Section 4.5 in the General Standard 60601-1 says: "Where this standard specifies a particular RISK CONTROL measure or test method, an alternative RISK CONTROL measure or method is acceptable, provided ...."

Does this also apply to the Particular Standards? If so, where is this stated?

Thanks!
 
Elsmar Forum Sponsor

Peter Selvey

Leader
Super Moderator
#2
Yes, it applies to particular standards.

The structure of particular standards is that they modify the general standard. It is not actually correct to read the standards separately or to consider them as two standards, rather it is the single merged version which is "legally" correct.

When I first started working in this field (35 years ago!) the test lab would prepare reports this way, using a merged version. However, the CB scheme got lazy and started publishing separate protocols. Admittedly it's a lot of work, standards have got a lot bigger, more particulars, amendments etc. Nevertheless, it can lead to errors if the standards are read and evaluated separately.
 

Tidge

Trusted Information Resource
#3
I worked on an ME device with a particular standard that (successfully) leveraged clause 4.5; although in our case the clause was specifically invoked for one small part of a collateral standard. Without going into detail: for the type of device in question there is a clause in a collateral standard (not addressed in the particular) that appears to exist because of some underlying assumption about the way all medical devices are used, but in practice does not apply to a majority of devices in a surgical theater. There is a lot of literature on the subject, our own formative testing revealed that complying with the clause was not only against established clinical practice but was interfering with task performance and generally making performing critical tasks very difficult. In other words: there was a requirement of the standard that was not really reducing any risks but was introducing (and increasing) risks.

It wasn't easy to get the NRTL to accept the alternate risk control method. An ironic element of this story is that despite us calling out that we were invoking clause 4.5, the NRTL missed it. We had to circle back to them to get them to recognize it because (a) it was the right thing to do and (b) we didn't want to risk a "last minute" test failure showing up... but it then became a nightmare to get them to accept the volumes of data (including a significant amount of user validation and published literature that defended the alternate risk control as more effective than the one in the collateral. We already had all of this in the DHF and it was all referenced within our risk files, but they all but made us repeat validation.

I was left with the impression that the NRTL took this approach with us for a mix of reasons:
  • Human factors: They were embarrassed at having missed something we called out for them on the checklist.
  • Work Experience: Despite the (now) long history of requiring risk management as the basis of 3rd edition, NRTLs have no practical experience with actually working through Risk Management. I literally had to explain what "verification of implementation" and "verification of effectiveness" meant.
  • An inherent suspicion of any ME manufacturer who can't/won't follow a standard (for any reason, or no reason)
This interaction really soured me on the NRTL, because the long-standing test lead (who used to instruct the class on 3rd edition compliance) explained in the week-long class what was necessary to support invocation of clause 4.5... and as part of the class would point to certain 'basic safety' elements of the standard as being areas where the standard didn't quite make sense and so strict compliance wasn't necessary (with explanation). I found it extremely frustrating that he'd casually accept alternate risk controls when energy was involved but in terms of usability he was worse than a novice.
 
Thread starter Similar threads Forum Replies Date
C GTIN & GS1 - what does the law say ? CE Marking (Conformité Européene) / CB Scheme 5
R Does anyone know how ASQ exam requirements are verified? Professional Certifications and Degrees 3
P Does cybersecurity apply to my product? EU Medical Device Regulations 5
Q Does a gage used for a visual verification need to be calibrated? Manufacturing and Related Processes 7
M Does any one know if company is allowed to submit appeal for rejected registration file Other Medical Device Related Standards 0
B Does TUS require load thermocouples? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
B Does anyone copyright their internal QMS documents? Document Control Systems, Procedures, Forms and Templates 24
D MDR technical file: does it need to contain duplicates of controlled documents EU Medical Device Regulations 2
D Does the MDR article 10 (14) needs to be documented word for word in a Procedure? EU Medical Device Regulations 1
S Why does GAGEpack drop the temperature and humidity readings out of the verification event card? Quality Assurance and Compliance Software Tools and Solutions 0
C Determining an as received OOT condition during calibration does not affect measurements taken prior to calibration. General Measurement Device and Calibration Topics 16
D Does the DoC require a technical documentation version? ISO 13485:2016 - Medical Device Quality Management Systems 1
S Local (country) registration of medical devices - Who does it in best case? Other Medical Device Regulations World-Wide 2
G When does containment event not require 100% sort? Nonconformance and Corrective Action 5
H Does this clause mean i must be in compliance with AS9100 or is 9001 enough? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
H Does ISO state in anyway that we MUST keep physical first piece parts? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
F Change in address for critical supplier - does NB need to be informed? CE Marking (Conformité Européene) / CB Scheme 20
S Brexit ISO 13485:2016 + Corrigendum - What does a UKCA DoC require? EU Medical Device Regulations 2
G Does anyone know about tobacco-free nicotine pouches? US Food and Drug Administration (FDA) 2
cgaro62 Does FDA apply to a non-medical 13485 certified custom manufacturing company? ISO 13485:2016 - Medical Device Quality Management Systems 11
R Does anyone use iQMS for their ISO based document control? Manufacturing and Related Processes 5
V How many hrs does it take to create a Technical File (TD)? EU Medical Device Regulations 4
M How does FDA determine OAI and VAI after inspection? ISO 13485:2016 - Medical Device Quality Management Systems 0
G Does TPI agencies comes under outsourced process as per Q1 Oil and Gas Industry Standards and Regulations 11
N FPY requested for an industry that does not have rework IATF 16949 - Automotive Quality Systems Standard 4
William55401 21 CFR 11 Password Aging - What does "periodically" revised mean in your org? Qualification and Validation (including 21 CFR Part 11) 3
I CQI-15 Welding System Assessment 2nd edition - Does not require Ultrasonic Welding anymore? Customer and Company Specific Requirements 0
Ajit Basrur Does "Refurbishment" fall under "Service"? ISO 13485:2016 - Medical Device Quality Management Systems 13
T Does marketing company require CE mark if manufacturer has CE mark on product? EU Medical Device Regulations 5
Watchcat REGULATORY WATCHCAT De Novo Post-Mortem 6 – How Does This Even Happen? Other US Medical Device Regulations 0
T What does VW mean by annual requalification? Customer and Company Specific Requirements 4
I Does BSI require suppliers to be ISO 9001 Certified? EU Medical Device Regulations 12
Watchcat Does Regulatory Affairs put the "no" in innovation? Other US Medical Device Regulations 14
Q Does anyone know what this call out means? Manufacturing and Related Processes 3
K COPLANARITY: Composite profile tolerance on multiple surfaces- what does" lower dimensional reference frame tolerance" control? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
P Does FDA require certification for quality system internal audit for auditor? Qualification and Validation (including 21 CFR Part 11) 1
H Does anyone has feedback on a notified body for small companies? EU Medical Device Regulations 3
E Does IVD need to integrate with hospital IT infrastructure? Medical Information Technology, Medical Software and Health Informatics 2
D What does a level 1 (PSW) PPAP actually promise? APQP and PPAP 19
R GS1 DataMatrix code does not meet GS1 Specification EU Medical Device Regulations 0
M Does the scope of ISO 9001:2015 applies to tenders, pricing and sales department of a medical devices distributor? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
K Why does load cell supplier requires force verification General Measurement Device and Calibration Topics 3
M Does C=0 strictly mean 1 bad, all bad, all the time? ISO 13485:2016 - Medical Device Quality Management Systems 6
C Does a medical device active (zinc oxide) needs BPR registration in EU? Other ISO and International Standards and European Regulations 6
D Does Risk Management apply to re-labeler (MDR) EU Medical Device Regulations 1
Ed Panek Does this FDA Requirement Apply to international (not USA) distributors for USA based manufacturing companies? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
S Does anyone have a checklist to prepare for ISO 13485, Stage I audit? ISO 13485:2016 - Medical Device Quality Management Systems 3
S Does a refurbished product required a new UDI? US Food and Drug Administration (FDA) 3
D Change to labelling - does it require a new 510(k)? US Food and Drug Administration (FDA) 5

Similar threads

Top Bottom