Do we need FDA Establishment Registration for Bone Screws?
RLDEARMOND
I work for a small component Medical Device manufacturer. At a recent conference our CEO was told by a consultent that because we machine bone screws that we have to have an FDA a facility license as bone screws are finished devices.
I diagree as we do not sterilize, package, label or market the screws. We only machine the screws and return them to the comapny that markets them.
I believe the consultant is just trying to get his foot in our door ($$$) but I would like a second opinion.
Coury Ferguson
Moderator here to help
Re: Component Manufacturer
Without knowing the entire situation, I tend to agree with you about the consultant. I am not a FDA expert, but there are some people here in the Cove that are.
Ajit Basrur
Re: Do we need to go in for FDA Establishment Registration ?
Welcome to the Cove :bigwave:
There is an entire section on Orthopedic Devices under 21 CFR Part 888 that you can refer - http://www.accessdata.fda.gov/scripts/cdrh/cfdocs/cfcfr/CFRSearch.cfm?CFRPart=888
Specifically, there is one medical device (Class II) under "metallic bone fixation fastener" -but not sure what your device is ?
Sec. 888.3040 Smooth or threaded metallic bone fixation fastener.
(a)Identification. A smooth or threaded metallic bone fixation fastener is a device intended to be implanted that consists of a stiff wire segment or rod made of alloys, such as cobalt-chromium-molybdenum and stainless steel, and that may be smooth on the outside, fully or partially threaded, straight or U-shaped; and may be either blunt pointed, sharp pointed, or have a formed, slotted head on the end. It may be used for fixation of bone fractures, for bone reconstructions, as a guide pin for insertion of other implants, or it may be implanted through the skin so that a pulling force (traction) may be applied to the skeletal system.
(b)Classification. Class II.
Ronen E
Problem Solver
Moderator
From your description it sounds that you may be a contract manufacturer of medical devices. There are new registration requirements coming into effect October 1st, under which you may be required to register and list (as all contract manufacturers), see here:
http://elsmar.com/Forums/showthread.php?t=55942
Cheers,
Ronen (another "greedy" consultant ;)).
PS Please note that from your description one can't tell whether you are the legal manufacturer of those devices, because you didn't say whose name is on the label. Theoretically you could be placing finished medical devices on the market, with contractors doing the packaging, labeling, sterilization and distribution for you. I'm not saying that this is your situation, nor that it's common or even likely, just pointing out that your description leaves that possibility in place (I'm mentioning it for others that may read this at a later time).
RLDEARMOND
I want to make it clear that we are a machine shop that produces many components of medical devices but we do not sterilze, package or market those components. We are just a machine shop. The customer does all the marketing, sterilization, packaging and shipping of the finished product.
If I interpret the regulation properly I don't believe the new (October 1st) regulations apply to us for facility registration.
I'm just trying to get some clarification as the regulation is vague as to who must list and register.
Ajit Basrur
If you do not make a finished device, you are exempted from Establishment Registration but I will advise you to get in touch with your customer to understand the Product Code
Ronen E
Problem Solver
Moderator
If your products go into assemblies which eventually make up finished medical devices, along with other non-packaging components, or are further worked to make them into medical devices (e.g. coated - just making up an imaginary example), then you are not considered a contract manufacturer of finished medical devices and are therefore exempt. However, if all that separates your products from being "finished" is sterilization, packaging and labeling, then there might be some scenarios (depending, among other factors, on the specific products you make) where your products could be considered "finished" just as well, and thus you'd be subject to registration requirements. For example, many dental devices are sold non-sterile, for consequent sterilization by the user prior to first use; and could also be sold (exact same device) single packed and already sterilized. I know that this is not exactly your situation, but I'm trying to make a generalized point. For an answer specific to your case one must have access to the details. You could also direct an explicit question to the FDA and get an authoritative interpretation of the regs.
Good luck,
Ronen.
MIREGMGR
No clarification of the 2013-rules boundary between a non-distributing contract manufacturer and a 21CFR 807.65(a) component maker has been provided yet.
As Ronen E notes, FDA has never made it fully clear whether a device that is "finished" but not yet ready for marketing because it hasn't been packaged, labeled and sterilized is a "finished device" or not. I hope that'll be sorted out as part of the 2013-rules transition.
Also note that under the different rules that the IRS is gradually rolling out for the 2.3% Medical Device Tax, you probably will be the manufacturer of record for the products you make as an FDA contract manufacturer or component maker, even though they are not yet packaged, labeled and sterilized, you do not distribute them and you are not FDA-responsible for them, unless the company for which you do the fabrication has a relevant patent on the mechanical or manufacturing aspects of their design. In that singular case, the company for which you do the work would be the manufacturer.
M Informational FDA Panel: Too early to pull textured breast implants over cancer risk, need more data Medical Device and FDA Regulations and Standards News 0
Leandro Need to register Medical Devices in Canada (FDA, HC, CMDCAS...) Canada Medical Device Regulations 3
P Do I need to Register with FDA? How shall my structure of processes look like? US Food and Drug Administration (FDA) 4
S Certificate of Need (CON) and FDA Requirements US Food and Drug Administration (FDA) 6
Q Need Outsourced ISO 13485/FDA Internal Auditor for a Medical Device Company Internal Auditing 3
E Do I need to notify the FDA of Management Representative changes? Other US Medical Device Regulations 3
B Need help in adding second location to FDA Registration Other US Medical Device Regulations 4
J Guidance on how often company SOPs need to be reviewed per FDA Regulations 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
C Medical Device Malfunction during misuse - Does this need to be reported to the FDA? Other US Medical Device Regulations 5
W Do 3rd party testing labs need NRTL certification to perform testing for FDA conforma ISO 13485:2016 - Medical Device Quality Management Systems 3
M Does a dental handpiece need FDA permission? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
F Conversion from Agile - Need cheaper FDA compliant software Qualification and Validation (including 21 CFR Part 11) 7
E Does European Manufacturer with ISO 13485 need QSR820 for FDA registration? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
C Do we need FDA Approval of the Manufacturing Facility and how do you go about it? US Food and Drug Administration (FDA) 5
J Start Up Medical Device Company - As a device company what FDA cert do we need? ISO 13485:2016 - Medical Device Quality Management Systems 5
S Do we need to submit our memory foam oxygen mask gasket for FDA approval? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
A Need Information - FDA Silver or Gray Sheets - August or September '06 ISO 13485:2016 - Medical Device Quality Management Systems 1
Gert Sorensen I need help regarding product responsibility in relationship to FDA ISO 13485:2016 - Medical Device Quality Management Systems 3
T Need Help on Interview Questions (FDA) Career and Occupation Discussions 10
L Turkish Requirements - Does the Software need to be translated? CE Marking (Conformité Européene) / CB Scheme 2
J Need for a cleanroom in the manufacture of a medical device for a clinical trial EU Medical Device Regulations 4
S Need help with analysing a survey on minitab Using Minitab Software 1
M IATF 16949 8.5.1.3 Verification of job set-ups - Do we need secondary check? IATF 16949 - Automotive Quality Systems Standard 7
P Electrosurgical Device User Need: Cord Flexibility -> Requirement Other Medical Device and Orthopedic Related Topics 4
P Do I need to get registered or have German entity to sell IVD products in Germany? CE Marking (Conformité Européene) / CB Scheme 2
J Documentation structure - Do I need Work Instructions? Document Control Systems, Procedures, Forms and Templates 23
G Need resources / tutorials about OPS (Operation) for ISO IT (Information Technology) Service Management 8
D Do non-IATF customers need to be included in audit scope? IATF 16949 - Automotive Quality Systems Standard 23
M Do you need an Applicable general safety and performance requirements Checklist? EU Medical Device Regulations 2
Y Does Solidworks (2D/3D drafting modules) need validation? Other Medical Device and Orthopedic Related Topics 5
M Do we need to create a new CER or can we just update the existing CER EU Medical Device Regulations 3
K A proposal for the model Quality Management - I need help for the project ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
B Record Management - Does the QMS need to control templates of records? Records and Data - Quality, Legal and Other Evidence 17
Q Need clarification on requirements.... Class i, gmp & 510(k) exempt Medical Device and FDA Regulations and Standards News 12
U Do we need clinical trial data for Class IIa medical device under MDR EU Medical Device Regulations 7
G Do we need to QA cert? We only plan to supply reagents Medical Device and FDA Regulations and Standards News 3
I MSA requirement for 5 Micrometers + CP changes need customer approval? IATF 16949 - Automotive Quality Systems Standard 2
R Evaluating the need for preventive action Preventive Action and Continuous Improvement 3
R Probability - Need a help to solve the below question Statistical Analysis Tools, Techniques and SPC 5
E In need of a new TGA sponsor - Small software company Other Medical Device Regulations World-Wide 4
8 Need Help - Runout - Function Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
B Need For BIS Standard Mark? Imported OEM's Power supply,Li-Ion Battery Other Medical Device Regulations World-Wide 0
F How many signatures do we need on calibration certificates? ISO 17025 related Discussions 8
B We need a QMS: file-based templates or software Other Medical Device Related Standards 23
M AQL table - I need to sample 1250pcs AQL - Acceptable Quality Level 3
atitheya Need of conducting medical trials in European Union EU Medical Device Regulations 2
L Gage R&R studies for identical Devices - Need to confirm the requirement to perform them Reliability Analysis - Predictions, Testing and Standards 2
D Does every piece of equipment used in a laboratory need to have an IQ protocol written and executed? ISO 13485:2016 - Medical Device Quality Management Systems 1
I IATF16949 Audit Preparation, Need "searchable" ANPQP 3.1 or latest IATF 16949 - Automotive Quality Systems Standard 1
A Does Class 1 Medical Device need to be certified to MDSAP? Canada Medical Device Regulations 5