Does IEC 60601-2-49 apply to Thoracic Bioimpedance?
craiglab
Involved In Discussions
#1
IEC 60601-2-49 Ed. 2.0 b:2011 has in the description “The scope of this standard is restricted to medical electrical equipment intended for connection to a single patient that has either two or more applied parts or multiple functions on an applied part.”
The device we are considering is similar to thoracic bioimpedance where you have 2 patient auxiliary circuits and 4+ monitoring leads. It is not clear if the auxiliary circuits and monitoring are each considered as seperate applied parts, and if this standard applies.
Elsmar Forum Sponsor
Peter Selvey
Staff member
Moderator
#2
Is the equipment intended to generate alarms? I don't have the 2011 edition yet, but the older edition (2001) only applies if there are alarms.
To understand any scope statement, always read the definitions used in the scope statement (and in some cases the definitions used in those definitions).
An "alarm" is not fully defined but normally means at least medium or high priority alarms as per IEC 60601-1-8, i.e. not just a message.
If there are alarms, then it seems the standard is intended to apply when there are at least two different functions that can trigger an alarm (e.g. one from ECG, one from SpO2). If the patient connections are all related to the same "function", then the standard would not apply (keep in mind though, IEC 60601-1-8 still applies).
R
raymond02302
#3
As you described, I imagine that the equipment is just for thoracic bioimpedance. the 2 patient auxiliary circuit and 4 monitoring leads are all for this clinical use. 1 circuit and 2 leads are used to generate the HF current and the other circuit and leads are for measuring the impedance.
If so, IEC 60601-2-49 is not applicable, as the euipment is just one applied part and only one physiological monitoring unit.
Thread starter Similar threads Forum Replies Date
shimonv IEC 60601-1-8 - When does an error message become an alarm signal? IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
MDD_QNA IEC 60601-1 Ed2 vs Ed3 - Does a product that conforms to Ed3 also fulfill Ed2? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-27: What does "linear within ±20% of the full scale output" mean? IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
E What does "other APPLIED PARTS" mean in IEC 60601-1 ed 3.0 8.7.4.7 b) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R Does long-term body worn medical monitoring device need to meet IEC 60601-1-11? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
S IEC 60601-1 3rd Gap analysis template TO BUY - Does anyone know where? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D Will buyers (hospitals) require IEC 60601, even when FDA does not? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R Does the Banana connector deviate from the IEC 60601-2-25? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
SteveK Does a 12V DC device need EMC (IEC 60601-1-2) Testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A Does Canade recognize IEC 60601-2-22? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
MDD_QNA QR Code Standard ISO/IEC 15417:2007 - Does anyone use it? Other Medical Device Related Standards 3
S Does IEC 62304 require documenting unresolved anomalies for all safety classes? IEC 62304 - Medical Device Software Life Cycle Processes 4
M Does Calibration to ISO/IEC 17025 conform to Z540.3? ISO 17025 related Discussions 1
M Does ISO/IEC 17021 Accreditation work for 13485 ISO 13485:2016 - Medical Device Quality Management Systems 2
L Does IEC 61010-1 apply for Surgical Microscopes? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Does ISO/IEC 17025:2005 require the Laboratory to have separate ISO 9001 Procedures ISO 17025 related Discussions 4
T Does anybody know about EN45014? (Replaced by ISO/IEC 17050-1) Other Medical Device Related Standards 1
A Sampling - What does it mean by Sampling in ISO:IEC 17025 ISO 17025 related Discussions 1
R Does ISO/IEC 17025 supersede ANSI/NCSL Z540 ISO 17025 related Discussions 6
A Does anybody have an audit check list for ISO/IEC 17025 to replace ISO/IEC Guide 25 ISO 17025 related Discussions 14
BeaBea Interesting Discussion Where Does Marketing/ Advertisement of Products fit in to ISO 9001? Process Maps, Process Mapping and Turtle Diagrams 35
P Does anyone have a API Q1 Documentation Package? Quality Management System (QMS) Manuals 1
N What is our product classification? (Does Unclassified classification still exists) Other US Medical Device Regulations 14
C Does a CE mark infer meeting all applicable standards? CE Marking (Conformité Européene) / CB Scheme 1
N Small Company - Internal audit process - Who does the audit? Internal Auditing 4
J Does anyone have an excel IATF 16949 Internal Audit checklist I could use? IATF 16949 - Automotive Quality Systems Standard 7
A Does AS9100 require traceability to operators performing the work? AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 4
I Does anybody use Detection in medical device Design FMEA? ISO 14971 - Medical Device Risk Management 18
A EN ISO 14971:2019 does not include the Annex Zs ISO 14971 - Medical Device Risk Management 4
Watchcat Does "Similar Device" = "Predicate"? EU Medical Device Regulations 7
C Document Control Stamps - Does anyone still stamp their documents? Document Control Systems, Procedures, Forms and Templates 21
Y Does Solidworks (2D/3D drafting modules) need validation? Other Medical Device and Orthopedic Related Topics 5
M Definition Open Audit - What does an Open Audit mean? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 3
I Does training have to be written? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
P What does UPH/m² mean? Quality Tools, Improvement and Analysis 3
L "IATF-Compliant" IATF 16949:2016 certification? What does this mean? IATF 16949 - Automotive Quality Systems Standard 13
D Why does Official Journal list superseded standards? EU Medical Device Regulations 6
B Record Management - Does the QMS need to control templates of records? Records and Data - Quality, Legal and Other Evidence 17
R Does any here use an internal auditing tool that works on different platforms? Internal Auditing 3
Sidney Vianna IAQG SCMH explains "positive risk"..........but does it? AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 3
D Where does "as far as possible" stop? FMEA - EN 14971 ISO 14971 - Medical Device Risk Management 24
W Does anyone have an API Q2 checklist for internal auditing? Oil and Gas Industry Standards and Regulations 1
G Does pitch/increment/resolution of a ruled scale apply to measurement uncertainty as line item? Measurement Uncertainty (MU) 10
L External power supplies: How close does the safety report have to match the end-use application? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
Watchcat Does 820.30 include the manufacturing process? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Watchcat Does ISO 13485 7.3 include the manufacturing process? ISO 13485:2016 - Medical Device Quality Management Systems 14
R When does the FDA consider a component a medical device? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 17
B Does EN ISO 15223-1:2016 include the graphic symbols to be added to software and IFU? Other Medical Device Related Standards 9
DitchDigger Boston Fire Code: Who Does What? (CAL TB 117-2013; CAL TB 133) Occupational Health & Safety Management Standards 2
GreatNate Metrotom - Does anyone have any exposure to the Zeiss Metrotom 800 or 1500? Manufacturing and Related Processes 0
Similar threads
Top Bottom