Does 21 CFR 820.80 (b) require a Supplier to have Receiving Inspection?

M

MJW66

#1
So I have a supplier that is debating that they do not need to have a receiving inspection function to be compliant to 21CFR820.80(b).
21CFR820.80(b): Receiving acceptance activities. Each manufacturer shall establish and maintain procedures for acceptance of incoming product. Incoming product shall be inspected, tested, or otherwise verified as conforming to specified requirements. Acceptance or rejection shall be documented.

The words that are the crux of the matter are "otherwise verified." Can they successfully argue that otherwise verification can occur in-process or even at final inspection? Can a company just put everything directly to inventory without having any verification activity? That is, no data from the supplier or any inspection data performed prior to the put away function.

At first glance, I would say "No." I would at least that you would need a CoC backed up by a supplier control process that shows that the supplier is producing conforming product.

Thoughts?
 
Elsmar Forum Sponsor
M

maaquilino

#3
So I have a supplier that is debating that they do not need to have a receiving inspection function to be compliant to 21CFR820.80(b).
21CFR820.80(b): Receiving acceptance activities. Each manufacturer shall establish and maintain procedures for acceptance of incoming product. Incoming product shall be inspected, tested, or otherwise verified as conforming to specified requirements. Acceptance or rejection shall be documented.

The words that are the crux of the matter are "otherwise verified." Can they successfully argue that otherwise verification can occur in-process or even at final inspection? Can a company just put everything directly to inventory without having any verification activity? That is, no data from the supplier or any inspection data performed prior to the put away function.

At first glance, I would say "No." I would at least that you would need a CoC backed up by a supplier control process that shows that the supplier is producing conforming product.

Thoughts?
Not sure I understand this completely, but it sounds like you?re asking if a supplier that you use, or are thinking of using, needs a receiving inspection function. If they?re your suppliers, they can essentially do what they want as you are still ultimately responsible for compliance to 820. That means if you have a supplier who doesn?t have a receiving inspection procedure that meets 820 then you have to ensure YOUR acceptance activities have the appropriate acceptance controls to meet compliance. You can require that your supplier does have a compliant receiving inspection process; but you can?t force them to have one. If they won?t do it, then you have to assess whether you want to do business with them. If you want your suppliers to be compliant to 820, the time to determine whether or not they are, or if they?ll do it for you, is when you select your suppliers and negotiate your contract with them.

If you look at the preamble to 820 and do a search on ?acceptance? and one on ?supplier?, you?ll find a lot of valuable information on how the FDA thinks when it comes to these subjects. I find the preambles invaluable whether I?m drafting policies and procedures or writing project documentation; they also come in handy during negotiations.

http://www.fda.gov/MedicalDevices/D...ments/QualitySystemsRegulations/ucm230127.htm
 
M

MIREGMGR

#4
For clarity: does this supplier provide you with components, or are they a contract manufacturer or responsible manufacturer?
 

Aphel

Involved In Discussions
#5
The QSR contains requirements for the medical device manufacturer... I guess in your case - you represent the manufacturer.

So - basically it is your duty to adress the requirements derived from ?820.80...

If you involve a supplier in manufacturing your devices - or components of your devices - you need to ensure, that the supplier monitors his processes - even his outsourced processes (e.g. incoming inspection). But you cannot force him to comply with ?820.80

Best regards,
Aphel
 
#6
I have run into these word games before. If a supplier builds you an item, and does not inspect one of the components of that item as they receive them, they argue that the final product would not pass inspection if they component was faulty. This could be true, I suppose a part that had no real failure mode on its own would be "inspected during assembly" and could even be "tested at final". In effect creating a 100% inspection during assembly.
Does that qualify as "otherwise verified"? Maybe. Is it risky? Possibly.
 
M

MIREGMGR

#7
There's a semantics issue underlying this thread. Some of us are discussing whether the supplier, if what they supply is unregulated "components" as defined in 21CFR 807.65, is legally subject to 21CFR 820.80. Others are looking at the problem as a matter of good business practice. The discussion responses sound almost as if they're addressing the same issue, but they're not.
 
M

MJW66

#8
Let me try to add more clarity. This supplier is a contract manufacturer to us where we hold the design. There are approximately 5 components that are assembled into a finished unit which is then considered a component to the overall finished medical device. This component is sold to end users as a replacement component, not to the OEM of the overall device. We are not the OEM of the overall device. Thus, this component is regulated as it is being sold directly to the end user market.
 
Last edited by a moderator:
M

MJW66

#9
I have run into these word games before. If a supplier builds you an item, and does not inspect one of the components of that item as they receive them, they argue that the final product would not pass inspection if they component was faulty. This could be true, I suppose a part that had no real failure mode on its own would be "inspected during assembly" and could even be "tested at final". In effect creating a 100% inspection during assembly.
Does that qualify as "otherwise verified"? Maybe. Is it risky? Possibly.
I agree. How can "otherwise verified" listed in "820.80(b) Receiving acceptance activities:" be shown to be verified or tested later on? Later on is not within the scope of "Receiving acceptance activities" in my opinion.
 
M

MIREGMGR

#10
This supplier is a contract manufacturer to us where we hold the design.
As before, it's helpful to distinguish between usual business use of terms, and FDA language. They may be a contract manufacturer, but--if what they make is not a finished device when they ship to you, not counting labeling, packaging and if applicable, sterilization--they are not a Contract Manufacturer. A Contract Manufacturer makes finished devices. If what they make is not finished devices, they are a 21CFR807.65 Component Maker.

A Contract Manufacturer (usually, to some extent) is subject to 21CFR820. A component maker is not.

There's always an exception, and in this case, there are two. If the component they supply is not part of the original device at original manufacture but instead is an Accessory, then it is regulated as a Device. If when that part is sent out to an existing device user it's sent out as a new Accessory instead of as a replacement part, then again it's regulated as a Device. In either case, they are a Contract Manufacturer instead of a component maker. Usually, though, replacement parts are for the device itself, and therefore aren't Accessories.

You on behalf of the responsible Manufacturer/Specification Provider of course may contract with them to perform certain actions and keep certain records, but they're doing that for the responsible Manufacturer/Specification Provider, not FDA. FDA will see the responsibility for whatever they do or don't do as belonging to that Manufacturer/Specification Provider.
 
Thread starter Similar threads Forum Replies Date
M Does the Scope of 21 CFR Part 820.72 (Equipment) apply to Design? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
M Does a process map satisfy 21 CFR 820.20(d) (Quality Plan)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
M FDA 21 CFR 820.250 - Does "valid statistical" always mean math? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
S Does FDA expect suppliers to be independently assessed to 21 CFR Part 820 ? US Food and Drug Administration (FDA) 4
I Does being a subsidiary fully absolve company B from complying with 21 CFR 820? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
A Does any one have FDA 21 CFR Part 820 QSR training material? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 13
Q Does 21 CFR 820 really revise every year? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
C Does ISO 13485 certification require a company to follow 21 CFR Part 820? ISO 13485:2016 - Medical Device Quality Management Systems 13
A What does the FDA think about 13485? Will the FDA upgrade 21 CFR 820? ISO 13485:2016 - Medical Device Quality Management Systems 3
William55401 21 CFR 11 Password Aging - What does "periodically" revised mean in your org? Qualification and Validation (including 21 CFR Part 11) 3
P Does 21 CFR Part 11 apply to Scanned Documents? Other US Medical Device Regulations 12
J GMP CFR 21 Part 210 or 211 - Where does GMP address Document Control Document Control Systems, Procedures, Forms and Templates 6
J Does EN ISO 13485 necessary for Class I (non-sterile) medical devices ISO 13485:2016 - Medical Device Quality Management Systems 0
C GTIN & GS1 - what does the law say ? CE Marking (Conformité Européene) / CB Scheme 5
R Does anyone know how ASQ exam requirements are verified? Professional Certifications and Degrees 3
P Does cybersecurity apply to my product? EU Medical Device Regulations 5
Q Does a gage used for a visual verification need to be calibrated? Manufacturing and Related Processes 7
M Does any one know if company is allowed to submit appeal for rejected registration file Other Medical Device Related Standards 0
B Does TUS require load thermocouples? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
B Does anyone copyright their internal QMS documents? Document Control Systems, Procedures, Forms and Templates 24
D MDR technical file: does it need to contain duplicates of controlled documents EU Medical Device Regulations 2
D Does the MDR article 10 (14) needs to be documented word for word in a Procedure? EU Medical Device Regulations 1
S Why does GAGEpack drop the temperature and humidity readings out of the verification event card? Quality Assurance and Compliance Software Tools and Solutions 0
C Determining an as received OOT condition during calibration does not affect measurements taken prior to calibration. General Measurement Device and Calibration Topics 16
D Does the DoC require a technical documentation version? ISO 13485:2016 - Medical Device Quality Management Systems 1
S Local (country) registration of medical devices - Who does it in best case? Other Medical Device Regulations World-Wide 2
G When does containment event not require 100% sort? Nonconformance and Corrective Action 5
H Does this clause mean i must be in compliance with AS9100 or is 9001 enough? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
H Does ISO state in anyway that we MUST keep physical first piece parts? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
F Change in address for critical supplier - does NB need to be informed? CE Marking (Conformité Européene) / CB Scheme 20
S Brexit ISO 13485:2016 + Corrigendum - What does a UKCA DoC require? EU Medical Device Regulations 2
G Does anyone know about tobacco-free nicotine pouches? US Food and Drug Administration (FDA) 2
cgaro62 Does FDA apply to a non-medical 13485 certified custom manufacturing company? ISO 13485:2016 - Medical Device Quality Management Systems 11
R Does anyone use iQMS for their ISO based document control? Manufacturing and Related Processes 5
V How many hrs does it take to create a Technical File (TD)? EU Medical Device Regulations 4
M How does FDA determine OAI and VAI after inspection? ISO 13485:2016 - Medical Device Quality Management Systems 0
G Does TPI agencies comes under outsourced process as per Q1 Oil and Gas Industry Standards and Regulations 11
N FPY requested for an industry that does not have rework IATF 16949 - Automotive Quality Systems Standard 4
I CQI-15 Welding System Assessment 2nd edition - Does not require Ultrasonic Welding anymore? Customer and Company Specific Requirements 0
Ajit Basrur Does "Refurbishment" fall under "Service"? ISO 13485:2016 - Medical Device Quality Management Systems 13
T Does marketing company require CE mark if manufacturer has CE mark on product? EU Medical Device Regulations 5
Watchcat REGULATORY WATCHCAT De Novo Post-Mortem 6 – How Does This Even Happen? Other US Medical Device Regulations 0
T What does VW mean by annual requalification? Customer and Company Specific Requirements 4
I Does BSI require suppliers to be ISO 9001 Certified? EU Medical Device Regulations 12
Watchcat Does Regulatory Affairs put the "no" in innovation? Other US Medical Device Regulations 14
Q Does anyone know what this call out means? Manufacturing and Related Processes 3
K COPLANARITY: Composite profile tolerance on multiple surfaces- what does" lower dimensional reference frame tolerance" control? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
P Does FDA require certification for quality system internal audit for auditor? Qualification and Validation (including 21 CFR Part 11) 1
H Does anyone has feedback on a notified body for small companies? EU Medical Device Regulations 3

Similar threads

Top Bottom