We have an accessory for a non-electrical medical device (the accessory has a circuit board, LEDs and solenoids etc) and is intended to be fitted in an ambulance and powered by the vehicle’s 12V DC supply. Is there a requirement for such a device to be tested to IEC 60601-1-2 (EMC), or is this standard intended for mains driven devices? I get a serious pain in the brain trying to understand/interpret the 60601-1 series standards.

Steve
 
Elsmar Forum Sponsor
G

George Weiss

#3
IEC 60601-1-2 is a base standard referring to other standards for EMC testing. The Standard mainly is referencing medical devices, specifically ones which could impact the patient and other equipment. Assuming your device is clearly classed as a non-medical device, then IEC 60601-1-2 testing is not required, but only maybe desired. 3.201.4 Non-medical equipment of 60601-1-2, suggests your device does not require EMC compliance.
For other discussion, I reference an article by a respected EMC lab. @
http://www.elliottlabs.com/documents/IEC%2060601%20for%20Conformity%20DEC%202005.pdf
A reference document I have suggests that a permanently affixed device in a vehicle/ambulance could qualify it as medical.
 
S

SteveK

#4
Hi George,

Do not think you read my post correctly. The 12V device is an accessory to a medical device (non-electrical). It does not have to be fiitted in an ambulance, but it is an option. Under MDD 93/42/EEC accessories are themselves classed as medical devices, thus my concern about EMC testing or non-testing requirements.

Steve
 
S

Sjoerd

#5
We have an accessory for a non-electrical medical device (the accessory has a circuit board, LEDs and solenoids etc) and is intended to be fitted in an ambulance and powered by the vehicle’s 12V DC supply. Is there a requirement for such a device to be tested to IEC 60601-1-2 (EMC), or is this standard intended for mains driven devices? I get a serious pain in the brain trying to understand/interpret the 60601-1 series standards.

Steve
From the rationale A of 60601-1-2:2007

"This collateral standard is applicable to ME EQUIPMENT and ME SYSTEMS. For the purposes of this collateral standard, an ME SYSTEM includes those ACCESSORIES that are needed for operating the ME SYSTEM as specified by the MANUFACTURER. The reader is reminded that an ME SYSTEM can consist of ME EQUIPMENT used in combination with other ME EQUIPMENT or with other electrical equipment that is not ME EQUIPMENT."

If a device is mains driven is not relevant, immunity, ESD and radiated-emmision can be of intrest for battery driven devices, however some tests can be skipped (for instance surge immunity).

Apart from this, if your accessory would not be medical, it is possibly no longer excluded in 2004-108-EEC, and you can end up still doing EMC tests.
 
Thread starter Similar threads Forum Replies Date
Ed Panek Does this FDA Requirement Apply to international (not USA) distributors for USA based manufacturing companies? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
S Does anyone have a checklist to prepare for ISO 13485, Stage I audit? ISO 13485:2016 - Medical Device Quality Management Systems 1
S Does a refurbished product required a new UDI? US Food and Drug Administration (FDA) 0
D Change to labelling - does it require a new 510(k)? US Food and Drug Administration (FDA) 5
M What does "constantly" mean ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
N Does anyone use SGS for ISO 13485 / CE certification Registrars and Notified Bodies 0
G Does FDA allows remote approvals of quality documentation. Is there any specific guidance on signing any quality records remotely? Document Control Systems, Procedures, Forms and Templates 1
B Does FDA Registration QSR need to cover non-medical devices for contract repackager? US Food and Drug Administration (FDA) 1
lanley liao Does all of the suppliers need to integrated into the supplier list qualified of the company? Oil and Gas Industry Standards and Regulations 2
0 To which part of 13485 does this refer? ISO 13485:2016 - Medical Device Quality Management Systems 3
A Medical Device Contract Manufacturer - Does the CM need to register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
J Records Control - Does each individual record need to be numbered? Records and Data - Quality, Legal and Other Evidence 2
lanley liao Does the customer`s trademark belong to customer-supplied property? Oil and Gas Industry Standards and Regulations 2
H How does a gas turbine work on diesel fuel? Oil and Gas Industry Standards and Regulations 12
G What does performance specification include? US Food and Drug Administration (FDA) 1
W Where does a coatings and paint company fall in IATF? IATF 16949 - Automotive Quality Systems Standard 5
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1
B Does anybody know how to get older versions of Minitab to work in Windows 10? Quality Tools, Improvement and Analysis 9
M Does the ISO 9001:2015 standard require a disaster recovery plan or emergency response plan ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
C Does an accessory need an IFU if it use is discussed in the Parent device IFU? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
S How long does it take to register a product with MHRA? UK Medical Device Regulations 8
M Quality Manual - Where does Revision History Section go? Document Control Systems, Procedures, Forms and Templates 8
U Does *anyone* know a lab that will test to EN 455-4 Medical Gloves shelf life determination? EU Medical Device Regulations 1
A Brexit When does the UK responsible person need to be in place? UK Medical Device Regulations 10
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
N Does anyone have experience of GB/T 34986-2017? China Medical Device Regulations 1
Z Does anyone have experience with EN ISO 17664 ? IEC 62366 - Medical Device Usability Engineering 9
F Does anyone have an ESD quality/cooler talk to share? Training - Internal, External, Online and Distance Learning 4
A What does this line from MDCG 2020-3 (MDR art. 120 substantial change) mean to you? EU Medical Device Regulations 4
D Change Approval Requirements - Does every change need formal customer approval? Design and Development of Products and Processes 17
T What does AS9100 mean when it says you must establish a process to do X? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 24
L Does a backdate form format can be changed if wrong revision is used? Document Control Systems, Procedures, Forms and Templates 8
B General Motors and Honda Alliance - What does this mean to suppliers? IATF 16949 - Automotive Quality Systems Standard 3
C ISO 13485 :2016 - CAPA - Does every CAPA need to be checked by regulations? ISO 13485:2016 - Medical Device Quality Management Systems 9
A Does ISO 9001:2015 cover all the requirements of ISO 10012:2003? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
N FDA UDI - Label vs. Labeling - Does the insert need to include UDI? Other US Medical Device Regulations 1
A Does anyone have a checklist of API Spec 650 13th Edition? Oil and Gas Industry Standards and Regulations 0
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 3
A What does this sentence "this symbol shall be used in the orientation shown" mean in ISO 780:2015? Other Medical Device Related Standards 4
L Turkish Requirements - Does the Software need to be translated? CE Marking (Conformité Européene) / CB Scheme 2
R Where does IATF 16949 address Process mapping? IATF 16949 - Automotive Quality Systems Standard 3
J Does Pakistan Medical Device Import License allows parallel import? Other Medical Device Regulations World-Wide 0
BeaBea Interesting Discussion Where Does Marketing/ Advertisement of Products fit in to ISO 9001? Process Maps, Process Mapping and Turtle Diagrams 39
P Does anyone have a API Q1 Documentation Package? Quality Management System (QMS) Manuals 1
N What is our product classification? (Does Unclassified classification still exists) Other US Medical Device Regulations 14
C Does a CE mark infer meeting all applicable standards? CE Marking (Conformité Européene) / CB Scheme 4
N Small Company - Internal audit process - Who does the audit? Internal Auditing 16
J Does anyone have an excel IATF 16949 Internal Audit checklist I could use? IATF 16949 - Automotive Quality Systems Standard 7
A Does AS9100 require traceability to operators performing the work? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
I Does anybody use Detection in medical device Design FMEA? ISO 14971 - Medical Device Risk Management 18

Similar threads

Top Bottom