkily trying to say that's not where they come from...Labs don't typically sell that stuff.

#1, you aint selling it (the buffers or oils) so PPAP with what you've got...if the customer signs it, you have customer approval.
#2... don't read the standard like "what extra do I need to do?"...read it like "I'm sure we already do this, where is my evidence?"

Only if you don't find any evidence should you start looking at adding things.
We're in the paint and coatings industry, specifically for wheels.
Seems like you're already doing it... so you should already have ample evidence of customer approval...a repeat PO if nothing else (though that's a bit thin). The customer signed PPAP would be way better...and if you're already selling into auto you may have that already.
 

John C. Abnet

Teacher, sensei, kennari
Staff member
#7
customers have required A2LA, none have required 17025
Good day @Ninja ;
So, hate to be "that guy" but thought it was worth while to point out that a2la is an organization and not a set of requirements. a2La accredits to given standards, such as...
ISO/IEC 17025 for Testing/Calibration Laboratories s
ISO/IEC 17020 for Inspection Bodies
ISO/IEC 17065 for Product Certification Bodies
ISO 20387 for Biobanking Facilities
ISO 15189 and CLIA for Clinical Testing Laboratories
ISO/IEC 17043 for Proficiency Testing Providers

ISO 17034 for Reference Material Producers

So if your customers "required a2la", I can only assume they were accrediting your organization to ISO 17025 (or other if applicable). Let me know if I'm missing something.

Be well
 

dwperron

Trusted Information Resource
#8
Do calibration oils and buffers (like viscosity oils and pH buffers) need to come from a 17025 lab to comply with IATF 16949 requirements?

We're slowly transitioning from ISO 9001 to IATF and am trying to figure out the nuances to present to the owner.

16949 does not require 17025 accreditation, it requires "conformity" to 17025. What it states is that if you and the outside sources you use are 17025 accredited then they consider to conform. If you choose not to get accredited to 17025, when an 16949 audit comes around they can ask for evidence as to how you conform to 17025.....
That's why a lot of 16949 organizations use 17025 accredited calibration houses and supplies. It just makes the audit process easier.
 

Ninja

Looking for Reality
Trusted Information Resource
#9
Thanks John,
You didn't miss anything...I simply wasn't precise (much at all, apparently as I reread) in what I wrote.

All of my auto customers have required calibration with NIST trace for measurements we did in house (reported or not) on their parts at any point in the process,
and that if we outsourced any measurements we couldn't do in house that we use an A2LA accredited lab if such a thing existed for the property measured.
We took that to additionally mean "if we send a piece out to an outside lab for verification that our measurement is adequate, it should be A2LA", and never had a customer push back on it.
 

Howard Atkins

Forum Administrator
Staff member
Admin
#10
16949 does not require 17025 accreditation, it requires "conformity" to 17025. What it states is that if you and the outside sources you use are 17025 accredited then they consider to conform. If you choose not to get accredited to 17025, when an 16949 audit comes around they can ask for evidence as to how you conform to 17025.....
That's why a lot of 16949 organizations use 17025 accredited calibration houses and supplies. It just makes the audit process easier.
Are you talking about internal or external labs
There are different requirements for both
7.1.5.3.1 Internal Laboratory
An organization's internal laboratory facility shall have a defined scope that includes its capability to perform the required inspection, test, or calibration services. This laboratory scope shall be included in the quality management system documentation. The laboratory shall specify
and implement, as a minimum, requirements for:
a) adequacy of the laboratory technical procedures;
b) competency of the laboratory personnel;
c) testing of the product;
d) capability to perform these services correctly, traceable to the relevant process standard (such as ASTM, EN, etc); when no national or international standard(s) is available, the organization shall define and implement a methodology to verify measurement system capability;
e) customer requirements, if any;
f) review of the related records.
NOTE Third-party accreditation to ISO/IEC 17025 or national equivalent may be used to demonstrate the organization's in-house laboratory conformity to this requirement.

The standard states
7.1.5.3.2 External Laboratory
External/commercial/independent laboratory facilities used for inspection, test, or calibration services by the organization shall have a defined laboratory scope that includes the capability to perform the required inspection, test, or calibration, and either:
- the laboratory shall be accredited to ISO/IEC 17025 or its national equivalent (e.g., CNAS-CL01 in China) by an accreditation body (Signatory) of the ILAC MRA (International Laboratory Accreditation Forum Mutual Recognition Arrangement – www.ilac.org) and include the relevant inspection, test, or calibration service in the scope of the accreditation (certificate); the certificate of calibration or test report shall include the mark of a national accreditation body; or - there shall be evidence that the external laboratory is acceptable to the customer.

If you use an external lab then it MUST be accredited or approved by your customer.
 
Thread starter Similar threads Forum Replies Date
R ASQ reference material clarification - Spiral bound materials allowed in ASQ Exam? Professional Certifications and Degrees 1
E Use of Certified Reference Materials for Quality Assurance General Measurement Device and Calibration Topics 6
H Can I get some information... CRM (Certified Reference Materials) General Measurement Device and Calibration Topics 3
R Policy Documents which contain Reference Materials Citations Document Control Systems, Procedures, Forms and Templates 1
S Using Prepared In-House Reference Materials (RM) for Equipment Calibration & Verifica General Measurement Device and Calibration Topics 3
S Training and Reference Materials for ISO 14971 ISO 14971 - Medical Device Risk Management 4
AnaMariaVR2 New NIST reference materials available for popular diet supplement Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
C What is reference standard and reference materials as per clause 5.4.5.2 note 2? General Measurement Device and Calibration Topics 1
G GD&T applied to reference dimension. Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
C IVD Metrological traceability without a reference measurement procedure General Measurement Device and Calibration Topics 11
M Reference part calibration or verification General Measurement Device and Calibration Topics 12
G Need journal and reference of abnormal bar chart SPC Statistical Analysis Tools, Techniques and SPC 4
W What is okay to use as a reference material for ISO 17025? ISO 17025 related Discussions 4
M SOPs - Reference Documents Section Document Control Systems, Procedures, Forms and Templates 4
B Reference to IEC 60950-1 in Amendment 2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Question on equipment - when to use reference only or research only stickers ISO 13485:2016 - Medical Device Quality Management Systems 5
J Marking "Distributed By" to product with distributors' reference-code EU Medical Device Regulations 3
Dazza 9001, 14001 and 45001 mandatory documents and records cross reference matrix Process Maps, Process Mapping and Turtle Diagrams 2
T CSE 5.01 reference on drawing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
CycleMike GD&T - Hole pattern - Print (attached) has a single Datum Reference Frame Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
J Sample size for creating a data base as a reference to a tested variable Other Medical Device and Orthopedic Related Topics 6
D Good Acceptance Activity/Inspection flowchart for reference - Wanted Please Process Maps, Process Mapping and Turtle Diagrams 2
G Customer complaint over a Reference Dimension Manufacturing and Related Processes 9
G Measure 3 times, for calibration report? Customer Owned Reference Standard General Measurement Device and Calibration Topics 4
I Data Backup Plan - Document Reference ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Q How to efficiently and compliantly reference external standards/regulations Other Medical Device Related Standards 2
S Lot assignment for secondary reference standard - Pharmaceuticals Manufacturing and Related Processes 0
V Inspection with 3D Scanning - Reference Data Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
E Ethylene Oxide Sterilization Validation-Reference Load Other Medical Device Related Standards 2
C Must your reference standard provider be ISO17034 certified to meet your testing lab's ISO 17025 certification requirements? Other ISO and International Standards and European Regulations 2
C Do Calibration Reference Standards require an MSA study? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
J Purpose of Reference Section in SOPs Document Control Systems, Procedures, Forms and Templates 6
M Quality Policy - Standards and reference numbers ISO 13485:2016 - Medical Device Quality Management Systems 14
M Reference to IATF 16949 clause 8.4.3 in clause 8.3.4.4, Is it right? IATF 16949 - Automotive Quality Systems Standard 1
S Reference calibration tools (traceable to NIST) that are not used in production ISO 13485:2016 - Medical Device Quality Management Systems 4
S Contract Manufacturer Question - Whether or not the reference to the ex-CMO ISO 13485:2016 - Medical Device Quality Management Systems 6
B Need to convert Reference Only M&TE to a calibration schedule General Measurement Device and Calibration Topics 3
M Looking for Cross reference of AS9100:2016 to ISO13485 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Print only has Reference Dimensions Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
V QMS / QSR - Reference to Framework - Templates - Procedures for Medical Devices 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
R How to Cross-Reference Controlled Documents Document Control Systems, Procedures, Forms and Templates 3
E IEC 60601-2-25 Reference to 60601-1 2nd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Visual Inspection with No Visual Reference General Measurement Device and Calibration Topics 2
M AS9100 Rev. D - Can anyone recommend a good reference book? Book, Video, Blog and Web Site Reviews and Recommendations 7
B When to use "For Reference Use Only" on documentation Document Control Systems, Procedures, Forms and Templates 4
G BCARs/EASA to ISO9001/AS9100 cross reference matrix/table wanted EASA and JAA Aviation Standards and Requirements 1
J In the ISO 9001:2008 company, can give reference to sister company also. ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
A Reference Source for Polarity Symbol Other ISO and International Standards and European Regulations 1
Q Multiple Quality Standards - AS9100 and ISO13485/9001 Cross Reference Matrix Quality Manager and Management Related Issues 3