I contacted the FDA directly and their response was that as long as we do not package, label or sterilize the device then we are exempt from registration.
Thanks to all those who gave their input.
I have the exact problem that you had . We are a machine shop that supplies medical implants and instruments to our customer's specifications. We do not label, package or sterilize our products. Yet our customer insists that, under the FDA definition, our products are considered "finished devices" - and therefore we are considered a "contract manufacturer" and are demanding we register and list with the FDA. Who did you contact at the FDA? Or where can I find the type of exemption you describe clearly stated in the FDA regulation.
Elsmar Forum Sponsor
M
MIREGMGR
#12
I have the exact problem that you had . We are a machine shop that supplies medical implants and instruments to our customer's specifications. We do not label, package or sterilize sterilize our products. Yet our customer insists that, under the FDA definition, our products are considered "finished devices" - and therefore we are considered a "contract supplier" and are demandeing we register and list with the FDA. Who did you contact at the FDA? Or where can I find the type of exemption you describe clearly stated in the FDA regulation.
The distinction between Contract Manufacturing and "component making" is defined at 21CFR 807.65.
Your customer of course can demand anything they want, and you can agree to meet their demands or not. If you choose not to meet their demands, they can take their business elsewhere. Or, you can decide that their demands are unreasonable and tell them to go elsewhere. That's the nature of business relationships. As long as you do not distribute to end users, there is no FDA requirement that determines whether your status is Contract Manufacturer or Component Maker.
The core reason why your customer might want to be Registered and to List as a Contract Manufacturer is that they are Registered and List as a Specification Provider. They however can change their Listings to Manufacturer status, in which case it's perfectly fine for them to buy from a component maker.
Some companies and consultants apparently think that a Manufacturer must perform physical manufacturing in order to be have that status. My understanding is that that belief is incorrect. The process of turning a non-medical-device object, such as the product of a component maker, into a medical device is a legal event. Who does what aspects of the physical processing--the Manufacturer, a contract manufacturer and/or a component maker--is a secondary consideration.
FDA DSMICA. 800-638-2041.
M
MIREGMGR
#13
I don't know the precise question that you put to DSMICA to which they responded, but per my understanding the key consideration is that a component maker must not either distribute to end users themselves, or provide their products to a Distributor, Reseller, Repackager or (in most cases) Kit Packer. They must provide their output only to a Manufacturer, or to a Contract Manufacturer on a Specification Developer's behalf.
The Manufacturer or Contract Manufacturer doesn't have to further physically process those objects, but the Manufacturer, Specification Provider or Contract Manufacturer does have to have legal possession of them. It is during this legal possession that the components legally become medical devices.
I'd think the most likely reason why a customer might insist that a supplier have a regulatorily responsible FDA status is that they (the customer) want to be a Distributor, and not be regulatorily responsible for the medical device products.
Someone must be regulatorily responsible for every medical device. If the maker of a product disclaims responsibility for it, but their customer is a distributor, my understanding of precedent is that FDA eventually will come after the product maker...not the distributor...even if the product maker claims that they're exempt under 21CFR 807.65(a).
T
treesei
#14
I want to make it clear that we are a machine shop that produces many components of medical devices but we do not sterilze, package or market those components. We are just a machine shop. The customer does all the marketing, sterilization, packaging and shipping of the finished product.
If I interpret the regulation properly I don't believe the new (October 1st) regulations apply to us for facility registration.
I'm just trying to get some clarification as the regulation is vague as to who must list and register.
Below is from FDA's website:
"A finished device is defined in 21 CFR 820.3(l) as any device or accessory to any device that is suitable for use or capable of functioning, whether or not it is packaged, labeled, or sterilized."
http://www.fda.gov/MedicalDevices/D...rements/QualitySystemsRegulations/default.htm
Also from FDA website (updated 10/11/2012):
"Contract Manufacturer - Manufactures a finished device to another establishment's specifications."
http://www.fda.gov/MedicalDevices/D...urDevice/RegistrationandListing/ucm053165.htm
The same website also describes who must register and list.
Y
ystark
#15
The short definition of "finished device" (and notice that it doesn't contain the word "medical") is trying to encompass thousands of devices produced throughout the world. Everything from surgical instruments, to implants or wheelchairs and crutches. Many of these must be sterilized to be "suitable for use". Some do not require sterilization at all. Is a surgical instrument or implant that is not sterilzed, yet "capable of functioning" to be considered a "finished device"?
M
MIREGMGR
#16
It must be noted, though, that notwithstanding the above definition in the law, FDA's working definition normally takes into account the intent of the maker and who they sell to.
For instance, a wood products company that bulk-fabricates flat, rounded-end wooden sticks in various sizes is not considered a medical device maker just because some sizes of those sticks are capable of being used as tongue depressors. Instead, that bulk wood products company is treated as a component maker, and their product...sold to a medical device company that does little more to the product than re-package and label it...legally becomes a medical device while in possession of that medical device company.
This would be true, in my understanding, even if the wood products company specialized in fabricating sticks that are within the size and shape ranges for tongue depressors...as long as their focus is on making wood products, and their sales of wood products that will end up as tongue depressors are to Manufacturers, Specification Providers and/or Contract Manufacturers, and not anyone else.
My view is that the OP's metal fabricator is an analogous example.
M
MIREGMGR
#17
The purpose of the FD&C Act definition is to provide a basis for the FDA to hold that someone in the realization chain for every medical device, prior to that medical device reaching an end user, is responsible for that device.
So, if you fabricate a surgical instrument and sell it in bulk to me, and I sell it to end users, either you or I is regulatorily responsible for that device. We can negotiate or fight about which of us it should be if we want, but most assuredly it will be one of the two of us. If I want to be regarded as a non-regulatorily-responsible Distributor, then you must be the Manufacturer. If you want to be regarded as a non-regulatorily-responsible component maker, then I must be the Manufacturer.
T
treesei
#18
By definition, yes.
I used to work for a company which has a facility whose products are 90% non-medical electronics (ISO 9001 QMS) and 10% medical devices (ISO 13485 QMS). It is registered and finished devices (all sold to OEM's) listed.
My take: If a machine shop makes a lot of products some of which happen to fall into the fit, form, and function of certain finished devices, as long as the products are not intended by the shop to be medical devices, the shop should not be considered a device manufacturer even company A may buy such "off-shelf" products, turn them into a device by packing, labeling (intended use comes into play here), sterilization, and assume full responsibilities. However, if the shop makes such products per company A's specifications or exclusively for company A, the shop becomes a contract manufacturer and the products become finished devices even they are only a small part in the product portfolio.
This will be an interesting situation. On one side, company A wants to mitigate risks and put some responsibility to the machine shop. On the other side, the shop wants to avoid such responsibility as much as possible.
Y
ystark
#19
I just spoke with the FDA and was given a different interpretation. These devices ARE considered finished devices and our company therefore is considered a "contract manufacturer". This means we must register and list our devices with the FDA and pay their annual fees.
Thread starter Similar threads Forum Replies Date
B Does FDA Registration QSR need to cover non-medical devices for contract repackager? US Food and Drug Administration (FDA) 1
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
M Informational FDA Panel: Too early to pull textured breast implants over cancer risk, need more data Medical Device and FDA Regulations and Standards News 0
L Need to register Medical Devices in Canada (FDA, HC, CMDCAS...) Canada Medical Device Regulations 3
P Do I need to Register with FDA? How shall my structure of processes look like? US Food and Drug Administration (FDA) 4
S Certificate of Need (CON) and FDA Requirements US Food and Drug Administration (FDA) 6
Q Need Outsourced ISO 13485/FDA Internal Auditor for a Medical Device Company Internal Auditing 3
E Do I need to notify the FDA of Management Representative changes? Other US Medical Device Regulations 3
B Need help in adding second location to FDA Registration Other US Medical Device Regulations 4
J Guidance on how often company SOPs need to be reviewed per FDA Regulations 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
C Medical Device Malfunction during misuse - Does this need to be reported to the FDA? Other US Medical Device Regulations 5
W Do 3rd party testing labs need NRTL certification to perform testing for FDA conforma ISO 13485:2016 - Medical Device Quality Management Systems 3
M Does a dental handpiece need FDA permission? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
F Conversion from Agile - Need cheaper FDA compliant software Qualification and Validation (including 21 CFR Part 11) 7
E Does European Manufacturer with ISO 13485 need QSR820 for FDA registration? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
C Do we need FDA Approval of the Manufacturing Facility and how do you go about it? US Food and Drug Administration (FDA) 5
J Start Up Medical Device Company - As a device company what FDA cert do we need? ISO 13485:2016 - Medical Device Quality Management Systems 5
S Do we need to submit our memory foam oxygen mask gasket for FDA approval? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
A Need Information - FDA Silver or Gray Sheets - August or September '06 ISO 13485:2016 - Medical Device Quality Management Systems 1
G I need help regarding product responsibility in relationship to FDA ISO 13485:2016 - Medical Device Quality Management Systems 3
T Need Help on Interview Questions (FDA) Career and Occupation Discussions 10
J Furnace repaired - Do I need a new initial TUS? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
S Need ISO 15189:2012 Documentation toolkit. Document Control Systems, Procedures, Forms and Templates 0
J Need a contract monitoring Tool General Information Resources 0
P UDI-PI requirements on reusable surgical device, do we need serialisation? ISO 13485:2016 - Medical Device Quality Management Systems 3
J Need Help with FPY Data in Assembly Process Manufacturing and Related Processes 7
W 17025 and NIST handbook relationship (need advice) ISO 17025 related Discussions 8
lanley liao Does all of the suppliers need to integrated into the supplier list qualified of the company? Oil and Gas Industry Standards and Regulations 2
K Need procedure for D&D inputs? ISO 13485:2016 - Medical Device Quality Management Systems 4
S Need help on "Country of Origin" Medical Device and FDA Regulations and Standards News 0
Ed Panek Immediate need for 80601-2-56 Consulting expert. PM me for details Career and Occupation Discussions 0
Tagin You're Gonna Need a Bigger Root Cause Coffee Break and Water Cooler Discussions 12
M PSA Suppliers - CSR matrix and need the quality manual of PSA APQP and PPAP 3
M Need Help With Information Security Asset Risk Register IEC 27001 - Information Security Management Systems (ISMS) 2
K "World Class Product" based QM. I need advice. Quality Management System (QMS) Manuals 14
I Do I need to sign off my annual audit calendar? Internal Auditing 2
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
M Do I need separation in my circuit with a medical charger? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
V Certified Auditor - Need of additional certification specific to industry ( GMPs) ASQ vs ECA vs others Professional Certifications and Degrees 1
D Low risk IVD study in the UK, do I need MHRA approval? UK Medical Device Regulations 1
J Records Control - Does each individual record need to be numbered? Records and Data - Quality, Legal and Other Evidence 2
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
J Do Software Subcontractors need to be ISO13485 compliant in the EU? EU Medical Device Regulations 3
K Do I need a "State of the art" plan? CE Marking (Conformité Européene) / CB Scheme 1
S Need advice for schooling Quality Manager and Management Related Issues 5
R What information do i need to get from the device manufacturer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
S Need guidance on project ISO 13485:2016 - Medical Device Quality Management Systems 2
H Need MSA 4th ed. compliant attribute MSA template General Measurement Device and Calibration Topics 4
J Need Change Control Yes/No Decision Tree Template ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
Similar threads
Top Bottom