We have a scenario in which we could use some input and guidance:
1. "Company A" is located in the US. They have a CE mark for a sterile set of products.
2. "Company A" wishes to market under "Brand X" in the EU.
3. "Company B" is the manufacturer in India that has a CE Mark for the devices.
4. These devices will be either class Ia or IIa sterile in the EU.
Under MDR, can "Company A" market the devices under "Brand X" using the "Company B" CE mark and technical file under a documented quality agreement? If so, how does the label need to be structured? It is our suspicion that this arrangement won't be able to be done this way and that "Company A" will need to get their own CE mark. Thoughts?
Elsmar Forum Sponsor
Ed Panek
QA RA Small Med Dev Company
Leader
Super Moderator
#2
I would start with the MDR Label requirements. Labels require the Manufacturer and Authorized rep on labels I think. Also consider how eudamed will handle the economic operators.
Other than that is Company A the designer of the device? What do they do? If they are similar to a distributor or the like relabelling is acceptable as long as the label has the man of record and Auth Rep listed.
#3
Thank you for your reply. In answer to your questions:
Company A is the designer and distributor. They would like to add their own brand name without needing to accept the obligations of a manufacturer (ie obtain their own CE mark for the device). It's our view that would not be necessary if they are able to establish an agreement with the manufacturer regarding the packaging, all of the parties are identified on the label, and the integrity of the product is not compromised if repackaging is necessary.
Do you concur?
Ronen E
Problem Solver
Moderator
#4
The above is a little confusing for me, so rather than trying to accurately understand "who does what" in this case, let me try and state some basics that seem relevant.
In MDR terms "Manufacturer" is the entity that bears the regulatory responsibility. Who physically makes the device (and where) is less relevant than the name on the device label (if more than one entity is there, it must be very clear who is the designated Manufacturer). Who the Manufacturer is is a matter of a conscious decision and it has serious ramifications.
If the Manufacturer on the label is different from what was previously cleared, or if any significant change to the label is made (e.g. the device name is changed), a "new CE mark" is required. More accurately, a new clearance process is required.
New branding or new packaging usually means taking full responsibility - it's a new product, regardless of the level of similarity to an existing/cleared one. The existing test data may be leveraged though (through reuse, not through a blanket pass), to a varying extent.
levatorsuperioris
Involved In Discussions
#5
For the Original OEM PLM relationship under the MDR, the PLM is required to have complete and current access to the technical file... This is not simply proven with a quality agreement, your NB will likely require you to prove you can get it immediately and completely... What you describe is one of the biggest changes to the MDD-MDR transition.
Junn1992
Quite Involved in Discussions
#6
If you want the product to bear the brand name of company A, then company A is the manufacturer under the MDR.
If you want company A to sell the product only, company A is considered a distributor, but it CANNOT have its brand on the product. The product brand will have to be held by the company with the CE mark, which in this case is your OEM.
Definition from the MDR:
(30)‘manufacturer’ means a natural or legal person who manufactures or fully refurbishes a device or has a device designed, manufactured or fully refurbished, and markets that device under its name or trademark;
There you go.
Thread starter Similar threads Forum Replies Date
BeaBea Interesting Discussion Where Does Marketing/ Advertisement of Products fit in to ISO 9001? Process Maps, Process Mapping and Turtle Diagrams 39
K Does ISO 13485:2016 - 7.2.3 Customer Communication include Marketing? ISO 13485:2016 - Medical Device Quality Management Systems 1
S Does anyone have a Marketing Intelligence template? Document Control Systems, Procedures, Forms and Templates 1
K Where and How Does Marketing Fit In? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
C GTIN & GS1 - what does the law say ? CE Marking (Conformité Européene) / CB Scheme 5
R Does anyone know how ASQ exam requirements are verified? Professional Certifications and Degrees 3
P Does cybersecurity apply to my product? EU Medical Device Regulations 5
Q Does a gage used for a visual verification need to be calibrated? Manufacturing and Related Processes 7
M Does any one know if company is allowed to submit appeal for rejected registration file Other Medical Device Related Standards 0
B Does TUS require load thermocouples? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
B Does anyone copyright their internal QMS documents? Document Control Systems, Procedures, Forms and Templates 24
D MDR technical file: does it need to contain duplicates of controlled documents EU Medical Device Regulations 2
D Does the MDR article 10 (14) needs to be documented word for word in a Procedure? EU Medical Device Regulations 1
S Why does GAGEpack drop the temperature and humidity readings out of the verification event card? Quality Assurance and Compliance Software Tools and Solutions 0
C Determining an as received OOT condition during calibration does not affect measurements taken prior to calibration. General Measurement Device and Calibration Topics 16
D Does the DoC require a technical documentation version? ISO 13485:2016 - Medical Device Quality Management Systems 1
S Local (country) registration of medical devices - Who does it in best case? Other Medical Device Regulations World-Wide 2
G When does containment event not require 100% sort? Nonconformance and Corrective Action 5
H Does this clause mean i must be in compliance with AS9100 or is 9001 enough? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
H Does ISO state in anyway that we MUST keep physical first piece parts? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
F Change in address for critical supplier - does NB need to be informed? CE Marking (Conformité Européene) / CB Scheme 20
S Brexit ISO 13485:2016 + Corrigendum - What does a UKCA DoC require? EU Medical Device Regulations 2
G Does anyone know about tobacco-free nicotine pouches? US Food and Drug Administration (FDA) 2
cgaro62 Does FDA apply to a non-medical 13485 certified custom manufacturing company? ISO 13485:2016 - Medical Device Quality Management Systems 11
R Does anyone use iQMS for their ISO based document control? Manufacturing and Related Processes 5
V How many hrs does it take to create a Technical File (TD)? EU Medical Device Regulations 4
M How does FDA determine OAI and VAI after inspection? ISO 13485:2016 - Medical Device Quality Management Systems 0
G Does TPI agencies comes under outsourced process as per Q1 Oil and Gas Industry Standards and Regulations 11
N FPY requested for an industry that does not have rework IATF 16949 - Automotive Quality Systems Standard 4
William55401 21 CFR 11 Password Aging - What does "periodically" revised mean in your org? Qualification and Validation (including 21 CFR Part 11) 3
I CQI-15 Welding System Assessment 2nd edition - Does not require Ultrasonic Welding anymore? Customer and Company Specific Requirements 0
Ajit Basrur Does "Refurbishment" fall under "Service"? ISO 13485:2016 - Medical Device Quality Management Systems 13
Watchcat REGULATORY WATCHCAT De Novo Post-Mortem 6 – How Does This Even Happen? Other US Medical Device Regulations 0
T What does VW mean by annual requalification? Customer and Company Specific Requirements 4
I Does BSI require suppliers to be ISO 9001 Certified? EU Medical Device Regulations 12
Watchcat Does Regulatory Affairs put the "no" in innovation? Other US Medical Device Regulations 14
Q Does anyone know what this call out means? Manufacturing and Related Processes 3
K COPLANARITY: Composite profile tolerance on multiple surfaces- what does" lower dimensional reference frame tolerance" control? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
P Does FDA require certification for quality system internal audit for auditor? Qualification and Validation (including 21 CFR Part 11) 1
H Does anyone has feedback on a notified body for small companies? EU Medical Device Regulations 3
M Does 4.5 - Alternative RISK CONTROL apply to the Particular Standards? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Does IVD need to integrate with hospital IT infrastructure? Medical Information Technology, Medical Software and Health Informatics 2
D What does a level 1 (PSW) PPAP actually promise? APQP and PPAP 19
R GS1 DataMatrix code does not meet GS1 Specification EU Medical Device Regulations 0
M Does the scope of ISO 9001:2015 applies to tenders, pricing and sales department of a medical devices distributor? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
K Why does load cell supplier requires force verification General Measurement Device and Calibration Topics 3
M Does C=0 strictly mean 1 bad, all bad, all the time? ISO 13485:2016 - Medical Device Quality Management Systems 6
C Does a medical device active (zinc oxide) needs BPR registration in EU? Other ISO and International Standards and European Regulations 6
D Does Risk Management apply to re-labeler (MDR) EU Medical Device Regulations 1
Similar threads
Top Bottom