Documentation for correction, corrective action, mini CAPA

psp1234

Involved In Discussions
#1
:confused:
Hi,
Correction is an action to eliminate a detected nonconformity; Corrective action is an action to eliminate the root cause. I get that. I think ;)
Is there a "mini corrective action" situation?

example (of a frequent occurrence):
A customer required a CAPA for a part that was found non-conforming. Let's say a screw in the assembly was missing (no risk to a patient, no safety issues etc.). We would do a containment (check warehouse, WIP etc...), get the part back and install the missing screw (correction) , we would also investigate the possible cause (maybe there wasn't a checklist to verify the count of screws to install), for the purpose of this example, the corrective action would be to "create a checklist to verify the count of screws etc." and of course the customer would expect verification of effectiveness. All "elements of corrective actions" are there, But - this corrective action is limited to a specific product and not really looking at possible similar issues in other products (and also the root cause efforts are minimal).
Customers typically are OK with that approach, but I had an auditor commenting about the "narrow" approach.

I do understand that other products may benefit from the same action etc., but it will be an extreme effort to think and implement this line of thinking (even though it sounds good).

In situations where a risk is identified or compliance risk is identified, the corrective action needs to be broad, that is a valid expectation.

Do any of you have experience addressing that? a "mini CA"...or "local CA"... it kind of not aligned with the essence of a CA, but very useful, considering risk vs. efforts. Would something be different in the documentation for it? like a "corrective action" with different levels and added requirements for the "BIG CA"?

Happy to hear other thoughts,
Thanks
Sue
 
Elsmar Forum Sponsor
#2
HI PSP1234 :)

"the corrective action would be to "create a checklist to verify the count of screws etc."

I believe that this is your Preventative Action, rather than your Corrective Action. the corrective is that you have replaced the component and you found during your Root Cause Analysis that it happened due to the lack of a checklist - therefore your checklist is preventing this from happening again in the future.

I think ultimately it is up to you to decide whether you would implement this change elsewhere, and you would have to identify and address where it was/was not necessary (pro's/con's/risks v likelihood etc) and it may be justifiable for you NOT to do this for all products....you just have to make sure that you document it, or it could get pulled up during an audit as having been unnoticed/ignored.
Something along the lines of stating that the risk to product conformity, usability, quality, patient etc is low; no other reports of this nature have been received; will be kept on file for trending and monitoring - then set yourself a 'deadline' to review it in the future during your CAPA reviews along the line and just check that the response is still adequate

hope this makes sense?
 

AMIT BALLAL

Super Moderator
#3
HI PSP1234 :)

"the corrective action would be to "create a checklist to verify the count of screws etc."

I believe that this is your Preventative Action, rather than your Corrective Action. the corrective is that you have replaced the component and you found during your Root Cause Analysis that it happened due to the lack of a checklist - therefore your checklist is preventing this from happening again in the future.

I think ultimately it is up to you to decide whether you would implement this change elsewhere, and you would have to identify and address where it was/was not necessary (pro's/con's/risks v likelihood etc) and it may be justifiable for you NOT to do this for all products....you just have to make sure that you document it, or it could get pulled up during an audit as having been unnoticed/ignored.
Something along the lines of stating that the risk to product conformity, usability, quality, patient etc is low; no other reports of this nature have been received; will be kept on file for trending and monitoring - then set yourself a 'deadline' to review it in the future during your CAPA reviews along the line and just check that the response is still adequate

hope this makes sense?
Thanks for your reply Kellylisa. But I beg to differ.
Once a problem is occurred, there cannot be any preventive action. Whether it is action against root cause related to occurrence, detection or system, all will become corrective actions since the problem has already occurred and we are trying to prevent recurrence.

psp1234, the purpose of horizontal deployment of corrective actions is to deploy same actions in similar product / process/ systems so that same problem can be prevented there. I understand horizontal deployment cannot be done immediately and will take time, but tracking its implementation and review will help. Develop an approach which helps horizontal deployment easy and practical to your organization in place of skipping it.
 

qualprod

Trusted Information Resource
#4
The standards says:

In one side
Corrective actions shall be appropriate to the effects of the nonconformities encountered.

So, it is not necessary to do a full analysis for every nonconformance.

For minor ncs, do a simple correction and turn the page.

But if you client asks you a CA with verification of effectiveness, well you have to hear him, even is minor.

Even in this mini CA´s is possible to spread the benefits ti other areas, according tho the standard in 10.2.1, b3

3) determining if similar nonconformities exist, or could potentially occur;

Hope this helps


Why you say mini, because doesn´t cver other nc´s
 

normzone

Trusted Information Resource
#5
Going down this very road myself at this moment, modifying some outdated process documentation and while I'm at it allowing us to take correction only with analysis of risk when appropriate.

9001:2015 references risk and opportunity, and notes productivity as an opportunity. It's not productive to do a full blown fishbone / 8D / 5 why for all issues when it's neither appropriate or necessary.

Sticking my neck out, two brand new external auditors in the house next week, we'll see how they feel about it.
 

psp1234

Involved In Discussions
#6
Hi normzone,
Although I am in the 13485, your audit results will be predictive enough...please do tell if the auditors accepted that approach!

Thanks,
Sue
 

RoxaneB

Change Agent and Data Storyteller
Super Moderator
#7
Going down this very road myself at this moment, modifying some outdated process documentation and while I'm at it allowing us to take correction only with analysis of risk when appropriate.

9001:2015 references risk and opportunity, and notes productivity as an opportunity. It's not productive to do a full blown fishbone / 8D / 5 why for all issues when it's neither appropriate or necessary.

Sticking my neck out, two brand new external auditors in the house next week, we'll see how they feel about it.
I have shared several times about a similar approach we employed in my "previous life."

To enable a consistent approach, we developed a document that identified areas and categories of known nonconformances. We identified the situations that would fall into "buckets" of 'Monitor/Track', 'Correction', 'Corrective Action.'

For example, a customer complaint under $X and associated with Product 1, 2, or 6 would fall under Correction. If there was a customer complain for Product 1,2, or 6 and > $X, then it was Corrective Action. There was no Monitoring/Tracking.

For process downtime <X, it was Monitored/Tracked. For process downtime >X but <Y, Correction. For downtime >Y, Corrective Action.

Our values were not arbitrary or selected based on a whim. We used historical data that allowed us to typically focus on 20% of our nonconformances, with this 20% typically being those situations of a more serious nature or larger magnitude.

Every year, as part of our larger management review, an analysis was conducted to determine if the triggers to determine what when into each bucked needed to be tightened. Triggers were never - from my recollection - loosened. That is to say, a corrective action for downtime might occur for all downtime > 2 hours this year. Next year, it could be > 1.5 hours, but it will not be increased to > 2.5 hours, because we are trying to stabilize our processes, not accept more variation.

This approach was not only accepted by our external auditors - ISO 9001, ISO 14001, ISRS - it was embraced by our employees, because it showed we made the standards work for us; we did not bend to fit the standards.
 

Ninja

Looking for Reality
Trusted Information Resource
#8
FWIW...and not disagreeing with any of the above (save for trying to prevent what already happened)...

When the error (nonconformance) has already touched the customer, and then back as a complaint, to then include return, rework and re-shipment...I don't see anything "mini" going on.
It may be a simple error, and a simple solution, but there are four separate items here, any one of which would have triggered my CA system.

1. Screw missing yet part approved (approval of NC part)
2. defective (nonconforming) product shipped (quality spill)
3. customer complaint
4. Return of product

I would put it through the CA system...but handle the CA as Roxanne described, based on risk/reward. It is not required that you do it that way...that's just what I would do.:2cents:
 

qualprod

Trusted Information Resource
#9
I fully agree with Ninja´s comments

Things have to be managed in a simple way.
I have tried as RoxaneB, suggested and was not as expected as we wanted, the main
problem was was the classification of the Nc´s. It was hard to people to have a full knowledge regarding where the nc fell, there were a lot of confusion for the classification, and frequently there were mistakes, because nc´s were classified in the wrong way.

Maybe there was something missing to be implemented, but that was my experience.

We changed to use a simple common sense to decide what to do, and now works better.

Hope this helps
 

psp1234

Involved In Discussions
#10
I fully agree with Ninja´s comments

Things have to be managed in a simple way.
I have tried as RoxaneB, suggested and was not as expected as we wanted, the main
problem was was the classification of the Nc´s. It was hard to people to have a full knowledge regarding where the nc fell, there were a lot of confusion for the classification, and frequently there were mistakes, because nc´s were classified in the wrong way.

Maybe there was something missing to be implemented, but that was my experience.

We changed to use a simple common sense to decide what to do, and now works better.

Hope this helps
Hi Qualprod,
Can you share what you ended up doing?
I am curious and eager to find a simple solution.

Thanks
Sue
 
Thread starter Similar threads Forum Replies Date
B Technical Documentation EU Medical Device Regulations 3
D Swiss Documentation Requirements for Class 1 Medical Device Other Medical Device Related Standards 0
M Should DoC be updated every time Technical Documentation is revised? EU Medical Device Regulations 2
P Controls over Systems Documentation in 21 CFR Part 11 Qualification and Validation (including 21 CFR Part 11) 1
D Does the DoC require a technical documentation version? ISO 13485:2016 - Medical Device Quality Management Systems 1
Y EU MDR Technical Documentation - Defined Review EU Medical Device Regulations 1
P AS9100D clause 8.6 - Documentation required to show evidence of conformity AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
E Discussion: How do you split up your Technical Documentation? EU Medical Device Regulations 4
B Technical Documentation for a medical device accessory Medical Device and FDA Regulations and Standards News 6
I Request for information regarding remote medical monitoring software (its technical documentation and the IUD system) IEC 62304 - Medical Device Software Life Cycle Processes 2
G Technical documentation for UKCA UK Medical Device Regulations 4
xfngrs Multi-lingual documentation IATF 16949 - Automotive Quality Systems Standard 6
R Configuration Management - Identifying Documentation IEC 62304 - Medical Device Software Life Cycle Processes 3
A Technical Documentation => Responsibilities when signing documents ISO 13485:2016 - Medical Device Quality Management Systems 7
A Update of Technical Documentation EU Medical Device Regulations 3
C FDA 510k documentation requirements US Food and Drug Administration (FDA) 1
M Documentation accompanying an aerospace part AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
G Seeking reference guides/ documentation /tips on verification best practices Other Medical Device Related Standards 6
P MDR documentation submissions EU Medical Device Regulations 4
dubrizo Good Documentation Practice (GDP) as it Applies to External Contractors or Vendors Document Control Systems, Procedures, Forms and Templates 4
A Updating Technical Documentation regularly => Leaner Release Advice EU Medical Device Regulations 2
P Updates to Technical Documentation EU Medical Device Regulations 7
R Good Documentation practices SOP US Food and Drug Administration (FDA) 2
S Documents belonging to Technical documentation to be stored in one file or spread in the QM system? EU Medical Device Regulations 10
O MDD 93/42 and making documentation available EU Medical Device Regulations 1
J Advice on documentation for project I am working on... ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
H Characteristics related to safety - format of documentation ISO 14971 - Medical Device Risk Management 3
H Do you repeat information throughout your documentation submission? EU Medical Device Regulations 4
J How many hours for a NB to review Technical Documentation? EU Medical Device Regulations 6
B Acquired Medical Device Product Line - Documentation Requirements for Device Master Record ISO 13485:2016 - Medical Device Quality Management Systems 7
J Technical Documentation as per MDR / subpoint 6.1 EU Medical Device Regulations 3
B UKRP to what level should you audit Class I Technical Documentation? UK Medical Device Regulations 0
A Technical documentation ce marking pippete tips CE Marking (Conformité Européene) / CB Scheme 0
S Medical Device - Technical Documentation structure EU Medical Device Regulations 1
I "Usability" for technical documentation EU Medical Device Regulations 1
Azaarus Help with CAPA for an IATF finding regarding inspection documentation IATF 16949 - Automotive Quality Systems Standard 10
E Technical documentation for IVDD General/ other CE Marking (Conformité Européene) / CB Scheme 3
B ISO13485 Standard Documentation Packages ISO 13485:2016 - Medical Device Quality Management Systems 2
C Requirement for Revision of Standard in all Documentation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
Z Data sheet from McMasterCarr enough for RoHS/REACH documentation? REACH and RoHS Conversations 4
K Addressing Missing Documentation Document Control Systems, Procedures, Forms and Templates 7
G Does FDA allows remote approvals of quality documentation. Is there any specific guidance on signing any quality records remotely? Document Control Systems, Procedures, Forms and Templates 1
S Need ISO 15189:2012 Documentation toolkit. Document Control Systems, Procedures, Forms and Templates 0
H MD Technical Documentation and SALES possibility EU Medical Device Regulations 0
A DMDIV-Variants : definition and technical documentation CE Marking (Conformité Européene) / CB Scheme 4
NDesouza COTS Items CoC for FAI Documentation AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 12
P MDR/IVDR Annex III - Technical documentation on PMS EU Medical Device Regulations 2
I IATF Lab Scope Testing Qualification and Competency Documentation IATF 16949 - Automotive Quality Systems Standard 3
C Documentation hold during submissions? ISO 13485:2016 - Medical Device Quality Management Systems 0
K SOUP (Software of Unknown Provenance) Anomaly Documentation IEC 62304 - Medical Device Software Life Cycle Processes 7

Similar threads

Top Bottom