Dear All,
I'd like to clarify with you the applicability of 60601 to this medical device:
it is a syringes filling equipment which is battery operated.
It has been designed in order to help the user (operator/patient) to automatically fill a syringe (50ml).
In my opinion this is not covered by the standard IEC 60601-1 because it
So for these reasons I think it is not in the scope of IEC 60601-1
Do you agree?
The situation could be different if during its use the patient should hold it but this is not the case.
So I'm wondering which is the standard to be used in order to ensure the electrical safety of this device
Thank you in advance to anyone who can help me
Regards!
G.
- doesn't have applied parts
- doesn't transfer energy to/from patient
Do you agree?
Regards!
G.