Does pitch/increment/resolution of a ruled scale apply to measurement uncertainty as line item?

greif

Involved In Discussions
#1
I have a guy at work that says that the resolution of a lined scale (such as a stage micrometer or steel ruler) should be included as a line item in the uncertainty calculation (because the resolution of the device under test is to be included). While this makes sense for measuring instruments with variable displays, I don't think it does for fixed increment linear scales. He is going to dig up documents that are supposed to back up his position, so I am hoping someone can point me to a specific document (and where in the doc!) that says a linear scale resolution does not apply.

Thanks!
 
Elsmar Forum Sponsor

normzone

Trusted Information Resource
#3
In the absence of any qualified inputs yet, let me play devil's advocate at the moment.

" While this makes sense for measuring instruments with variable displays, I don't think it does for fixed increment linear scales. "

How did you arrive at this conclusion? From my point of view, resolution is resolution, regardless of whether it's analog, digital, or a cubit.

And while we're here, I'll plug a song I like -

 

greif

Involved In Discussions
#4
I guess the main difference is that we are considering the calibration of high quality scales (chrome on glass typically). It is a line standard not a ruler.
Uncertainty calc. Example of why it does not make sense;

(uncertainty line items)
Resolution (of scale) 1mm
Repeatability (of measuring device) 0.001mm
Cal ref std 0.001mm
Meas. Equip cal tol. 0.002mm

K=2 uncertainty from this comes out to be 1.2mm

I am thinking only the resolution of the measuring device should be applied. I need some standard documentation to show this is so.
Help!
 

normzone

Trusted Information Resource
#5
" He is going to dig up documents that are supposed to back up his position, so I am hoping someone can point me to a specific document (and where in the doc!) that says a linear scale resolution does not apply. "

At that point in the game, as usual, the influences to consider are

Industry standards
Regulatory requirements
Organizational procedures

I don't know what team you play for or in what league, but unless there are related requirements in the above three then it's just your preferences vs his.
 

normzone

Trusted Information Resource
#7
Oh, we're talking about what I find ludicrous ...

Well, you did use the words " makes sense ".

For purposes of self preservation I tend to find the world to be ludicrous, so I'm afraid I'm not always the best technical consultant.

But sometimes the area of uncertainty exceeds the information you wish to know. This is what drives us to use different tools, or re-frame the method, or the criteria.

I'd also hoped to draw in a real professional opinion - what is that rule about the fastest way to get the correct answer on the internet is to post the incorrect answer?
 

greif

Involved In Discussions
#8
I had not heard of that particular internet rule; but now that you mention it- very true!

I may have found some evidence to help my case (but nothing as clear cut as "here is the standards doc, and it states such in such at paragraph # xxx".
What I found was a round robin test of a 100mm scale sent to many national metrology centers (doc at: https://www.bipm.org/utils/common/pdf/final_reports/L/K7/EUROMET.L-K7.2006.pdf )

This doc has wonderful descriptions of the machine used, the methodology and the detailed uncertainty stack. Some of the machine are quite similar to our setup (optical microscope to camera to screen). None of them mention the pitch of the scale read!
 

dwperron

Trusted Information Resource
#9
I have a guy at work that says that the resolution of a lined scale (such as a stage micrometer or steel ruler) should be included as a line item in the uncertainty calculation (because the resolution of the device under test is to be included). While this makes sense for measuring instruments with variable displays, I don't think it does for fixed increment linear scales. He is going to dig up documents that are supposed to back up his position, so I am hoping someone can point me to a specific document (and where in the doc!) that says a linear scale resolution does not apply.

Thanks!
If you are using the scale to make a measurement then the resolution of the measurement must be included as a contributor. For instance, if you are using an 8 1/2 digit multimeter to make a measurement but only report out to 0.001 V on your results it is the resolution of the measurement, not the meter, that gets reported.

If you are only reporting your results to the cardinal points on the scale, that is the resolution. If you are interpolating between lines, that is the resolution. Whatever assumptions you make regarding rounding up or down need to be included. The pitch of the gradations should be considered as they might be "significant" in some measurements, and significant is the threshold for reporting uncertainty contributors.
 
Last edited:

greif

Involved In Discussions
#10
I note your words in italics (" then the resolution of the measurement ") and hope that this wording might be a quote of a standard document.
Is it?
This makes my point nicely; we are measuring the scale, and that measurement has an uncertainty. Using the scale to measure has it's own uncertainty.
 
Thread starter Similar threads Forum Replies Date
M Does any one know if company is allowed to submit appeal for rejected registration file Other Medical Device Related Standards 0
B Does TUS require load thermocouples? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
B Does anyone copyright their internal QMS documents? Document Control Systems, Procedures, Forms and Templates 24
D MDR technical file: does it need to contain duplicates of controlled documents EU Medical Device Regulations 2
D Does the MDR article 10 (14) needs to be documented word for word in a Procedure? EU Medical Device Regulations 1
S Why does GAGEpack drop the temperature and humidity readings out of the verification event card? Quality Assurance and Compliance Software Tools and Solutions 0
C Determining an as received OOT condition during calibration does not affect measurements taken prior to calibration. General Measurement Device and Calibration Topics 16
D Does the DoC require a technical documentation version? ISO 13485:2016 - Medical Device Quality Management Systems 1
S Local (country) registration of medical devices - Who does it in best case? Other Medical Device Regulations World-Wide 2
G When does containment event not require 100% sort? Nonconformance and Corrective Action 5
H Does this clause mean i must be in compliance with AS9100 or is 9001 enough? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
H Does ISO state in anyway that we MUST keep physical first piece parts? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
XRAY_3121 Class II heating pad - does the power supply have to have IEC 60601-1 testing?? US Medical Device Regulations 5
F Change in address for critical supplier - does NB need to be informed? CE Marking (Conformité Européene) / CB Scheme 20
S Brexit ISO 13485:2016 + Corrigendum - What does a UKCA DoC require? EU Medical Device Regulations 2
G Does anyone know about tobacco-free nicotine pouches? US Food and Drug Administration (FDA) 2
cgaro62 Does FDA apply to a non-medical 13485 certified custom manufacturing company? ISO 13485:2016 - Medical Device Quality Management Systems 11
R Does anyone use iQMS for their ISO based document control? Manufacturing and Related Processes 1
V How many hrs does it take to create a Technical File (TD)? EU Medical Device Regulations 4
M How does FDA determine OAI and VAI after inspection? ISO 13485:2016 - Medical Device Quality Management Systems 0
G Does TPI agencies comes under outsourced process as per Q1 Oil and Gas Industry Standards and Regulations 11
N FPY requested for an industry that does not have rework IATF 16949 - Automotive Quality Systems Standard 4
William55401 21 CFR 11 Password Aging - What does "periodically" revised mean in your org? Qualification and Validation (including 21 CFR Part 11) 3
I CQI-15 Welding System Assessment 2nd edition - Does not require Ultrasonic Welding anymore? Customer and Company Specific Requirements 0
Ajit Basrur Does "Refurbishment" fall under "Service"? ISO 13485:2016 - Medical Device Quality Management Systems 13
T Does marketing company require CE mark if manufacturer has CE mark on product? EU Medical Device Regulations 5
Watchcat REGULATORY WATCHCAT De Novo Post-Mortem 6 – How Does This Even Happen? Other US Medical Device Regulations 0
T What does VW mean by annual requalification? Customer and Company Specific Requirements 4
I Does BSI require suppliers to be ISO 9001 Certified? EU Medical Device Regulations 12
Watchcat Does Regulatory Affairs put the "no" in innovation? Other US Medical Device Regulations 14
Q Does anyone know what this call out means? Manufacturing and Related Processes 3
K COPLANARITY: Composite profile tolerance on multiple surfaces- what does" lower dimensional reference frame tolerance" control? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
P Does FDA require certification for quality system internal audit for auditor? Qualification and Validation (including 21 CFR Part 11) 1
H Does anyone has feedback on a notified body for small companies? EU Medical Device Regulations 3
M Does 4.5 - Alternative RISK CONTROL apply to the Particular Standards? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Does IVD need to integrate with hospital IT infrastructure? Medical Information Technology, Medical Software and Health Informatics 2
D What does a level 1 (PSW) PPAP actually promise? APQP and PPAP 19
R GS1 DataMatrix code does not meet GS1 Specification EU Medical Device Regulations 0
M Does the scope of ISO 9001:2015 applies to tenders, pricing and sales department of a medical devices distributor? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
K Why does load cell supplier requires force verification General Measurement Device and Calibration Topics 3
M Does C=0 strictly mean 1 bad, all bad, all the time? ISO 13485:2016 - Medical Device Quality Management Systems 6
C Does a medical device active (zinc oxide) needs BPR registration in EU? Other ISO and International Standards and European Regulations 6
D Does Risk Management apply to re-labeler (MDR) EU Medical Device Regulations 1
Ed Panek Does this FDA Requirement Apply to international (not USA) distributors for USA based manufacturing companies? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
S Does anyone have a checklist to prepare for ISO 13485, Stage I audit? ISO 13485:2016 - Medical Device Quality Management Systems 3
S Does a refurbished product required a new UDI? US Food and Drug Administration (FDA) 3
D Change to labelling - does it require a new 510(k)? US Food and Drug Administration (FDA) 5
M What does "constantly" mean ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
N Does anyone use SGS for ISO 13485 / CE certification Registrars and Notified Bodies 0
G Does FDA allows remote approvals of quality documentation. Is there any specific guidance on signing any quality records remotely? Document Control Systems, Procedures, Forms and Templates 1

Similar threads

Top Bottom