Does 21 CFR 820.80 (b) require a Supplier to have Receiving Inspection?

M

MJW66

#1
So I have a supplier that is debating that they do not need to have a receiving inspection function to be compliant to 21CFR820.80(b).
21CFR820.80(b): Receiving acceptance activities. Each manufacturer shall establish and maintain procedures for acceptance of incoming product. Incoming product shall be inspected, tested, or otherwise verified as conforming to specified requirements. Acceptance or rejection shall be documented.

The words that are the crux of the matter are "otherwise verified." Can they successfully argue that otherwise verification can occur in-process or even at final inspection? Can a company just put everything directly to inventory without having any verification activity? That is, no data from the supplier or any inspection data performed prior to the put away function.

At first glance, I would say "No." I would at least that you would need a CoC backed up by a supplier control process that shows that the supplier is producing conforming product.

Thoughts?
 
Elsmar Forum Sponsor
M

maaquilino

#3
So I have a supplier that is debating that they do not need to have a receiving inspection function to be compliant to 21CFR820.80(b).
21CFR820.80(b): Receiving acceptance activities. Each manufacturer shall establish and maintain procedures for acceptance of incoming product. Incoming product shall be inspected, tested, or otherwise verified as conforming to specified requirements. Acceptance or rejection shall be documented.

The words that are the crux of the matter are "otherwise verified." Can they successfully argue that otherwise verification can occur in-process or even at final inspection? Can a company just put everything directly to inventory without having any verification activity? That is, no data from the supplier or any inspection data performed prior to the put away function.

At first glance, I would say "No." I would at least that you would need a CoC backed up by a supplier control process that shows that the supplier is producing conforming product.

Thoughts?
Not sure I understand this completely, but it sounds like you?re asking if a supplier that you use, or are thinking of using, needs a receiving inspection function. If they?re your suppliers, they can essentially do what they want as you are still ultimately responsible for compliance to 820. That means if you have a supplier who doesn?t have a receiving inspection procedure that meets 820 then you have to ensure YOUR acceptance activities have the appropriate acceptance controls to meet compliance. You can require that your supplier does have a compliant receiving inspection process; but you can?t force them to have one. If they won?t do it, then you have to assess whether you want to do business with them. If you want your suppliers to be compliant to 820, the time to determine whether or not they are, or if they?ll do it for you, is when you select your suppliers and negotiate your contract with them.

If you look at the preamble to 820 and do a search on ?acceptance? and one on ?supplier?, you?ll find a lot of valuable information on how the FDA thinks when it comes to these subjects. I find the preambles invaluable whether I?m drafting policies and procedures or writing project documentation; they also come in handy during negotiations.

http://www.fda.gov/MedicalDevices/D...ments/QualitySystemsRegulations/ucm230127.htm
 
M

MIREGMGR

#4
For clarity: does this supplier provide you with components, or are they a contract manufacturer or responsible manufacturer?
 

Aphel

Involved In Discussions
#5
The QSR contains requirements for the medical device manufacturer... I guess in your case - you represent the manufacturer.

So - basically it is your duty to adress the requirements derived from ?820.80...

If you involve a supplier in manufacturing your devices - or components of your devices - you need to ensure, that the supplier monitors his processes - even his outsourced processes (e.g. incoming inspection). But you cannot force him to comply with ?820.80

Best regards,
Aphel
 

hogheavenfarm

Quite Involved in Discussions
#6
I have run into these word games before. If a supplier builds you an item, and does not inspect one of the components of that item as they receive them, they argue that the final product would not pass inspection if they component was faulty. This could be true, I suppose a part that had no real failure mode on its own would be "inspected during assembly" and could even be "tested at final". In effect creating a 100% inspection during assembly.
Does that qualify as "otherwise verified"? Maybe. Is it risky? Possibly.
 
M

MIREGMGR

#7
There's a semantics issue underlying this thread. Some of us are discussing whether the supplier, if what they supply is unregulated "components" as defined in 21CFR 807.65, is legally subject to 21CFR 820.80. Others are looking at the problem as a matter of good business practice. The discussion responses sound almost as if they're addressing the same issue, but they're not.
 
M

MJW66

#8
Let me try to add more clarity. This supplier is a contract manufacturer to us where we hold the design. There are approximately 5 components that are assembled into a finished unit which is then considered a component to the overall finished medical device. This component is sold to end users as a replacement component, not to the OEM of the overall device. We are not the OEM of the overall device. Thus, this component is regulated as it is being sold directly to the end user market.
 
Last edited by a moderator:
M

MJW66

#9
I have run into these word games before. If a supplier builds you an item, and does not inspect one of the components of that item as they receive them, they argue that the final product would not pass inspection if they component was faulty. This could be true, I suppose a part that had no real failure mode on its own would be "inspected during assembly" and could even be "tested at final". In effect creating a 100% inspection during assembly.
Does that qualify as "otherwise verified"? Maybe. Is it risky? Possibly.
I agree. How can "otherwise verified" listed in "820.80(b) Receiving acceptance activities:" be shown to be verified or tested later on? Later on is not within the scope of "Receiving acceptance activities" in my opinion.
 
M

MIREGMGR

#10
This supplier is a contract manufacturer to us where we hold the design.
As before, it's helpful to distinguish between usual business use of terms, and FDA language. They may be a contract manufacturer, but--if what they make is not a finished device when they ship to you, not counting labeling, packaging and if applicable, sterilization--they are not a Contract Manufacturer. A Contract Manufacturer makes finished devices. If what they make is not finished devices, they are a 21CFR807.65 Component Maker.

A Contract Manufacturer (usually, to some extent) is subject to 21CFR820. A component maker is not.

There's always an exception, and in this case, there are two. If the component they supply is not part of the original device at original manufacture but instead is an Accessory, then it is regulated as a Device. If when that part is sent out to an existing device user it's sent out as a new Accessory instead of as a replacement part, then again it's regulated as a Device. In either case, they are a Contract Manufacturer instead of a component maker. Usually, though, replacement parts are for the device itself, and therefore aren't Accessories.

You on behalf of the responsible Manufacturer/Specification Provider of course may contract with them to perform certain actions and keep certain records, but they're doing that for the responsible Manufacturer/Specification Provider, not FDA. FDA will see the responsibility for whatever they do or don't do as belonging to that Manufacturer/Specification Provider.
 
Thread starter Similar threads Forum Replies Date
M Does the Scope of 21 CFR Part 820.72 (Equipment) apply to Design? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
M Does a process map satisfy 21 CFR 820.20(d) (Quality Plan)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
M FDA 21 CFR 820.250 - Does "valid statistical" always mean math? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
S Does FDA expect suppliers to be independently assessed to 21 CFR Part 820 ? US Food and Drug Administration (FDA) 4
I Does being a subsidiary fully absolve company B from complying with 21 CFR 820? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
A Does any one have FDA 21 CFR Part 820 QSR training material? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 13
Q Does 21 CFR 820 really revise every year? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
C Does ISO 13485 certification require a company to follow 21 CFR Part 820? ISO 13485:2016 - Medical Device Quality Management Systems 13
A What does the FDA think about 13485? Will the FDA upgrade 21 CFR 820? ISO 13485:2016 - Medical Device Quality Management Systems 3
P Does 21 CFR Part 11 apply to Scanned Documents? Other US Medical Device Regulations 12
J GMP CFR 21 Part 210 or 211 - Where does GMP address Document Control Document Control Systems, Procedures, Forms and Templates 6
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
J Records Control - Does each individual record need to be numbered? Records and Data - Quality, Legal and Other Evidence 2
lanley liao Does the customer`s trademark belong to customer-supplied property? Oil and Gas Industry Standards and Regulations 2
H How does a gas turbine work on diesel fuel? Oil and Gas Industry Standards and Regulations 12
G What does performance specification include? US Food and Drug Administration (FDA) 1
W Where does a coatings and paint company fall in IATF? IATF 16949 - Automotive Quality Systems Standard 5
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
B Does anybody know how to get older versions of Minitab to work in Windows 10? Quality Tools, Improvement and Analysis 9
M Does the ISO 9001:2015 standard require a disaster recovery plan or emergency response plan ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
C Does an accessory need an IFU if it use is discussed in the Parent device IFU? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
S How long does it take to register a product with MHRA? UK Medical Device Regulations 3
M Quality Manual - Where does Revision History Section go? Document Control Systems, Procedures, Forms and Templates 8
U Does *anyone* know a lab that will test to EN 455-4 Medical Gloves shelf life determination? EU Medical Device Regulations 1
A Brexit When does the UK responsible person need to be in place? UK Medical Device Regulations 10
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
N Does anyone have experience of GB/T 34986-2017? China Medical Device Regulations 1
Z Does anyone have experience with EN ISO 17664 ? IEC 62366 - Medical Device Usability Engineering 9
F Does anyone have an ESD quality/cooler talk to share? Training - Internal, External, Online and Distance Learning 4
A What does this line from MDCG 2020-3 (MDR art. 120 substantial change) mean to you? EU Medical Device Regulations 4
D Change Approval Requirements - Does every change need formal customer approval? Design and Development of Products and Processes 17
T What does AS9100 mean when it says you must establish a process to do X? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 24
L Does a backdate form format can be changed if wrong revision is used? Document Control Systems, Procedures, Forms and Templates 8
B General Motors and Honda Alliance - What does this mean to suppliers? IATF 16949 - Automotive Quality Systems Standard 3
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
A Does ISO 9001:2015 cover all the requirements of ISO 10012:2003? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
A Does anyone have a checklist of API Spec 650 13th Edition? Oil and Gas Industry Standards and Regulations 0
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 3
A What does this sentence "this symbol shall be used in the orientation shown" mean in ISO 780:2015? Other Medical Device Related Standards 4
L Turkish Requirements - Does the Software need to be translated? CE Marking (Conformité Européene) / CB Scheme 2
R Where does IATF 16949 address Process mapping? IATF 16949 - Automotive Quality Systems Standard 3
J Does Pakistan Medical Device Import License allows parallel import? Other Medical Device Regulations World-Wide 0
BeaBea Interesting Discussion Where Does Marketing/ Advertisement of Products fit in to ISO 9001? Process Maps, Process Mapping and Turtle Diagrams 39
P Does anyone have a API Q1 Documentation Package? Quality Management System (QMS) Manuals 1
N What is our product classification? (Does Unclassified classification still exists) Other US Medical Device Regulations 14
C Does a CE mark infer meeting all applicable standards? CE Marking (Conformité Européene) / CB Scheme 4
N Small Company - Internal audit process - Who does the audit? Internal Auditing 16
J Does anyone have an excel IATF 16949 Internal Audit checklist I could use? IATF 16949 - Automotive Quality Systems Standard 7
A Does AS9100 require traceability to operators performing the work? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4