Early Research & Development - ISO 13485:2003 requirements Clarification

  • Thread starter TempusVernumProcella
  • Start date
T

TempusVernumProcella

#11
Re: R&D 13485:2003 Clarification...

Tempus,

I've highlighted your words that indicate to me that you are trying to impose your will on your R&D team instead of listening to them.

I agree with Laura's advice on this in #7 and Somashekar's in #8.

To whom does the QMS belong? How is the QMS fulfilling the needs of the R&D team. To paraphrase Edision, many R&D failures may be necessary for their eventual success.

They may already realize the importance recording what does not work but see the recording methods mandated by your QMS as unhelpful.

Lastly, periodic multidisciplinary design reviews for resulting in effective design output may not be appropriate to the sand-box or skunk works activities at the front-end of truly innovative R&D.

John


All excellent points, to be sure. Here's where I'm getting confused, though- ISO 13485:2003, 7.3 Design and Development states that "the organization shall plan and control the design and development of product.
During the design and development planning, the organization shall determine
a) the design and development stages,
b) the review, verification, validation and design transfer activities that are appropriate to each design and development stage, and
c) the responsibilities and authorities for design and development.
The organization shall manage the interfaces between different groups involved in design and development to ensure effective communication and clear assignment of responsibility.
Planning output shall be documented, and updated as appropriate, as the design and development progresses (see ISO 13485:2003, 4.2.3 Control of Documents)."


So... having regulated documents IS necessary... :frust:
 
Elsmar Forum Sponsor

Marcelo

Inactive Registered Visitor
#13
All excellent points, to be sure. Here's where I'm getting confused, though- ISO 13485:2003, 7.3 Design and Development states that "the organization shall plan and control the design and development of product.
During the design and development planning, the organization shall determine
a) the design and development stages,
b) the review, verification, validation and design transfer activities that are appropriate to each design and development stage, and
c) the responsibilities and authorities for design and development.
The organization shall manage the interfaces between different groups involved in design and development to ensure effective communication and clear assignment of responsibility.
Planning output shall be documented, and updated as appropriate, as the design and development progresses (see ISO 13485:2003, 4.2.3 Control of Documents)."


So... having regulated documents IS necessary...
And that?s the problem qith most of the comments.

The thing is, the standard has not defined WAY to control research. But it does not mean that it expects that research is done in a total chaotic way.

If you go to the first "control", it is for design inputs. Those are the product requirements (so you already has decided in which product to develop). After that, there?s no specific way to control developt itself (besides design review at stages), and then design output.

So, there?s a little step-by-step control scheme after design input.

There?s no step-by-step control scheme prior to that (research), which means that you can control the way you want. There?s no specific requiremet, such as "resarch input" or "research output".

But that does not mean that you should not include some, even as a way to manage how thngs are done.

That?s why I mentioned in the comment before, if you defined that research has to follow some steps, those would became part of your QMS (even if the standard does not require specific controls).
 

Marcelo

Inactive Registered Visitor
#14
Also, you should take care because, a lot of times people THINK they are doing only research, they already are doing some development (which should follow the mentioned step-by-step controls). One examle is when people create prototypes, a lot of times people create initial prototypes then use then further down the development process..so, since the beginning, those would need some control.
 
T

TempusVernumProcella

#15
Also, you should take care because, a lot of times people THINK they are doing only research, they already are doing some development (which should follow the mentioned step-by-step controls). One examle is when people create prototypes, a lot of times people create initial prototypes then use then further down the development process..so, since the beginning, those would need some control.


Marcelo,
Thank you for your feedback. :agree1: I know for fact that initial design phases for prototype's are indeed cataloged/recorded as soon as an idea becomes a physical manifestation. The grand issue, unfortunately, is the bureaucratic paperwork... Our QMS, that was written by a previous QMS manager, states that R&D starts with an idea and then must be logged; which is, in my opinion, highly unrealistic. Because of this current process, R&D is 'falling out of compliance'. So, yes, restructuring our current system and working with R&D to effectively generate a functional process is most necessary.
 

John Broomfield

Staff member
Super Moderator
#16
Last edited:
M

MIREGMGR

#18
I read the thread from the beginning, and my initial impression was that the "R&D" people were product creators/developers who just wanted to avoid having to do any documentation work, or as little as possible.

The last few posts seem to have gone in that direction.

My view is that, except maybe in academic labs and a few pure-science organizations, "R&D" is just early stage product development.
 

Ronen E

Problem Solver
Staff member
Moderator
#19
Also, you should take care because, a lot of times people THINK they are doing only research, they already are doing some development (which should follow the mentioned step-by-step controls). One examle is when people create prototypes, a lot of times people create initial prototypes then use then further down the development process..so, since the beginning, those would need some control.
There should be a clear product development project kick-off. At that point anything brought forward from the Fuzzy Front End (FFE) phase should be documented. This can be done retroactively. Sometimes prototypes are even rebuilt, in a more orderly way, so that a good documented and controled basis exists for the next phases. This is not necessarily a bad thing.
 
Last edited:

somashekar

Staff member
Super Moderator
#20
Marcelo,
Thank you for your feedback. :agree1: I know for fact that initial design phases for prototype's are indeed cataloged/recorded as soon as an idea becomes a physical manifestation. The grand issue, unfortunately, is the bureaucratic paperwork... Our QMS, that was written by a previous QMS manager, states that R&D starts with an idea and then must be logged; which is, in my opinion, highly unrealistic. Because of this current process, R&D is 'falling out of compliance'. So, yes, restructuring our current system and working with R&D to effectively generate a functional process is most necessary.
You have such a good point here...
Research feeds design and development. Design and development feeds the Purchasing and the product & service provision.
There is no fine line dividing the Research and D&D. It is how you set up your QMS and how you want it to work for you effectively.
All you need are :
A Design plan, updated as appropriate, as the design and development progress.
Approved design inputs, which are complete and unambiguous. (Do not miss the risk management)
The Design outputs (As in note of 7.3.3 of the ISO 13485)
Records of the results of the design reviews, and necessary actions thereof
Records of the results of the design verification and necessary actions thereof
Records of the results of the design validation and necessary actions thereof
 
Thread starter Similar threads Forum Replies Date
J EDMS for Very Early Drug Development (Research/Drug Discovery) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 4
P IEC 62304:2006 A1:2015 - Software from the early 1990s IEC 62304 - Medical Device Software Life Cycle Processes 4
M Informational FDA Panel: Too early to pull textured breast implants over cancer risk, need more data Medical Device and FDA Regulations and Standards News 0
Sidney Vianna Interesting Discussion World wide energy demand will peak in the early 2030's Sustainability, Green Initiatives and Ecology 0
Q Is EN 55011:2009+A1:2010 is expiring in early 2019? Other ISO and International Standards and European Regulations 1
Marc Air Pollution in the Early 21st Century Coffee Break and Water Cooler Discussions 5
S Final FDA Guidance on IDEs for Early Feasibility Clinical Studies - 2013 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
B Best Attribute Data Sampling Plan - Early Production Containment Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
K Tips 'n Tricks for getting a Job in the early 21st Century Career and Occupation Discussions 1
Q Reduced Device License Fee was rejected for being sent in too early Canada Medical Device Regulations 2
G ISO/DIS 14045 - Anyone have an early draft copy? Miscellaneous Environmental Standards and EMS Related Discussions 2
V Evidence for Preventive Action & Early Warning Vigilance procedure Preventive Action and Continuous Improvement 4
J Today it's 'Leave The Office Early' Day Career and Occupation Discussions 2
P Early versus Late delivery from suppliers - what's the acceptable ratio? Quality Tools, Improvement and Analysis 6
Stijloor ASQ Membership Renewal Request Comes Early - Will you renew?? ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 35
M Risk Management - AS9100 revision C - Early advice please! AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
A Juran is dead - Why the thread is closed so early? Philosophy, Gurus, Innovation and Evolution 2
Jim Wynne Never believe early news reports Coffee Break and Water Cooler Discussions 7
Marc When Flying, How early do you get to the airport? Travel - Hotels, Motels, Planes and Trains 25
Q Early Spring Cleaning - Finding Old Documents and Specifications ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
B Retiring Early or Never? Coffee Break and Water Cooler Discussions 65
T "Creative" ways to handle Early Production Containment Misc. Quality Assurance and Business Systems Related Topics 9
Marc GM in early talks to buy Chrysler World News 6
C Graph Choice For Monitoring DOA's (Dead On Arrival) and ELF's (Early Life Failures) Statistical Analysis Tools, Techniques and SPC 5
T AS9003 becoming obsolete as early as 2008??? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
D An Early Brush with Design of Experiments Design and Development of Products and Processes 7
M Aspect Indentification? Early stages of ISO 14001 implementation ISO 14001:2015 Specific Discussions 4
C Reasons for early termination of an audit? General Auditing Discussions 23
J GM's GP-12 Early Production Containment applicable to Tier 2 Suppliers? Customer and Company Specific Requirements 14
Marc BP Meets ISO 14001 Targets 8 Years Early Miscellaneous Environmental Standards and EMS Related Discussions 5
Marc Containment FMEA and GP-12 (GM procedure for early production containment) FMEA and Control Plans 9
Marc The Future of Quality - An Early Discussion Philosophy, Gurus, Innovation and Evolution 11
D Question on equipment - when to use reference only or research only stickers ISO 13485:2016 - Medical Device Quality Management Systems 5
L Clinical research - Request from ANVISA ISO 14971 - Medical Device Risk Management 1
Watchcat Authoritative References about the Research Question? Quality Tools, Improvement and Analysis 0
A VDmax25 and cGMP requirements for "research use only" products Other Medical Device Related Standards 1
M Risk Classification For Supplier - Clinical Research Organisation (CRO) Supply Chain Security Management Systems 3
M Exemption Clauses for Research, Demonstration, etc. Other Medical Device Regulations World-Wide 2
N Time source for paper-based documentation (research nurses) US Food and Drug Administration (FDA) 1
M Medical Device News CDRH Research Programs – VICTRE: Virtual Imaging Clinical Trials for Regulatory Evaluation Medical Device and FDA Regulations and Standards News 0
S Medical Device Research Use Only (RUO) label Other Medical Device Regulations World-Wide 4
M Acceptance of remote auditing techniques - Can you help me with my research? General Auditing Discussions 0
N Dataset for Academic Research - Infusion pump flow IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T Research Medical Device EU - Requirements and Regulations EU Medical Device Regulations 13
Y Change Management in a very complex Research Organization Misc. Quality Assurance and Business Systems Related Topics 3
L EU MDR CRO (Contract Research Organization) requirements for Clinical Investigations EU Medical Device Regulations 1
T Control of R&D (Research and Development) Parts and Equipment ISO 13485:2016 - Medical Device Quality Management Systems 3
I AS9100D - Interview Request - I'm doing research Misc. Quality Assurance and Business Systems Related Topics 1
I Design & Development, Research and Innovation in the Medical Device framework ISO 13485:2016 - Medical Device Quality Management Systems 3
R University Research Project - Management Review Management Review Meetings and related Processes 17