Dose Consistency of Insulin Formulations

S

sreeranjini

#1
Hello,

I am working on an insulin delivery device. We came across a requirement for checking the consistency of the dose delivered from the device.
Is this a regulatiry requirement? If yes, what should be the test methodology or strategy of conducting this test.

Thanks
Sree
 
Elsmar Forum Sponsor

v9991

Trusted Information Resource
#2
i am not into devices/insulin mfg but here's 1 reference from FDA,...
but here';s starting point,
"APDS control algorithms currently recommend periodic insulin dosing with frequencies ranging from 1-30 minutes. These frequent bolus doses are extremely small and ask the pump to perform accurately near the lowest doses available in the pump. In order to understand how well the APDS can deliver the recommended insulin or drug, bolus dose accuracy testing should be performed. This testing should evaluate the APDS using the most frequent dosing rate (i.e., the shortest time between dosing adjustments) and accuracy measurements should be tested using the minimum dose, maximum dose and incremental doses between the min and max. The testing of this system should mimic the conditions of the clinical study and the measurement technique should account for evaporation of small doses during the testing. Temperature should be controlled and reflect the use environment."
http://www.fda.gov/MedicalDevices/DeviceRegulationandGuidance/GuidanceDocuments/ucm281706.htm

also refer...sec. E.1
FDA recommends completion of the following dose accuracy testing.
? Testing to demonstrate that the volume/weight of drug/biological product expelled through the injector is the same as the set dose
? Testing to ensure that multi-dose cartridge injectors satisfy the requirement
that each successive dose is the same as the first set dose
? Testing to ensure that dose settings/markings correlate with the volume of drug/biological product delivered

Because diluents may affect the drug/biological product and the injector, the above testing should ensure that the delivery volume meets the dose accuracy specification for the specific drug/biological product when reconstituted according to its approved or proposed labeling, as applicable.

We recommend, where appropriate, that injectors include a mechanism or feature that notifies the user that an incorrect dose has been delivered. Performance testing should include testing of such mechanism(s). Product labeling should also include information instructing the user on what action should be taken in the event of incorrect dosage.

http://www.fda.gov/downloads/RegulatoryInformation/Guidances/UCM147095.pdf

hope this helps...
just in case if you have not already seen this...
 

Ajit Basrur

Staff member
Admin
#3
Homogeneity within a batch and consistency between batches are goals of Process Validation activities. Hence if you say that your process is validated, this becomes mandatory.
 
S

sreeranjini

#4
Homogeneity within a batch and consistency between batches are goals of Process Validation activities. Hence if you say that your process is validated, this becomes mandatory.
Dear Ajit,

We have already established consistency between different batches of the formulation. So I assume we have validated our process. Please correct me if I am wrong :).
My question was regarding the consistency of the doses delivered from the device. Whether it should be based on the volume or weight of the drug delivered or do we need to conduct any chemical analysis?

Thanks
Sree
 
S

sreeranjini

#5
Dear V9991,

Thank you for the valuable guidance considering you are not into this particular field. I have seen this guidance, but in this guidance, "• Testing to ensure that multi-dose cartridge injectors satisfy the requirement that each successive dose is the same as the first set dose" this line confused me. We are checking the accuracy of the dose delivered. But I wanted to know whether it has to be the drug product or the device that should be analysed here.

Thanks
Sree
 
Last edited by a moderator:
I

isoalchemist

#6
You should be looking at the device. You are really trying to show that the delivery is consistent if the vial is full, half full, or nearly empty and when you change vials it stays the same.

Weight or volume come down to whatever is easier and more accurate to measure since they are related by specific gravity.
 

v9991

Trusted Information Resource
#7
Re: Dose Consistency of Insulin Formulations ...

If you are not looking for specific methods & numbers ... let try to put my understanding forward......(typically what we do in validation procotols in similar context)

Protocol/study plan :-
a) identify the range of device characteristics (basically batch to batch variation of its accuracy, precision etc.,)
b) identify the range of product characteristics which device has to handle... ( different levels of pH, viscisity, assay, CU or any other relevant/specific parameter )
c) arrive at an test-method(sequence viz., device operating conditions, settings, ) and acceptance criteria (of accuracy and precision.... )

b) & c) are usually an extension(matrix) of product requirements and device performance capabilities.(as selected by you or as stated by vendor)

while execution/implementing the protocol ...
* first characterize the batch at the required levels as identified above
* execute the experiment.
* analyze the data.
*conclude (addressing the recommendations, / deviations etc)

hope this helps.


further,
We have already established consistency between different batches of the formulation. So I assume we have validated our process. Please correct me if I am wrong
yes, your formulation is consistent not the product/process!!!
for us product is formulation+device; hence consistency of formulation alone wouldn't really be complete w/o addressing rest of the part as well.!!!
 

Ajit Basrur

Staff member
Admin
#8
Dear Ajit,

We have already established consistency between different batches of the formulation. So I assume we have validated our process. Please correct me if I am wrong :).
My question was regarding the consistency of the doses delivered from the device. Whether it should be based on the volume or weight of the drug delivered or do we need to conduct any chemical analysis?

Thanks
Sree
You should comply per stated and labeled claim.
 
Thread starter Similar threads Forum Replies Date
Judy Abbott Increase the maximum allowable dose for E-beam sterilization Reliability Analysis - Predictions, Testing and Standards 2
E Unit Dose Packaging Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
M Informational MHRA Guidance – Dental radiographic X-ray imaging: dose to patients Medical Device and FDA Regulations and Standards News 0
K Calibration of Dose Area Product (DAP) Meter in our Class IIb X-Ray Medical Device General Measurement Device and Calibration Topics 1
F Irradiation Sterilization Validation Maximum Dose - VDmax 25 method (ISO 11137) EU Medical Device Regulations 2
B Radiation Sterilization Dose Audit: Bioburden frequency Other Medical Device Related Standards 1
R ISO 11137 Gamma sterilization validation, dose setting, dose audit Other Medical Device Related Standards 1
H Change of a medical device sterilization dose - New 510(k) necessary? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
H Gamma Irradiation - dose audit and source replenishment Other Medical Device Related Standards 13
K Bioburden 0 cfu for Dose Setting using the VDmax25 Other Medical Device Related Standards 9
M ISO 11137 - Verification dose audit for VD max method Other Medical Device Regulations World-Wide 7
alonFAI Quarterly Dose Audit Medical Device VDmax Other Medical Device Related Standards 2
R Irradiation Sterilization Dose Setting and Dose Audit Flow Charts ISO 13485:2016 - Medical Device Quality Management Systems 0
C Dose Mapping - How to relate back to Routine Monitoring Position ISO 13485:2016 - Medical Device Quality Management Systems 3
P Quarterly Dose Verification for e-beam Sterilization Question ISO 13485:2016 - Medical Device Quality Management Systems 1
N Gamma Validation vs. Quarterly Dose Audit Differences ISO 13485:2016 - Medical Device Quality Management Systems 3
M Setting Solid Dose Drug Fill Count Ranges Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 8
K Gamma Ray Sterilization Quarterly Dose Auditing Supplier Requirements ISO 13485:2016 - Medical Device Quality Management Systems 5
A Help needed for Fighting a specifically identified Organism found during Dose Audit Miscellaneous Environmental Standards and EMS Related Discussions 2
E Auditing sterilization dose substantiated Vdmax 25 Other Medical Device Related Standards 4
G Dose Audit Requirements Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 7
I Quarterly Gamma Dose Audit Question Other Medical Device Related Standards 4
T Missed Quarterly Dose Audit! Sterilization Validation of Medical Devices Other Medical Device and Orthopedic Related Topics 15
M Radiation (Gamma or E-beam) Dose Audit SOP - Quarterly dose audits based on ISO 11137 Other Medical Device Related Standards 8
E ISO 11137 gamma sterilized device dose audit - Lead time? Other ISO and International Standards and European Regulations 7
T Sterilization Dose Audit - Recommended Quantity of Samples/Batches ISO 13485:2016 - Medical Device Quality Management Systems 1
M Pharmaceutical Non conformances SOP - Solid oral dose start up company Nonconformance and Corrective Action 2
J Nuclear Medicine Dose Calibrator Classification (MDD) ISO 13485:2016 - Medical Device Quality Management Systems 5
D US-EU Medical Device Labeling Consistency 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S Consistency of Laboratory results between different labs US Food and Drug Administration (FDA) 2
B New ISO 10993-12 - Consistency of Extraction Conditions through different Assays Other ISO and International Standards and European Regulations 1
S Consistency in Organization Documents Document Control Systems, Procedures, Forms and Templates 5
L How to achieve Inspector Consistency? Specialty vehicle manufacturer Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
K Variations in Tool Use Techniques and How to Improve Consistency? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
H Auditor Opinion and consistency - Each auditor has different 'opinions' ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
S Difference between Stability and Consistency Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
RoxaneB Lack of Registration Body Consistency in Transitioning ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
B ISO certification - consistency or quality?? General Auditing Discussions 3
M Determining if an Insulin Pen Testing Machine is a Medical Device? EU Medical Device Regulations 4
M Classifying Software - Insulin Calculator EU Medical Device Regulations 13
S Insulin Pen Injector Storage Requirements Other Medical Device and Orthopedic Related Topics 1
W Measuring the Dial Torque of Plastic Insulin Pen devices Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Ronen E Covidien recalls some Monoject Insulin Safety Syringe lots Other Medical Device and Orthopedic Related Topics 0
B What kind of insulin syringe is this for? ISO 8537:2007 Section 8, Table 1 Other Medical Device Related Standards 1
Ajit Basrur My iPod, My Cell, My Insulin Pump - Advances of Diabetics Management ISO 13485:2016 - Medical Device Quality Management Systems 1