We have an accessory for a non-electrical medical device (the accessory has a circuit board, LEDs and solenoids etc) and is intended to be fitted in an ambulance and powered by the vehicle’s 12V DC supply. Is there a requirement for such a device to be tested to IEC 60601-1-2 (EMC), or is this standard intended for mains driven devices? I get a serious pain in the brain trying to understand/interpret the 60601-1 series standards.
Steve
From the rationale A of 60601-1-2:2007
"This collateral standard is applicable to ME EQUIPMENT and ME SYSTEMS. For the purposes of this collateral standard, an ME SYSTEM includes those ACCESSORIES that are needed for operating the ME SYSTEM as specified by the MANUFACTURER. The reader is reminded that an ME SYSTEM can consist of ME EQUIPMENT used in combination with other ME EQUIPMENT or with other electrical equipment that is not ME EQUIPMENT."
If a device is mains driven is not relevant, immunity, ESD and radiated-emmision can be of intrest for battery driven devices, however some tests can be skipped (for instance surge immunity).
Apart from this, if your accessory would not be medical, it is possibly no longer excluded in 2004-108-EEC, and you can end up still doing EMC tests.