Do We Have to Comply with Our Own Procedures?

Big Jim

Super Moderator
#81
Reference to 7.5.2 is a red herring IMHO. There is sufficent reference to use of documented procedures (confusingly called work instructions) here:

There are plenty of processes where it is possible to verify the output (so not covered by clause 7.5.2) but, for reasons of speed or cost you wouldn't want to do this on every product. instead an organisation might choose to document the production procedure as a means of providing additional control of the process.

Clause 7.5.2 only applies to processes and the requirement for is after the process has been validated. An example might help:

Welding is one of the classic special process that comes under 7.5.2. In many industries where welding is used control is exercised through use of competent people. In others it is through use of calibrated equipment and in many industries - particularly using structural or safety critical welds welding procedures are developed covering:
  • welder qualification;
  • equipment type and calibration;
  • number of welds;
  • direction and orientation of each weld pass;
  • the welding consumables to be used for each pass;
  • speed and feed rates for weld consumables to control material deposition; and
  • post weld treatment processes
The weld procedure is developed during product design and development and generally a completed weld is sectioned and destructively tested to validate that the weld will give the necessary structural strength before the weld procedure is approved and issued for production.

Of course when this procedure is issued people can choose whether they follow it or not. :sarcasm:
A better example of the need to follow a procedure for the process to come out correctly might be plating. Plating follows a specification, which may loosely be called a recipe. Proof that you follow the recipe (records of process controls) would be records of validation.

Welding using certified welders as you described probably better fits under 7.5.2 b "approval of equipment of and qualification of personnel".
 
Elsmar Forum Sponsor

Paul Simpson

Trusted Information Resource
#82
We're rapidly approaching 'I wish I hadn't bothered' territory! I should have known but ignored the signs. :nope:
A better example of the need to follow a procedure for the process to come out correctly might be plating.
I specifically chose welding 1) because it's a classic example used everywhere, and 2) because I used it to highlight alternative methods for 'arrangements for ... processes' post validation to 'achieve planned results' - but I'm happy to use your example.
Plating follows a specification, which may loosely be called a recipe. Proof that you follow the recipe (records of process controls) would be records of validation.
In the process of developing the 'recipe' the organisation will normally produce parts, carry out metallurgical sections of the plated component(s), carry out some accelerated life testing using salt spray (for example) etc. The records of these tests are the record of validation of the process.

Proof that you follow the recipe is evidence you are working to the process as validated or, to paraphrase my words above you are following 'arrangements for ... processes' to 'achieve planned results'

Welding using certified welders as you described probably better fits under 7.5.2 b "approval of equipment of and qualification of personnel".
Agreed. I try not to use the clause numbering unless I feel the need to explain where my thinking comes from. So in the example list I quoted (reproduced below) all the items would generally be included in a welding procedure / specification (under 7.5.2 c). Where the procedure includes another part of the clause is (again IMHO) irrelevant - but here they are anyway.
- welder qualification (b);
- equipment type and calibration(b);
- number of welds (c);
- direction and orientation of each weld pass (c);
- the welding consumables to be used for each pass (c);
- speed and feed rates for weld consumables to control material deposition (c); and
- post weld treatment processes (c)
  • :deadhorse: 1 There are certain situations in most organisations where inputs are turned into outputs (process) in a better way if the person / people working in the process follow a prescribed set of steps in a certain way (procedure). For ease of reference this is often captured in media in some way (documented). It may or may not be possible to verify through monitoring or measuring that the output is correct.
  • :deadhorse: 2 The simple fact that someone with the authority to document a procedure has done so means that anyone who works in the process must / shall / has to follow the procedure.
  • :deadhorse: 3 An audit of the area must / shall / has to include some assessment of whether the people in the process are following the documented procedure.
 
P

plantengineer

#83
Just my take...if you have it written in your procedure, then you have to perform it. Your procedure states how you are going to handle something, and even if it's "no longer required" by an updated revision, you still have it in YOUR procedure...you do it.

If you don't want to perform it, then you should revise your procedure as such, but if it's in, then it's done...
 
Q

Quality_Steve

#84
Thank you all for your posts.

I have learned so much from both sides of this debate.

I will begin training our auditors first to follow the process. I don't feel our auditors are seasoned enough to effectively audit the monitoring of the process, yet.

I do hope to get them there. :cfingers:

Hitting the process from both ends seems like it would produce great results.

Ideally, I would like to see our auditors follow the process and find what's working and what's not. Then bring it back to the process owner and audit them about what was found. Basically audit the monitoring of the process backwards.

That's the direction I will take for now.

:thanks:

Steve
 

Big Jim

Super Moderator
#85
We're rapidly approaching 'I wish I hadn't bothered' territory! I should have known but ignored the signs. :nope:
I specifically chose welding 1) because it's a classic example used everywhere, and 2) because I used it to highlight alternative methods for 'arrangements for ... processes' post validation to 'achieve planned results' - but I'm happy to use your example.
In the process of developing the 'recipe' the organisation will normally produce parts, carry out metallurgical sections of the plated component(s), carry out some accelerated life testing using salt spray (for example) etc. The records of these tests are the record of validation of the process.

Proof that you follow the recipe is evidence you are working to the process as validated or, to paraphrase my words above you are following 'arrangements for ... processes' to 'achieve planned results'

Agreed. I try not to use the clause numbering unless I feel the need to explain where my thinking comes from. So in the example list I quoted (reproduced below) all the items would generally be included in a welding procedure / specification (under 7.5.2 c). Where the procedure includes another part of the clause is (again IMHO) irrelevant - but here they are anyway.


  • :deadhorse: 1 There are certain situations in most organisations where inputs are turned into outputs (process) in a better way if the person / people working in the process follow a prescribed set of steps in a certain way (procedure). For ease of reference this is often captured in media in some way (documented). It may or may not be possible to verify through monitoring or measuring that the output is correct.
  • :deadhorse: 2 The simple fact that someone with the authority to document a procedure has done so means that anyone who works in the process must / shall / has to follow the procedure.
  • :deadhorse: 3 An audit of the area must / shall / has to include some assessment of whether the people in the process are following the documented procedure.
In the plating example, the plating shop usually does not develop the recipe. That would be the MIL spec or the like that the organization told them to plate to as part of the purchasing information. Someone else figured out the recipe, usually long ago.
 

Paul Simpson

Trusted Information Resource
#86
In the plating example, the plating shop usually does not develop the recipe. That would be the MIL spec or the like that the organization told them to plate to as part of the purchasing information. Someone else figured out the recipe, usually long ago.
Really? :confused:

My experience of specifications (and I confess it doesn't extend to MIL Specs for plating. :)) is that the specifier normally writes up what they want achieved:
  • required plating thickness
  • adherence of the deposited material to the substrate
  • durability - accelerated testing requirements e.g. salt spray, abrasion testing

They leave the how to the process experts. So for a component of a given size / shape / complexity the plater decides:
  • what is the best method for preparation of substrate - pretreatment
  • Anode composition
  • How long in the bath
  • key chemical(s) in the solution
  • bath temperature
  • applied voltage
  • desired current

But perhaps you could post an extract so we can compare your ideas and mine?
 

Big Jim

Super Moderator
#87
Really? :confused:

My experience of specifications (and I confess it doesn't extend to MIL Specs for plating. :)) is that the specifier normally writes up what they want achieved:
  • required plating thickness
  • adherence of the deposited material to the substrate
  • durability - accelerated testing requirements e.g. salt spray, abrasion testing

They leave the how to the process experts. So for a component of a given size / shape / complexity the plater decides:
  • what is the best method for preparation of substrate - pretreatment
  • Anode composition
  • How long in the bath
  • key chemical(s) in the solution
  • bath temperature
  • applied voltage
  • desired current

But perhaps you could post an extract so we can compare your ideas and mine?
I guess that you and I hang out at different plating shops and different customers that use their service.

It is true that the plating shop has to allow for thickness, and do the math to determine how long to leave it in the tank as well as how to set the rectifier, but the spec still provides critical parameters. In particular the key chemicals in the solution is determined by the spec, not by the plater. The plater needs to constantly monitor and adjust the tank chemistries to keep them in spec.
 

Jim Wynne

Staff member
Admin
#88
I guess that you and I hang out at different plating shops and different customers that use their service.

It is true that the plating shop has to allow for thickness, and do the math to determine how long to leave it in the tank as well as how to set the rectifier, but the spec still provides critical parameters. In particular the key chemicals in the solution is determined by the spec, not by the plater. The plater needs to constantly monitor and adjust the tank chemistries to keep them in spec.
Typical plating specs (ASTM B633 for zinc on steel, e.g.) do not specify bath chemistry.
 

Paul Simpson

Trusted Information Resource
#89
I guess that you and I hang out at different plating shops and different customers that use their service.
I don't know, Jim. You raised the topic.

It is true that the plating shop has to allow for thickness, and do the math to determine how long to leave it in the tank as well as how to set the rectifier, but the spec still provides critical parameters. In particular the key chemicals in the solution is determined by the spec, not by the plater. The plater needs to constantly monitor and adjust the tank chemistries to keep them in spec.
Again, rather than discuss the differences between my opinion and yours perhaps you could post an excerpt of one of the specs for people to have a look at. Factual approach to decision making.
 
#90
A better example of the need to follow a procedure for the process to come out correctly might be plating. Plating follows a specification, which may loosely be called a recipe. Proof that you follow the recipe (records of process controls) would be records of validation.

Welding using certified welders as you described probably better fits under 7.5.2 b "approval of equipment of and qualification of personnel".
You must hang out in different weld shops than I've experienced, Jim! It may not have appeared on your radar screen, but there's a professional qualification, called "Welding Engineer'. In my experience, these are the engineers who determine what's the most appropriate welding procedure to accomplish specific joint. This can include such diverse topics (unrelated to the actual welding equipment) as:

Materials selection
Material pre-treatment (heating annealing)
Physical geometry of the edge profiles
Weld run layouts, sequence, duration, length (to control distortion etc)
Post weld treatments, weld dressing, heat treatments etc.

I fully agree with Paul assertions. I was also heavily involved with platers and anodizers and no-where in the specifications did it give the 'recipe' as you put it, but it did describe the sequences (in chroming for example) and the relative thicknesses which may typically be achieved. In a related world, that of printed circuit manufacture, those suppliers would have thrown us out, if we'd put the recipe on the drawings - we simply stated the tin/lead ratio, thickness and such like.
 
Thread starter Similar threads Forum Replies Date
J WAIVED ON Q1 - We Don't have to comply with FORDS customer specific requirements IATF 16949 - Automotive Quality Systems Standard 2
M FULFILMENT of compliance obligation versus COMPLY with compliance obligations ISO 14001:2015 Specific Discussions 2
L Proof of Concept Studies - Do we need to comply with SAE reporting? Medical Device and FDA Regulations and Standards News 3
M MDR, RED and LVD - Should our device comply with them? EU Medical Device Regulations 3
R MDR standards - which standards to comply with ? EU Medical Device Regulations 3
M Informational Some things the EU MDR 2017/745 does not tell you, but you may need to know to comply with it effectively – Part 1 Medical Device and FDA Regulations and Standards News 0
B How to comply with IATF 16949:2016 9.3.2.1k - Management review IATF 16949 - Automotive Quality Systems Standard 2
S How to make Single Sign On (SSO) Comply e-sig requirements? ISO 13485:2016 - Medical Device Quality Management Systems 4
Z Does a website needs to comply with Part 11? Qualification and Validation (including 21 CFR Part 11) 6
Q How to comply with ISO 9001:2015 Clause 7.4 Communication ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
B Quality Policy does not include a commitment to comply with legal requirements Quality Management System (QMS) Manuals 5
S A CE Marked Product that does not comply with the Standard EU Medical Device Regulations 7
H Job Descriptions to comply with ISO Standards ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
M Medical devices are CE mark but not sold in EU - Need to comply with REACH? RoHS, REACH, ELV, IMDS and Restricted Substances 9
M How do you comply with 7.2.3 Customer Comunication AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
M Do I need to comply with both the MDD and the PED for my Medical Device? EU Medical Device Regulations 8
S Selecting materials for implants to comply with ISO 10993 biocompatibility Other Medical Device Related Standards 4
A Type of Materials to comply with IEC 60601 (Dental Laser Case) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Definition Comply - What does 'comply' mean to you? (Definition) Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
O Comply with 21 CFR 11, but no other FDA regulations? Qualification and Validation (including 21 CFR Part 11) 4
J ISO 9001 Clause 7.5.2 Validation of Processes - How to comply? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 32
A ISO/TS 16949 - Comply SPC requirements Statistical Analysis Tools, Techniques and SPC 22
K Identifying Required Testing to comply with IEC 60601 EU Medical Device Regulations 4
G What is meant by FAI (First Article Inspection) and how do we comply? AS9102 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 26
C Do all Class 1 Medical Devices (Electrical) have to comply with IEC60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
P Outsourced Manufacturing - Making Subcontractors comply with TL9000 TL 9000 Telecommunications Standard and QuEST 2
R Dielectric Strength of Triple Insulated Wire to comply with 2MOPP IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Fender1 How to comply with ISO 9001 and provide quick/short lead-time orders? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
B Commitment to Comply - ISO 14001 Clause 4.2 - Environmental Policy ISO 14001:2015 Specific Discussions 6
G Are Component Manufacturers required to comply with ISO13485:2003 ISO 13485:2016 - Medical Device Quality Management Systems 4
G How to comply with 4.4.5 Contol of Documents - Documents of External Origion Miscellaneous Environmental Standards and EMS Related Discussions 6
J Which kind of medical equipment must comply with IEC60601-1-8? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Procedure for Translating User Documentation to comply with MEDDEV 2.5/5 Other Medical Device Regulations World-Wide 14
H Proprietary Processes - How to Protect and still comply when performing an FAI? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
D How To Comply with ISO 9001 Clause 6.2.2 d (Personnel Awareness) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
P ISO 14971 - Is it a guidance document or should we fully comply with it? ISO 14971 - Medical Device Risk Management 14
J Does my company's Business Plan Contents comply with requirements of TS 16949? IATF 16949 - Automotive Quality Systems Standard 2
AnaMariaVR2 Link between failure to comply w/ Lab PPE standards & ISO9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
harrysons Automotive product transferring process what requirement to comply? IATF 16949 - Automotive Quality Systems Standard 3
A How to comply with ISO/IEC 17025 Laboratory Requirements ISO 17025 related Discussions 7
L Questions: Plastic Food Container to comply with FDA US Food and Drug Administration (FDA) 6
Q Product Realization Procedure - How to comply with ISO 9001 Clause 7.1 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
Q Suppliers Monitoring their Processes - How to comply with Clause 7.4.3.2 IATF 16949 - Automotive Quality Systems Standard 6
L How to comply with AS9100 Clause 7.6 - Monitoring and measuring devices to be used AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 31
I AS9100B Cl 7.5.1.3 - How to comply with validation of production tools requirement? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
K Quality Policy - Potential problem? Adding a Paragraph to Comply ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 27
N Requirements to Comply With FDA 21CFR820 For Invitro Diagnostic Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
N IVD Manafucturer in India - What Regulatory Requirements to Comply? ISO 13485:2016 - Medical Device Quality Management Systems 7
J How should customer complaints be handled to comply with ISO 9001? Customer Complaints 4
J How to comply with 7.2.1 - Customer Related Process ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6

Similar threads

Top Bottom