Earthing system (IT, TT or TN) and IEC 60601 Requirements

F

fonk140

#1
Does the type of earthing system used in a hospital (TT, IT ot TN) affects the electrical safety of the medical device?

As far as i know the 60601 never specifies the type of earthing system but for example in a TT connection a short between mains and earth maybe wouldn't blow the fuse of the device doesn't it?

I don't know if what im saying makes sense or not but i would appreciate any comments.

Thanks in advance!
 
Elsmar Forum Sponsor

Peter Selvey

Leader
Super Moderator
#3
IEC 60601-1 assumes the worst case, for example leakage currents assume an earthed neutral which results in the highest values.

If the mains is floating from earth, and there is a Line - Earth fault in the equipment, this is not a hazard since it just pulls the line to earth potential which should be the same potential as other earthed objects in environment (other equipment, plumbing, ground).

If the environment earthing is not reliable, and a fault occurs, it could be an issue.
 
Thread starter Similar threads Forum Replies Date
S Earthing of moving parts, designing tips IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Protective Earthing of Medical Device Moving Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J Protective Earthing for Medical Device with two wire power cord in metal enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
A Document control system for QMS in Google Drive? Document Control Systems, Procedures, Forms and Templates 6
M On-demand EU MDR Compliance toolkit with online training courses, templates & document management system Training - Internal, External, Online and Distance Learning 0
S ISO 9001 Quality Management System and its processes ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
P Learning Management System (LMS) recommendations? Training - Internal, External, Online and Distance Learning 2
T SaMD or Software system? EU Medical Device Regulations 2
D Conformity Assessment for a System Class IIb EU Medical Device Regulations 0
T Determination of Software/system end-of-life IEC 62304 - Medical Device Software Life Cycle Processes 5
H A better Lot/bin system Preventive Action and Continuous Improvement 2
S Environment Monitoring System Validation ISO 14001:2015 Specific Discussions 1
R Clinical evaluation for instrument of a navigation system EU Medical Device Regulations 1
A Non-medical device testing in the medical system IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D Repeat Build of a validated system ISO 13485:2016 - Medical Device Quality Management Systems 2
A LEDs in the system to indicate a status or fault IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Surveillance Audit for AS9100D and new ERP system incorporated Document Control Systems, Procedures, Forms and Templates 4
A ISO 9001- any advise please with transitioning from a paper based system to eQMS and multi site scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
PQ Systems 10 Ways to Improve Your SPC System Using SQCpack Software 0
R Quality System Software AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
D Integrated Management System (IMS) - Process Vs Clause Based Audits ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
I Request for information regarding remote medical monitoring software (its technical documentation and the IUD system) IEC 62304 - Medical Device Software Life Cycle Processes 2
Q The Management system is too much ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A Recommendations for affordable CAPA system Preventive Action and Continuous Improvement 4
B Classification of screening system EU Medical Device Regulations 14
G Software verification vs. system verification IEC 62304 - Medical Device Software Life Cycle Processes 3
D Electronic Document Management System ISO 13485:2016 - Medical Device Quality Management Systems 16
N Computerized System Validation ISO 13485:2016 - Medical Device Quality Management Systems 12
R First Time Managing Calibration and Measurement System. Need Help. General Measurement Device and Calibration Topics 19
F Protection from mains in a ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
T What is a Quality Management System? Manufacturing and Related Processes 20
W Posting Non-Quality System instructions on the Production Floor Manufacturing and Related Processes 13
U FSC Due Diligence System (DDS) summary report... Sustainability, Green Initiatives and Ecology 0
T MDR PMS System Vs. Plan Vs. Procedures EU Medical Device Regulations 3
C. Tejeda Computer system validation approach for Minitab Statistical software Software Quality Assurance 11
M Sterile barrier system with outer protective packaging Symbol EU Medical Device Regulations 11
L ISO 11607-1 Packaging system validation Design and Development of Products and Processes 9
Sidney Vianna IATF 16949 News IATF Performance Complaint Management System (IATF CMS) within the IATF Database IATF 16949 - Automotive Quality Systems Standard 4
J Cloud Based System Qualification and Validation (including 21 CFR Part 11) 7
E Verifying our build system IEC 62304 - Medical Device Software Life Cycle Processes 2
D Class III license: Add new device to the license system Canada Medical Device Regulations 1
PQ Systems Recipe for Success! 10 Ingredients for a Healthy Gage Management System Using GAGEpack Software 0
T ISO quality system software such as MQ1 (which is what we currently use) Document Control Systems, Procedures, Forms and Templates 9
I CQI-15 Welding System Assessment 2nd edition - Does not require Ultrasonic Welding anymore? Customer and Company Specific Requirements 0
A Product Update and executing only affected System Tests, leaving out unaffected ones IEC 62304 - Medical Device Software Life Cycle Processes 5
L Language of quality system in case of FDA inspection US Food and Drug Administration (FDA) 1
blackholequasar Validation of new ERP system ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
B Vision system process validation Manufacturing and Related Processes 1
W Strategy for determining which components from a system should be "ME EQUIPMENT" -- home healthcare environment IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
E Marketing, product and system requirements IEC 62304 - Medical Device Software Life Cycle Processes 2

Similar threads

Top Bottom