Effective Date on Documents

tebusse

Involved In Discussions
#1
All,

Please correct me if I'm wrong - within 21 CFR 820, ISO 13485:2016, and the MDD, I cannot find that having the effective date on a document is required. All that is required is that the document was reviewed and approved, has an identifier (e.g. name, number, revision level, etc.).

Here's my conundrum...

We currently place effective dates on our documents. However, we utilize document management software that appends a page at the end of each document with signature of the reviewers, the document approval date and the publication date. I would think that the publication date can serve as the effective date. Otherwise, we always have to carefully coordinate the effective date on the document and the publication within the software system. Please note, reviewers applying their signatures are in a queue. If someone doesn't approve the document until after the effective date we outline on the document, the process has to start over.

Please advise.

Tonia
 
Elsmar Forum Sponsor

Ninja

Looking for Reality
Staff member
Super Moderator
#2
A curiosity...
Is there a requirement that the effective date must be after all signatures?

I am not in medical, nor was ever under 21CFR820...but under ISO9001 and TS16949, I had effective dates months in advance of signatures in some cases...

The root of my question is: Are you creating your own problem?
 

tebusse

Involved In Discussions
#3
A curiosity...
Is there a requirement that the effective date must be after all signatures?

I am not in medical, nor was ever under 21CFR820...but under ISO9001 and TS16949, I had effective dates months in advance of signatures in some cases...

The root of my question is: Are you creating your own problem?
Ninja,

The way the document management system software works is that a document is routed for review, reviewers approve (e.g. signatures) and then the document is published once all reviewers have approved it. Please note, the publication is not automatic once the final reviewer signs. Document control my manually publish the document within the system.
 

Ninja

Looking for Reality
Staff member
Super Moderator
#4
Understood...but you didn't answer the question...

Is there a requirement that the effective date be after all signatory dates?

Wording it another way...
You have an effective date of November 1, 2018.
You have two signatures: one on 10/25/18, one on 11/25/18...
what is the problem with the effective date being between those two dates?
Where does the "It must be after all signature dates" requirement come from?
 

tebusse

Involved In Discussions
#5
For me it's a moot point as our electronic document system doesn't allow signatures to be placed after the document becoming effective.
 

DEVigil

Involved In Discussions
#6
Understood...but you didn't answer the question...

Is there a requirement that the effective date be after all signatory dates?

Wording it another way...
You have an effective date of November 1, 2018.
You have two signatures: one on 10/25/18, one on 11/25/18...
what is the problem with the effective date being between those two dates?
Where does the "It must be after all signature dates" requirement come from?
That depends.

ISO 13485:2016 4.2.4 requires that documents shall be reviewed and approved for adequacy prior to issue. Approval should come from someone with access to pertinent background information upon which to base the decision. So, if the signature is of a required approver, the document cannot be effective before it's been approved by them. This is mirrored in the language in 21 CFR 820.40.

However, some organizations have other functions approve documents as well (e.g., it has to pass through a more administrative QC-type review to make sure the format is correct, etc.). If it is an approval that is not related to the content, then you could argue the document was effective when approved by the required approving function.
 
Last edited:

Ninja

Looking for Reality
Staff member
Super Moderator
#7
our electronic document system doesn't allow signatures to be placed after the document becoming effective.
Got it...it is your software imposing the requirement.
In that case, it seems like you need a business system of "forewarned and expected" in order to get around the software imposed restriction.
have everyone notified and in agreement before you release it through the software.

reviewed and approved for adequacy prior to issue.
Don't mean to be semantic or pedantic...but "issue" and "effective as of" are two different things...aren't they?

As said many times, I am not in medical...I understand that words may mean different things in a different field.
 

DEVigil

Involved In Discussions
#8
Don't mean to be semantic or pedantic...but "issue" and "effective as of" are two different things...aren't they?

As said many times, I am not in medical...I understand that words may mean different things in a different field.
Yes, but I have trouble envisioning how something could be effective before it was issued, so it is both pedantic and moot.
 

Ninja

Looking for Reality
Staff member
Super Moderator
#9
I have trouble envisioning how something could be effective before it was issued
The document effective before issue...can't happen (-ish).
The process documented is often effective before the document is issued (at least in non-med).
It has to exist before you can document it.

We find a better way to do things...so we start doing them that way immediately...why wouldn't we?
A waiver is written to allow this deviation.
Then the document is updated to reflect the new process tweaks...effective date is the day we started using the new process, which is prior to the document catching up. Thus "Issue date" is often later than "effective date"....hardly moot, but I'll agree on pedantic.

I suppose folks handle the concept of "effective date" differently...I always put it at when the process being documented became effective.
 
Thread starter Similar threads Forum Replies Date
A Effective Date (of Documents) issue outlined in 21 CFR Part 820.4 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
I What is the proper date to use as effective date? Old 'procedures', New documents Document Control Systems, Procedures, Forms and Templates 10
Q Predating Document Effective Date Other Medical Device and Orthopedic Related Topics 18
D Effective Date of updated SOP Document Control Systems, Procedures, Forms and Templates 3
B Document Effective Date vs. Implementation Date Document Control Systems, Procedures, Forms and Templates 7
D What document will indicate the date the QMS became effective? Misc. Quality Assurance and Business Systems Related Topics 4
A PRRC effective role and activities EU Medical Device Regulations 1
T How you ensure that QMS is effective and efficient? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
S Effective nonconforming process for AS9100 8.7 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
M Informational TGA presentation: How to submit an effective good manufacturing practice clearance application Medical Device and FDA Regulations and Standards News 0
N Validation procedure, Major NC CAP - Equipment Validation process is not effective ISO 13485:2016 - Medical Device Quality Management Systems 5
A Cost effective thread gaging with ISO 9001 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S Root cause analysis - The system of developing a work instruction is not effective Problem Solving, Root Cause Fault and Failure Analysis 1
qualprod Effective Auditing advice needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
Q Simple and effective way to maintain Training Matrix ISO 13485:2016 - Medical Device Quality Management Systems 16
J If Corrective and Preventive Action were truly Effective IATF 16949 - Automotive Quality Systems Standard 3
qualprod Risk closeout , mitigation was not effective, next? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
J Specific Actions where Timely and/or Effective Actions are not Achieved Nonconformance and Corrective Action 6
P Thai FDA postpone AMDD to be effective in 2020 Other Medical Device Regulations World-Wide 1
S Sell a Non-RoHS device, manufactured before RoHs regulations became effective? EU Medical Device Regulations 2
J What does "effective implementation of quality management system" mean? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
N What does an effective design process look like? (shoots flare in the air) Design and Development of Products and Processes 2
M Corrective Action - For a corrective action to be effective, a change is required? Nonconformance and Corrective Action 5
R 5.4.1 Quality Objectives - Nonconformity: The process was not fully effective ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 36
R Training to Effective Documents - Transitioning to Agile Document Control Systems, Procedures, Forms and Templates 5
R Good Supply Practices that will be effective on March 1 2015 in China China Medical Device Regulations 2
K IATF has posted new GM Customer Specifics, effective January 2015 Customer and Company Specific Requirements 5
B Medical Device Labeling - Order No. 6 Effective Oct 01, 2014 China Medical Device Regulations 7
R CFDA updated five regulations to be effective from October 1, 2014 China Medical Device Regulations 16
T Qualification System of Internal Auditors is not Effective General Auditing Discussions 5
A A Guide to Effective Internal Management System Audits Book, Video, Blog and Web Site Reviews and Recommendations 2
P Effective way for DC/DC conversion for 2 MOPP @ 250VAC IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J Want best and effective solution for mouth infection Coffee Break and Water Cooler Discussions 11
S New FAA Form 8130-3 Effective 2/1/14 - Unlocked version attached Federal Aviation Administration (FAA) Standards and Requirements 3
P Effective Road Usage Coffee Break and Water Cooler Discussions 9
H How to make the Design Review Process Effective Design and Development of Products and Processes 1
Q Effective 1st Piece Inspection Sign Off Procedure Quality Manager and Management Related Issues 8
QMMike What is the most cost effective SPC software Statistical Analysis Tools, Techniques and SPC 5
Y Delivering an effective and all inclusive EHS program in a Food company Food Safety - ISO 22000, HACCP (21 CFR 120) 2
F Suggestions for quick & effective solution for Continuous Improvement Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
T Very effective movie about clean-up - Tool for 5S Training Lean in Manufacturing and Service Industries 4
M AS 9100 4.2.4 Process for making changes to records is not fully effective Quality Manager and Management Related Issues 4
Q Process Approach & Continual Improvement Clauses and Evidence of its Effective Use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
M 8.2.2 Internal Audit - Determination of Effective Implementation IATF 16949 - Automotive Quality Systems Standard 4
L Audit NC - Lack of Effective Corrective Action Process ISO 13485:2016 - Medical Device Quality Management Systems 4
P Which Root Cause Analysis Method is the most Effective? Quality Tools, Improvement and Analysis 16
J Cost Effective Solution for Isolating USB 5V 500mA Power IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Q Are ECO's closed after they are shown to be Effective? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
V What are the Steps in Ensuring a Definition/Implementation of an Effective Procedure Document Control Systems, Procedures, Forms and Templates 3
P What is Effective Maintenance? Auditing a Maintenance Department Manufacturing and Related Processes 5